  • MiSeq error: Through-focus peak did not exceed SNR threshold

    Hi,

    Using a 10 pM pooled library of 6 indexed samples (4 from FFPE + 2 from fresh tissues) prepared with TruSeq Exome kit, I got this error message when running on MiSeq with no data generated (except the log files):
    'Through-focus peak did not exceed SNR threshold (Illumina.Instrument.Bolt.Logic.AutoFocus.DarkSampleException)'

    Do you have any idea what it means?
    What could be the cause of this error?

    Thanks !

  • #2
    That probably means that no signal was detected (no clusters generated). Was the pool QC'ed?

    You should contact Illumina tech support and have them take a look at the run. If this is a hardware/reagent failure, they will be able to determine and replace as needed. Include a screenshot of the error on MiSeq with your ticket.

    Comment


    • #3
      Ok.
      I checked the concentration of the final library with PicoGreen.

      How do you QC the final pooled library?
      Any method/kit in particular?

      Comment


      • #4
        We use the qPCR based method for quantification, such as the KAPA kit.

        Comment


        • #5
          To expand on lxkx's answer:

          One of the issues with PicoGreen is that it only looks for dsDNA, which you might have plenty of, but ignores the potential complication that not all DNA fragments will have sequencing adapters in the correct orientation. qPCR amplification only occurs when adapters are in the correct orientation, so it's a more accurate method when quantifying material for sequencing. I've used the KAPA kits for several years now and have had very good luck with them. The standards are remarkably consistent lot-to-lot, too.

          Comment

          Latest Articles

          Collapse

          • seqadmin
            Current Approaches to Protein Sequencing
            by seqadmin


            Proteins are often described as the workhorses of the cell, and identifying their sequences is key to understanding their role in biological processes and disease. Currently, the most common technique used to determine protein sequences is mass spectrometry. While still a valuable tool, mass spectrometry faces several limitations and requires a highly experienced scientist familiar with the equipment to operate it. Additionally, other proteomic methods, like affinity assays, are constrained...
            04-04-2024, 04:25 PM
          • seqadmin
            Strategies for Sequencing Challenging Samples
            by seqadmin


            Despite advancements in sequencing platforms and related sample preparation technologies, certain sample types continue to present significant challenges that can compromise sequencing results. Pedro Echave, Senior Manager of the Global Business Segment at Revvity, explained that the success of a sequencing experiment ultimately depends on the amount and integrity of the nucleic acid template (RNA or DNA) obtained from a sample. “The better the quality of the nucleic acid isolated...
            03-22-2024, 06:39 AM

          ad_right_rmr

          Collapse

          News

          Collapse

          Topics Statistics Last Post
          Started by seqadmin, 04-11-2024, 12:08 PM
          0 responses
          18 views
          0 likes
          Last Post seqadmin  
          Started by seqadmin, 04-10-2024, 10:19 PM
          0 responses
          22 views
          0 likes
          Last Post seqadmin  
          Started by seqadmin, 04-10-2024, 09:21 AM
          0 responses
          16 views
          0 likes
          Last Post seqadmin  
          Started by seqadmin, 04-04-2024, 09:00 AM
          0 responses
          47 views
          0 likes
          Last Post seqadmin  
          Working...
          X