  • Error rate of 51 cycle mRNA Seq

    We have been running mRNAseq on an Illumina GAll. At cycle 36 we are seeing a spike in our error rate. Has anyone else observed this? We do not see it when we run genomic dna or in our standard.
    Attached Files

  • #2
    Bruce,

    I have what may be a solution/explanation of your problem, but I'd strongly recommend that you contact Illumina's tech support team as they can help you work through the issue.

    Our R&D group used to see higher error rates when they first starting running mRNA-Seq samples longer than 36 reads. It turns out they were sequencing libraries with relatively short inserts (<50 bases). They were reading across the entire insert and in to the adapter sequence on the other side (and since the adapter doesn't align to the genome, you get higher error rates). Libraries with consistently larger inserts are needed when sequencing out to 75 or 100 bases (or beyond). Hopefully this will help you with what you're seeing.

    Sincerely,

    Shawn C. Baker
    Market Manager, Expression and Regulation
    Illumina

    Comment

    Latest Articles

    Collapse

    • seqadmin
      Current Approaches to Protein Sequencing
      by seqadmin


      Proteins are often described as the workhorses of the cell, and identifying their sequences is key to understanding their role in biological processes and disease. Currently, the most common technique used to determine protein sequences is mass spectrometry. While still a valuable tool, mass spectrometry faces several limitations and requires a highly experienced scientist familiar with the equipment to operate it. Additionally, other proteomic methods, like affinity assays, are constrained...
      04-04-2024, 04:25 PM
    • seqadmin
      Strategies for Sequencing Challenging Samples
      by seqadmin


      Despite advancements in sequencing platforms and related sample preparation technologies, certain sample types continue to present significant challenges that can compromise sequencing results. Pedro Echave, Senior Manager of the Global Business Segment at Revvity, explained that the success of a sequencing experiment ultimately depends on the amount and integrity of the nucleic acid template (RNA or DNA) obtained from a sample. “The better the quality of the nucleic acid isolated...
      03-22-2024, 06:39 AM

    ad_right_rmr

    Collapse

    News

    Collapse

    Topics Statistics Last Post
    Started by seqadmin, 04-11-2024, 12:08 PM
    0 responses
    18 views
    0 likes
    Last Post seqadmin  
    Started by seqadmin, 04-10-2024, 10:19 PM
    0 responses
    22 views
    0 likes
    Last Post seqadmin  
    Started by seqadmin, 04-10-2024, 09:21 AM
    0 responses
    16 views
    0 likes
    Last Post seqadmin  
    Started by seqadmin, 04-04-2024, 09:00 AM
    0 responses
    47 views
    0 likes
    Last Post seqadmin  
    Working...
    X