Old 10-19-2016, 11:27 AM   #1
pmiguel
Senior Member
 
Location: Purdue University, West Lafayette, Indiana

Join Date: Aug 2008
Posts: 2,317
Exclamation Agilent Bioanalyzer Negative Peaks?!?

Here is what I mean:

NegativePeak.png

This is some input DNA obtained by reverse-crosslinking sonicated chromatin submitted by someone interested in our doing library construction for them. It was run on an Agilent Bioanalyzer High Sensitivity Chip.

Everything looks great except -- the peak produced has negative amplitude!
The Agilent Bioanalyzer Trouble shooting manual suggests this could happen as a result of the sample containing "detergents" or "dyes".

Anyone seen this and been able to figure out the cause?

(Yes, I know I answered the post of someone asking nearly the same question 4 years ago. But having seen this a few time recently in our lab, I'm finding all the responses in that thread, including my own, unhelpful.)

--
Phillip

Last edited by pmiguel; 10-19-2016 at 11:39 AM.
pmiguel is offline   Reply With Quote
Old 10-19-2016, 12:02 PM   #2
luc
Senior Member
 
Location: US

Join Date: Dec 2010
Posts: 451
Default

We have seen this before (slightly less dramatic) also with ChIP DNA from customers. Perhaps there is a reagent in the Diagenode ChIP kits that can have this effect? The library prep and sequencing went well nevertheless.
luc is offline   Reply With Quote
Old 10-19-2016, 01:49 PM   #3
nucacidhunter
Jafar Jabbari
 
Location: Melbourne

Join Date: Jan 2013
Posts: 1,238
Default

I have seen it when DNA is eluted with buffer containing 0.1% Tween 20 even though in the Chip it is further diluted. Running the same sample on Screentape HS5000 is OK.
nucacidhunter is offline   Reply With Quote
Old 10-19-2016, 01:55 PM   #4
RickC7
Member
 
Location: Baton Rouge, Louisiana

Join Date: Feb 2010
Posts: 31
Default

Did you check concentration on qubit? May be outside of upper range? Have seen similar, not so pronounced, but diluting helped in our case.
RickC7 is offline   Reply With Quote
Old 10-20-2016, 12:08 PM   #5
pmiguel
Senior Member
 
Location: Purdue University, West Lafayette, Indiana

Join Date: Aug 2008
Posts: 2,317
Default

Quote:
Originally Posted by RickC7 View Post
Did you check concentration on qubit? May be outside of upper range? Have seen similar, not so pronounced, but diluting helped in our case.
We re-ran the samples, loading 1/4th the amount. Same effect:
NegativePeak2.png

--
Phillip
pmiguel is offline   Reply With Quote
Old 10-20-2016, 12:13 PM   #6
pmiguel
Senior Member
 
Location: Purdue University, West Lafayette, Indiana

Join Date: Aug 2008
Posts: 2,317
Default

Quote:
Originally Posted by nucacidhunter View Post
I have seen it when DNA is eluted with buffer containing 0.1% Tween 20 even though in the Chip it is further diluted. Running the same sample on Screentape HS5000 is OK.
Interesting!

Thanks,
--
Phillip
pmiguel is offline   Reply With Quote
Old 10-20-2016, 12:22 PM   #7
pmiguel
Senior Member
 
Location: Purdue University, West Lafayette, Indiana

Join Date: Aug 2008
Posts: 2,317
Default

Quote:
Originally Posted by nucacidhunter View Post
I have seen it when DNA is eluted with buffer containing 0.1% Tween 20 even though in the Chip it is further diluted. Running the same sample on Screentape HS5000 is OK.
Alas, after checking with the people doing the sample prep, they don't use Tween. They use SDS. They also do a minelute clean-up afterwards.

--
Phillip
pmiguel is offline   Reply With Quote
Old 10-20-2016, 05:31 PM   #8
seqdavis
Junior Member
 
Location: Midwest

Join Date: Sep 2015
Posts: 5
Default

Phillip,

Have you checked this resource from Agilent - slide 12??

http://www.agilent.com/cs/library/sl...ng_Feb2014.pdf

Maybe the sample needs one more clean up??
__________________
SeqDavis
seqdavis is offline   Reply With Quote
Old 10-20-2016, 05:35 PM   #9
nucacidhunter
Jafar Jabbari
 
Location: Melbourne

Join Date: Jan 2013
Posts: 1,238
Default

This suggests multiple causes including detergents and possibly chaotropic salts remains from column purification. Probably columns need to be washed more times than manufacturer recommends.
nucacidhunter is offline   Reply With Quote
Reply

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 12:07 AM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2020, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO