![]() |
|
![]() |
||||
Thread | Thread Starter | Forum | Replies | Last Post |
qPCR reagent recommendation | HICORE | Sample Prep / Library Generation | 2 | 08-29-2011 01:55 PM |
Sequencing reagent stability | AciBASE | 454 Pyrosequencing | 6 | 03-02-2011 03:27 AM |
library reagent units and conc | niceday | Sample Prep / Library Generation | 0 | 02-15-2011 05:45 AM |
Reagent costs per run | gavin.oliver | General | 1 | 02-12-2010 02:40 AM |
GS-FLX reagent shelf lives | bbeitzel | 454 Pyrosequencing | 2 | 10-08-2008 07:02 AM |
![]() |
|
Thread Tools |
![]() |
#1 |
Member
Location: seoul Korea Join Date: Mar 2011
Posts: 19
|
![]()
Hello
I observed a strange sequence result for two times. The run seemed OK until a certain point, and sequnce quality collapse and the low sequence quality retained until the end of the run. Two Times !!! I found that the sequnce quality drop down right after resetting new primer. and I found that there was a reset process right before applying new primer from the log file. I checked the reset reagent, and the reset reagent bottle was almost empty!!! I am not sure that the low sequence quality mainly from the shortage of the reset reagent, but I am kind of positive. Anyone has an experience of running out of the reset reagent? Any idea about the low sequence quality observed from the first cycle of new primer? |
![]() |
![]() |
![]() |
#2 |
Member
Location: USA Join Date: Aug 2010
Posts: 23
|
![]()
Since there is no cleave step after last cycle of each primer, if you run out of reset buffer, any further ligation won'thappen and most likely the instrument will just repeat to scan the last cycle of the primer before reset. What you can do is to get a new bottle of reset buffer and rerun the primer with low quality as far as you haven't clear the run and slide is still sitting on the instrument.
|
![]() |
![]() |
![]() |
#3 | |
Senior Member
Location: East Coast USA Join Date: Feb 2008
Posts: 7,143
|
![]() Quote:
Last edited by GenoMax; 07-07-2011 at 10:21 AM. Reason: Wrong forum post |
|
![]() |
![]() |
![]() |
#4 |
Guest
Posts: n/a
|
![]()
There is a list of required volumes for each buffer in the appendix of the instrument manual. You sometimes have to add more reset depending on the type of run.
|
![]() |
![]() |
#5 |
Member
Location: Maryland Join Date: Jun 2011
Posts: 16
|
![]()
Agreed. I have noticed also that the reset buffer get sucked up a little faster than the manual states on my solid 4. However I think it is always a good idea to pause at a safe pause point and check your buffer levels if you think any of your reagents will come close to running out based on your calculations at the beginning of your run.
|
![]() |
![]() |
![]() |
#6 |
Junior Member
Location: Philly Join Date: Oct 2010
Posts: 4
|
![]()
As a general rule we change the bottle of reset buffer when we change the tags. We have a huge surplus of that particular buffer so "when in doubt, change it out".
|
![]() |
![]() |
![]() |
Thread Tools | |
|
|