Default problem with the ligation of adaptator- Rapid library

Dear All,
In the aim of a prokaryotic RNA sequencing on Roche 454; I used the kit Rapid cDNA Library (GS FLX Titanium).
To recover the small RNA, I modified slightly the protocol at the level of purification (calibration of beads Ampur) between the enzymatic steps (RT, Fragment repair, ..)
Unfortunately, it seems that the step of the ligation of double-stranded adapters doesn't work very well because the controls checked using Bioanalyser of Agilent (high sensitivity) show the presence of my product, but no matches were exploitable using the TBS fluorometer and at the stage of titration.
Has anyone already meet this kind of problem on the kit Rapid library?
thanks for your advices.
cheers.
doc_lille is offline   Reply With Quote
Old 07-27-2010, 08:16 AM   #2
pmiguel
Senior Member
 
Location: Purdue University, West Lafayette, Indiana

Join Date: Aug 2008
Posts: 2,317
Default

Can you post your Bioanalyzer trace?
pmiguel is offline   Reply With Quote
Old 07-27-2010, 11:21 PM   #3
doc_lille
Junior Member
 
Location: france

Join Date: Mar 2009
Posts: 5
Default with the bioanalyser profil

To be a little more precise, my starting material is composed exclusively of messenger RNA.
In order to detect if I effectively linked the adapters on the cDNA double stranded, I made a conventional PCR (35 cycles) with primers A / B (sequences ligated adapters) but my PCR still always negative even after an developpement concerning a gradient of the concentration in MgCl2 and gradient of temperatures.
thank you for hypothetical explanations
Attached Images
File Type: jpg bioanalyser.jpg (16.3 KB, 41 views)
doc_lille is offline   Reply With Quote
Old 07-28-2010, 04:41 AM   #4
pmiguel
Senior Member
 
Location: Purdue University, West Lafayette, Indiana

Join Date: Aug 2008
Posts: 2,317
Default

Did you do a positive control to make sure the primers you are using will work?

If so, then it is likely that your adaptors failed to ligate to the double stranded cDNA. Possible explanations for this would include:

(1) Contaminants from clean-up steps/washes (eg, chaotropes or alcohols) prevented one or more of the blunting, a-tailing or ligation reactions.
(2) First or second strand cDNA synthesis failed.
(3) exonucleases destroyed your adaptors during the ligation step.

The above is little more than my writing "either you QC assay has failed, or, indeed, you failed to produce amplicons".

--
Phillip
pmiguel is offline   Reply With Quote
Old 07-28-2010, 07:04 AM   #5
doc_lille
Junior Member
 
Location: france

Join Date: Mar 2009
Posts: 5
Default

Dear phillip,
thank you for your quick and accurate response.
In fact I tested the primers A and B on a previous library prepared from a kit GS FLX Titanium (not quick Libary preparation) and saw a good smear on the agarose gel after the PCR.
In terms of hypothesis No. 2, I do not think that it was a trouble with the RT or the double strand synthesis because the Bioanalyser (DNA Chip) posts a explanable profile, it is therefore clearly an dsDNA. Moreover, With the same RNA, i tested another kit from Ambion which included a RT and double strand synthesis, then after purification with Ampure, i came back into the Roche kit at the Fragment repair step until the end.
After all, i observed the same profil using the Bioanalyser and my PCR control A and B , still not working.
I keep in memory the prospective problem with the contaminant about the wash step, i will test this hypothesis using the MinElute kit (Qiagen ).
Keep in touch.
thanks
doc_lille is offline   Reply With Quote
Old 07-28-2010, 08:14 AM   #6
pmiguel
Senior Member
 
Location: Purdue University, West Lafayette, Indiana

Join Date: Aug 2008
Posts: 2,317
Default

The most common issue I see with MinElute is ethanol contamination. This is easy to fix using a speed vap. If you nose is sensitive to ethanol, you can smell it.

Second most common, I think, are Acetate salts, I think. You can actually see these, if they are really high, in the UV spectrum as a peak in the 230 nm region.

--
Phillip
pmiguel is offline   Reply With Quote
Reply

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 06:34 PM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2020, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO