Old 08-28-2017, 01:29 PM   #21
Joanna
Junior Member
 
Location: North Carolina

Join Date: Jul 2017
Posts: 9
Default

Hi guys,

I want to post some updates. As people suggested, we ordered fresh primers and made new libraries. Then loaded 6pM with 30% PhiX spike in. And we also doubled the volume of sequencing primers.

This time we had a cluster density of 854 K/mm2 which has improved but PF% was still very low (48.78%) and yield was also low. Aligned PhiX% was just 15% instead of 30%.

My thought is:
Based on the attached thumbnail images, I feel it was over-clustered this time. And since we sequenced low diversity 16S rRNA gene libraries so we had uneven signal distribution for each cycle. This may cause trouble for the camera to correctly identify clusters. So we saw an overall drop in Q30% and "bleeding" (see attached). But this couldn't explain the low PhiX% than expected. And I also noticed that the Q30% of R2 (the first index read) was especially low. Wondering if others met this before? What do you think is the problem?

Thanks!
Attached Images
File Type: png Q30.png (99.2 KB, 9 views)
Attached Files
File Type: pdf run metrics.pdf (141.9 KB, 12 views)
File Type: pdf thumbnails.pdf (1.05 MB, 4 views)
Joanna is offline   Reply With Quote
Old 08-29-2017, 06:11 AM   #22
MU Core
Member
 
Location: Columbia, Missouri

Join Date: Apr 2008
Posts: 50
Default

Joanna, looks to me as though you've resolved the library issue with the new primers. I would agree that this particular run is over-clustered. We observe similar quality issues when 16S libraries are clustered at this density. Reduce the cluster density and the Q30% will improve. We target cluster density of 650-750 K/mm2 with a 15% spike-in which works well for us on a MiSeq run. I've provided two attachments that display typical results.
Attached Images
File Type: jpg Run metric.JPG (57.0 KB, 5 views)
File Type: jpg Q30plot.JPG (26.1 KB, 7 views)
MU Core is offline   Reply With Quote
Old 08-30-2017, 12:30 PM   #23
thermophile
Senior Member
 
Location: CT

Join Date: Apr 2015
Posts: 201
Default

lower than expected phiX indicates that your quant was off. If you loaded your library to get 30% and only could align 15% phiX, that means that your 16S was approximately twice as concentrated as you thought so took up twice as many spots on the flowcell as you planned. Recovering roughly the amount of phiX as you put on is a decent indication that your library quant was correct
__________________
Microbial ecologist, running a sequencing core. I have lots of strong opinions on how to survey communities, pretty sure some are even correct.
thermophile is offline   Reply With Quote
Reply

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 07:04 AM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2017, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO