  • HiSeq vs MiSeq read-length and quality

    Hi,
    I have a very basic question.
    Why is the read highest achievable read-length of MiSeq compared to HiSeq longer although, as far as I understand, they use the same chemistry?

    Second question would be, why is it usual that the second read of a paired end read has lower quality than the first one?

    I am a novice in the NGS field. I searched the forum but couldn't find an answer.
    Are there any comprehensive introduction to illumina sequencing technology?
    The manuals on the illumina webpage are nice, but only give a very rough overview.

    Thanks a lot.

    kind regards

  • #2
    MiSeq - 2 x 300 bp (or 600 bp single-end, theoretically possible though you probably don't want that).
    HiSeq 2500 - Max supported read length is 2 x 250 bp in Rapid-mode.

    R2 generally has lower quality as:
    a) cluster grow fatter over time and the sequencer may have trouble basecalling if a run is overloaded
    b) reagents are held at 4-6 C in sequencer but deteriorate some over length of the run

    Comment


    • #3
      Thanks a lot, very nice explanations!

      Comment


      • #4
        My understanding that the read lengths are limited due to DNA damage.

        Firstly, incorporation chemistry/removal of the dye terminator is somewhat corrosive to the DNA - so this restricts the single end read length to roughly 300 bases, and requires the paired end turn around to regenerate the template.
        Secondly, the HiSeq uses fairly powerful lasers which also damages the DNA. The MiSeq on the other hand has LEDs, which are not as damaging.

        Comment


        • #5
          Originally posted by bbp View Post
          My understanding that the read lengths are limited due to DNA damage.

          Firstly, incorporation chemistry/removal of the dye terminator is somewhat corrosive to the DNA - so this restricts the single end read length to roughly 300 bases, and requires the paired end turn around to regenerate the template.
          I doubt that this is the case. The limitation is more likely phasing/prephasing -- which has to do with the ensemble of product molecule containing members slightly behind or slightly ahead of the bulk of the molecules.
          If the template molecules were actually damaged, then turn around would not be able to generate reverse-complement strands for sequencing the reverse read(s).
          Originally posted by bbp View Post
          Secondly, the HiSeq uses fairly powerful lasers which also damages the DNA. The MiSeq on the other hand has LEDs, which are not as damaging.
          Nope. The HiSeq 2500 in Rapid mode can do 2x250 base reads at roughly the same quality as the MiSeq does 2x250 base reads. So, whatever the issues are limiting the read lengths, it won't be the laser.
          --
          Phillip

          Comment


          • #6
            Hmm. That was the answer given to me by Illumina. It's possible that they were not providing the correct information, but I'm inclined to believe it.

            I see no increase in my phasing/prephasing rates on a HiSeq 2500 1x 200 cycle rapid run, and I'd have expected this to be cumulative. Perhaps someone with a 2x 250 rapid run could report their phasing/prephasing stats over the course of a run.

            With regards to DNA damage, I'd suspect that the max cycle length has been optimised to ensure that information is not lost beyond some acceptable standard. If illumina could acceptably perform longer cycle reads, they most certainly would do it.

            Comment


            • #7
              Wait bbp. You are sure this is really Illumina who gave you that answer? Because Illumina usually sounds just as portrayed by GenoMax above.
              And if it isn't Illumina, then who is it?
              I didn't mean there was a difference in phasing/prephasing between the HiSeq and the MiSeq. That isn't really an issue because they both produce basically the same read lengths max. But phasing/prephasing is ultimately the main limiter of Illumina read lengths.
              Also, the marginal improvement in useful results when moving from 150 base to 250 base reads is small for most users. The usefulness of the 250 base reads tends to be in edge cases. So the desire to increase read lengths beyond 150 base on Illumina instruments will be low.
              --
              Phillip

              Comment


              • #8
                In my experience very few people at Illumina have an in depth knowledge of the sequencing chemistry.
                I had also wondered about this question; why was the Miseq (especially in the good old days when the V3 chemistry came out) able to deliver high quality longer reads?
                It seems to me that the HiSeq and MiSeq chemistries are optimized differetnly; the fluorescence levels tend to down with increasing cycle numbers for HiSeq runs and up for MiSeq runs.

                Comment

                Latest Articles

                Collapse

                • seqadmin
                  Current Approaches to Protein Sequencing
                  by seqadmin


                  Proteins are often described as the workhorses of the cell, and identifying their sequences is key to understanding their role in biological processes and disease. Currently, the most common technique used to determine protein sequences is mass spectrometry. While still a valuable tool, mass spectrometry faces several limitations and requires a highly experienced scientist familiar with the equipment to operate it. Additionally, other proteomic methods, like affinity assays, are constrained...
                  04-04-2024, 04:25 PM
                • seqadmin
                  Strategies for Sequencing Challenging Samples
                  by seqadmin


                  Despite advancements in sequencing platforms and related sample preparation technologies, certain sample types continue to present significant challenges that can compromise sequencing results. Pedro Echave, Senior Manager of the Global Business Segment at Revvity, explained that the success of a sequencing experiment ultimately depends on the amount and integrity of the nucleic acid template (RNA or DNA) obtained from a sample. “The better the quality of the nucleic acid isolated...
                  03-22-2024, 06:39 AM

                ad_right_rmr

                Collapse

                News

                Collapse

                Topics Statistics Last Post
                Started by seqadmin, 04-11-2024, 12:08 PM
                0 responses
                17 views
                0 likes
                Last Post seqadmin  
                Started by seqadmin, 04-10-2024, 10:19 PM
                0 responses
                22 views
                0 likes
                Last Post seqadmin  
                Started by seqadmin, 04-10-2024, 09:21 AM
                0 responses
                16 views
                0 likes
                Last Post seqadmin  
                Started by seqadmin, 04-04-2024, 09:00 AM
                0 responses
                46 views
                0 likes
                Last Post seqadmin  
                Working...
                X