Old 02-15-2017, 12:14 PM   #1
soleulloa
Member
 
Location: Chile

Join Date: Aug 2011
Posts: 33
Default Discordance between Cluster Density and Thumbnail images

Hi all,

I ran MiSeq using a 500 cycles kit.
The value for cluster density was 310K/mm2 but if I'm going to look the thumbnail images, I think I have much more cluster than that.

Illumina tech looked the results and says "you are overclusterized".
But for me cluster density is OK when I've reviewed the images.

I'm attaching here 4 images from cycle 1.

Could you tell me your opinion?
Thank you!!
Soledad
Attached Images
File Type: jpg s_1_1102_a.jpg (58.3 KB, 23 views)
File Type: jpg s_1_1102_t.jpg (57.6 KB, 22 views)
File Type: jpg s_1_1102_c.jpg (62.8 KB, 13 views)
File Type: jpg s_1_1102_g.jpg (68.3 KB, 18 views)
soleulloa is offline   Reply With Quote
Old 02-15-2017, 02:41 PM   #2
nucacidhunter
Senior Member
 
Location: Iran

Join Date: Jan 2013
Posts: 1,013
Default

To me reported cluster density is what I would expect from the images. The run has been underclustered for a library that looks high diversity although I am assuming that the base diversity of other cycles are similar to the first cycle.
nucacidhunter is offline   Reply With Quote
Old 02-15-2017, 07:17 PM   #3
Brian Bushnell
Super Moderator
 
Location: Walnut Creek, CA

Join Date: Jan 2014
Posts: 2,637
Default

What's the scale on the little yellow boxes?
Brian Bushnell is offline   Reply With Quote
Old 02-15-2017, 08:34 PM   #4
misterc
Member
 
Location: Livermore, CA

Join Date: Jan 2016
Posts: 20
Default

That looks about right to me for 300-500K/mm2 raw density (and NOT overclustered). Those intensities look reasonable to me too for cycle 1 given the MiSeq ramps up exposure on the cameras at each cycle to counter the typical decreasing intensity plots on their instruments.
misterc is offline   Reply With Quote
Old 02-16-2017, 03:02 AM   #5
soleulloa
Member
 
Location: Chile

Join Date: Aug 2011
Posts: 33
Default

Ok! Thank you.
I'm using 10 pM (from 4 nM) to get this result.
What do you think I have to use now? 12 or 15 pM to get a good run?
soleulloa is offline   Reply With Quote
Old 02-16-2017, 05:11 AM   #6
microgirl123
Senior Member
 
Location: New England

Join Date: Jun 2012
Posts: 188
Default

What's your percentage passing filter?
microgirl123 is offline   Reply With Quote
Old 02-16-2017, 05:15 AM   #7
soleulloa
Member
 
Location: Chile

Join Date: Aug 2011
Posts: 33
Default

Hi,

95.27% PF
soleulloa is offline   Reply With Quote
Old 02-16-2017, 05:17 AM   #8
microgirl123
Senior Member
 
Location: New England

Join Date: Jun 2012
Posts: 188
Default

If 95% of them are passing filter, then you are definitely not overclustered. If you were overclustered your percentage passing filter would be low.

Unfortunately, the relationship between loading concentration and clusters is not linear so it's always a guess. I find that getting good cluster density is the Achilles' heel of the Illumina Miseq. My guess would be to go with the 15 pM since your density was so low.
microgirl123 is offline   Reply With Quote
Old 02-17-2017, 07:21 AM   #9
pmiguel
Senior Member
 
Location: Purdue University, West Lafayette, Indiana

Join Date: Aug 2008
Posts: 2,212
Default

Quote:
Originally Posted by Brian Bushnell View Post
What's the scale on the little yellow boxes?
I don't know, but when I looked into it in the past, it seemed that the size of a cluster was around 1um in diameter. But that may have been on a HiSeq. I'm not sure whether the scales would be different on a MiSeq.

Old-timers like @ECO would be more likely to know. (Alas, I don't think @ callouts actually work on SeqAnswers...)

--
Phillip
pmiguel is offline   Reply With Quote
Old 02-17-2017, 07:46 AM   #10
GenoMax
Senior Member
 
Location: East Coast USA

Join Date: Feb 2008
Posts: 6,409
Default

Illumina has a note on optimizing cluster density available here.

@soledad: Have you checked what the images look like in the middle and at the end of run?

Last edited by GenoMax; 02-17-2017 at 07:51 AM.
GenoMax is offline   Reply With Quote
Old 02-17-2017, 08:27 AM   #11
soleulloa
Member
 
Location: Chile

Join Date: Aug 2011
Posts: 33
Default

Someone has quantified libraries before dilution? When you prepare 2 nM dilutions?
I've quantified before and after that final dilution now because this is the first run when I have this problem and I'd like to compare these results.
Always I was between 1000-1200 K/mm2 using a 300 cycles kit.
This time I did the same process... and the only variable is I'm using 500 cycles kit.
But the library concentration will be the same.. so I don't understand what happens.

soleulloa is offline   Reply With Quote
Old 02-17-2017, 02:13 PM   #12
nucacidhunter
Senior Member
 
Location: Iran

Join Date: Jan 2013
Posts: 1,013
Default

It is most likely overestimating library concentration if you have used the same type of library. qPCR is the best method for quantification, other methods that quantify dsDNA are inadequate for libraries that contain non-sequenceable dsDNA fragments.
nucacidhunter is offline   Reply With Quote
Reply

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 03:46 PM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2017, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO