  • #16
    Ok sorry for the delay all...please see attached document for my results/experience. There is a review of standard truseq, the design that I came up with with help from this thread, and an example samplesheet.

    Everything comes with my standard disclaimer of "if you don't know what you're doing you WILL mess something up" and...this is based on one successful run of testing...so this is very experimental.
    Attached Files

    Comment


    • #17
      Originally posted by pmiguel View Post
      So that would be a hemi-methylated DpnI site? With the flowcell oligo contributing the methylated adenine required for DpnI to cut?

      --
      Phillip
      Hi Philip,
      that was the idea. But ECO pointed out that it is probably uracil at some position and some uracil dependant cleavage.

      Comment


      • #18
        I could definitely be wrong though!! It seems crazy to rely on restriction when more specificity would be achieved by uracil dependent cleavage.

        The flow cell primer having a uracil at that position is a bit of a clue too...assuming that info is correct.

        Comment


        • #19
          Hi all, I'm new to NGS so approach this post with a healthy dose of skepticism.

          I'm in the process of designing PCR primers for a targeted approach to sequencing dual indexed amplicons on the MiSeq. After reading ECO's posts, this is what I've deduced about primer design:

          FORWARD PCR PRIMER:
          5’ – AATGATACGGCGACCACCGAGA{TCTACAC}nnnnnnnnACACTCTTTCCCTACACGACGCTCTTCCGATCT[Forward primer sequence]
          REVERSE PCR PRIMER:
          5’ – CAAGCAGAAGACGGCATACGAGATnnnnnnnnGTGACTGGAGTTCAGACGTGTGCTCTTCCGATCT[Reverse primer sequence]

          ...where bold = i5 and i7 indices, respectively.

          This confuses me because the sequences following the indices are different from those introduced by the transposon. According to Illumina's Customer Sequence Letter, a targeted PCR strategy that mimics samples prepped using Nextera would have primers that appear as follows:

          FORWARD PCR PRIMER:
          5' - AATGATACGGCGACCACCGAGA{TCTACAC}nnnnnnnnTCGTCGGCAGCGTCAGATGTGTATAAGAGACAG[Forward primer sequence]
          REVERSE PCR PRIMER:
          5’ – CAAGCAGAAGACGGCATACGAGATnnnnnnnnGTCTCGTGGGCTCGGAGATGTGTATAAGAGACAG[Reverse primer sequence]

          ...which would mean Read 1, Index 1, and Read 2 sequencing primers are different, hence my confusion. Will either of these strategies work on the MiSeq? Has the above strategy worked for some folk? My thanks in advance.

          Comment


          • #20
            I think both should work.
            The Illumina read primer is actually a primer mix of several primers. I guess there's probably primers for both sequences in there (TruSeq and Nextera).

            However, if you're generating library directly from PCR, why not remove those sequences all together and just use custom read primers. Your custom read primers will just be versions of your fwd and rev primer sequences. You'll also waste less reagent by not sequencing your primers. See my post further up.

            Comment


            • #21
              Originally posted by TonyBrooks View Post
              I think both should work.
              The Illumina read primer is actually a primer mix of several primers. I guess there's probably primers for both sequences in there (TruSeq and Nextera).

              However, if you're generating library directly from PCR, why not remove those sequences all together and just use custom read primers. Your custom read primers will just be versions of your fwd and rev primer sequences. You'll also waste less reagent by not sequencing your primers. See my post further up.
              Got it. Thanks!

              Comment


              • #22
                Originally posted by TonyBrooks View Post
                I think both should work.
                The Illumina read primer is actually a primer mix of several primers. I guess there's probably primers for both sequences in there (TruSeq and Nextera).

                However, if you're generating library directly from PCR, why not remove those sequences all together and just use custom read primers. Your custom read primers will just be versions of your fwd and rev primer sequences. You'll also waste less reagent by not sequencing your primers. See my post further up.
                We've been using custom read primers on an amplicon library successfully on the HiSeq. However, we just tried the exact same strategy on the MiSeq and it failed utterly (nothing but PhiX in the first (index) read.)

                Are the annealing temperatures very different on the HiSeq and MiSeq?

                Comment


                • #23
                  See this document (taken from an Illumina bulletin on their website).
                  Attached Files

                  Comment


                  • #24
                    MiSeq fail with Nextera adapters

                    I used the following adapters on our MiSeq:

                    N701:
                    CAAGCAGAAGACGGCATACGAGATTCGCCTTAGTCTCGTGGGCTCGGAGATGTGTATAAGAGACAG

                    N501:
                    AATGATACGGCGACCACCGAGATCTACACTAGATCGCTCGTCGGCAGCGTCAGATGTGTATAAGAGACAG

                    It didn't detect clusters (I used the provided sequencing primers).

                    Anyone have any ideas of what went wrong? I'm dumbfounded.

                    Comment


                    • #25
                      Originally posted by uberfinch View Post
                      I used the following adapters on our MiSeq:

                      N701:
                      CAAGCAGAAGACGGCATACGAGATTCGCCTTAGTCTCGTGGGCTCGGAGATGTGTATAAGAGACAG

                      N501:
                      AATGATACGGCGACCACCGAGATCTACACTAGATCGCTCGTCGGCAGCGTCAGATGTGTATAAGAGACAG

                      It didn't detect clusters (I used the provided sequencing primers).

                      Anyone have any ideas of what went wrong? I'm dumbfounded.
                      I guess I don't understand your strategy.
                      Looks to me like clusters should have formed, but I don't see anywhere for the standard sequencing primers to anneal. Which would make the clusters undetectable unless you added custom sequencing primer. If you did add custom sequencing primer, does it have an annealing temp of 65 oC (or higher)? The Illumina document that Tony attached above points out that primers with lower annealing temps might not work.

                      Eg, your i5 adapter:

                      AATGATACGGCGACCACCGAGATCTACA | CTAGATCG | CTCGTCGGCAGCGTCAGATGTGTATAAGAGACAG
                      Flow cell binding----------- | index??? | ?____________________what is this?


                      Oh wait, that is the old Nextera transposase sequence. Guess Illumina does not include it in their standard i5 primer cocktail. You would have to spike it in.

                      --
                      Phillip
                      Last edited by pmiguel; 07-27-2012, 03:55 AM.

                      Comment


                      • #26
                        Thanks for your reply Philip. Yes, as you surmised those 19 bp on the end are the nextera sequence taken from the sequence reported in Illumina's letter for version 2. To make absolutely sure it was right I TOPO cloned a library made with the Nextera XT kit that is supposedly compatible with the miseq on got that exact adapter sequence.

                        Is there any reason to believe the PCR primers for nextera are somehow modified?

                        Comment


                        • #27
                          PCR primers? No.
                          It is the internal segment, closest to the "insert" that has been replaced in the Illumina Nextera kits.
                          Details in the "Illumina Customer Sequence Letter".
                          Again, I think you probably did get clusters on your flow cell. They just were invisible because Illumina sequencers only detect clusters to which a read1 sequencing primers can anneal.

                          --
                          Phillip

                          Comment


                          • #28
                            Originally posted by uberfinch View Post
                            To make absolutely sure it was right I TOPO cloned a library made with the Nextera XT kit that is supposedly compatible with the miseq on got that exact adapter sequence.
                            Are you sure it was a Nextera XT kit, not an old (possibly Epicentre) Nextera kit that you topo cloned and Sanger sequenced?

                            --
                            Phillip

                            Comment


                            • #29
                              Yes, I'm positive it is a Nextera XT kit (just bought it!). If you look on the illumina letter, those sequences are from the Illumina V.2 nextera kits.

                              Anyway I guess there isn't a good explanation for why it didn't work. Maybe I'll just have to risk burning another kit! There's only so many times you can stare at the sequences looking for an error.

                              Comment


                              • #30
                                Okay, my mistake. Your adapters should be fine.
                                Is this just a normal NexteraXT kit library, or a set of amplicons you constructed in another manner to mimic the structure of a Nextera library?

                                We just did a MiSeq run on a NexteraXT library (our first Nextera run). It worked. I even spiked in 30% phiX, just in case. (Denature the normal, NaOH way, diluted to 8pM and mixed 30:70 vol:vol with the heat denatured NexteraXT library.)

                                Actually, that was our second Nextera run. We ran the same library (no phiX) the previous day on another MiSeq and it terminated prematurely (crashed) without imaging apparently. Maybe I should have mentioned this earlier? But I was pretty sure this was a fluidics issue because the instrument failed a subsequent volume test.

                                --
                                Phillip

                                Comment

                                Latest Articles

                                Collapse

                                • seqadmin
                                  Techniques and Challenges in Conservation Genomics
                                  by seqadmin



                                  The field of conservation genomics centers on applying genomics technologies in support of conservation efforts and the preservation of biodiversity. This article features interviews with two researchers who showcase their innovative work and highlight the current state and future of conservation genomics.

                                  Avian Conservation
                                  Matthew DeSaix, a recent doctoral graduate from Kristen Ruegg’s lab at The University of Colorado, shared that most of his research...
                                  03-08-2024, 10:41 AM
                                • seqadmin
                                  The Impact of AI in Genomic Medicine
                                  by seqadmin



                                  Artificial intelligence (AI) has evolved from a futuristic vision to a mainstream technology, highlighted by the introduction of tools like OpenAI's ChatGPT and Google's Gemini. In recent years, AI has become increasingly integrated into the field of genomics. This integration has enabled new scientific discoveries while simultaneously raising important ethical questions1. Interviews with two researchers at the center of this intersection provide insightful perspectives into...
                                  02-26-2024, 02:07 PM

                                ad_right_rmr

                                Collapse

                                News

                                Collapse

                                Topics Statistics Last Post
                                Started by seqadmin, 03-14-2024, 06:13 AM
                                0 responses
                                33 views
                                0 likes
                                Last Post seqadmin  
                                Started by seqadmin, 03-08-2024, 08:03 AM
                                0 responses
                                72 views
                                0 likes
                                Last Post seqadmin  
                                Started by seqadmin, 03-07-2024, 08:13 AM
                                0 responses
                                80 views
                                0 likes
                                Last Post seqadmin  
                                Started by seqadmin, 03-06-2024, 09:51 AM
                                0 responses
                                68 views
                                0 likes
                                Last Post seqadmin  
                                Working...
                                X