Question Agilent TapeStation vs. Qubit: RNA concentration.

SCENARIO:

After RNA extraction (RNA-seq is my aim), I check the concentration of the extracted RNA with Qubit, and then run the extracted RNA on Agilent TapeStation 2200 with HSRNA ScreenTape. The concentrations are listed below and the RINe values from the TapeStation are shown in the picture.

Qubit concentrations vs. TapeStation concentrations:
A1 = 14.9 ng/uL vs. 40.6 ng/uL
B1 = 19.4 ng/uL vs. 60.2 ng/uL
C1 = 24.4 ng/uL vs. 135 ng/uL
D1 = 21.5 ng/uL vs. 110 ng/uL
E1 = 32.3 ng/uL vs. ?

QUESTIONS:

1. Which concentration values should I trust: Qubit or TapeStation?

2. Should I trust the RINe values of the TapeStation, even though the error message says "RNA concentration outside the recommended range for RINe", or should I dilute the sample according to the concentration given by the TapeStation and run them again?

3. I've been told to use samples with approximately the same concentration on the TapeStation because the bad concentration of one sample (either too high or too low) can affect the results of the other samples. Is that true? How can one sample affect the result of another?
evobio is offline   Reply With Quote
Old 07-29-2015, 06:58 AM   #2
pmiguel
Senior Member
 
Location: Purdue University, West Lafayette, Indiana

Join Date: Aug 2008
Posts: 2,300
Default

Quote:
Originally Posted by evobio View Post
SCENARIO:

After RNA extraction (RNA-seq is my aim), I check the concentration of the extracted RNA with Qubit, and then run the extracted RNA on Agilent TapeStation 2200 with HSRNA ScreenTape. The concentrations are listed below and the RINe values from the TapeStation are shown in the picture.

Qubit concentrations vs. TapeStation concentrations:
A1 = 14.9 ng/uL vs. 40.6 ng/uL
B1 = 19.4 ng/uL vs. 60.2 ng/uL
C1 = 24.4 ng/uL vs. 135 ng/uL
D1 = 21.5 ng/uL vs. 110 ng/uL
E1 = 32.3 ng/uL vs. ?

QUESTIONS:

1. Which concentration values should I trust: Qubit or TapeStation?
Probably Qubit. But which dye were you using? As long as you were using the RNA-specific dye, then trust the Qubit.
Quote:
Originally Posted by evobio View Post
2. Should I trust the RINe values of the TapeStation, even though the error message says "RNA concentration outside the recommended range for RINe", or should I dilute the sample according to the concentration given by the TapeStation and run them again?
I don't know. Do you want to for some reason? What will you do with the RIN score?
Quote:
Originally Posted by evobio View Post
3. I've been told to use samples with approximately the same concentration on the TapeStation because the bad concentration of one sample (either too high or too low) can affect the results of the other samples. Is that true? How can one sample affect the result of another?
Seems crazy to me, but there are crazy aspects to the Agilent Bioanalyzer. So maybe there are also goofy issues with the TapeStation as well.

--
Phillip
pmiguel is offline   Reply With Quote
Old 08-03-2015, 04:19 AM   #3
evobio
Junior Member
 
Location: Sweden

Join Date: Jul 2015
Posts: 3
Smile

1. I am using the appropriate Qubit dye (Qubit RNA HS Assay Kit).

2. The extracted RNA will be used in the preparation of RNA-seq libraries, so I'm using the RINe scores from the TapeStation as a proxy of the quality of the RNA. So I guess I need to run the samples again with the new concentrations in order to trust completely the RINe scores of the TapeStation.

3. To me that doesn't make sense, because that means one bad sample can give you bad scores for all other samples (even if they are good). If that really is the case, each run should have only one sample, which is absurd.

Last edited by evobio; 08-03-2015 at 04:25 AM.
evobio is offline   Reply With Quote
Old 08-03-2015, 05:55 AM   #4
pmiguel
Senior Member
 
Location: Purdue University, West Lafayette, Indiana

Join Date: Aug 2008
Posts: 2,300
Default

2. Okay.
3. Actually, Agilent has decent tech support. Maybe give them a call about this issue. See if it is true...

--
Phillip
pmiguel is offline   Reply With Quote
Reply

Tags
agilent 2200, qubit, rna concentration, tapestation

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 10:19 PM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2018, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO