Exclamation Miseq Flow cell leakage Problems

I have been having problems with leakage from one of my Miseq flow cells. There is pitting on the heater/slide holder, from salt/fluid build up it seems and resulting erosion - I get contradictory responses from Illumina, some tech people say it is a know problem, others say not!

I have another system and that does not show the same.

Anyone else experienced this problem?

Thanks
dannyhi321 is offline   Reply With Quote
Old 03-07-2016, 09:58 AM   #2
cmbetts
Senior Member
 
Location: Bay Area

Join Date: Jun 2012
Posts: 114
Default

Quote:
Originally Posted by dannyhi321 View Post
I have been having problems with leakage from one of my Miseq flow cells. There is pitting on the heater/slide holder, from salt/fluid build up it seems and resulting erosion - I get contradictory responses from Illumina, some tech people say it is a know problem, others say not!

I have another system and that does not show the same.

Anyone else experienced this problem?

Thanks
The MiSeq at my former company had issues where there would be a poor seal with the flowcell gaskets, resulting in an accumulation of salts where the flowcell seats on the instrument. Occasionally wiping it down with some dI water when salts became visible seemed sufficient to prevent issues.

More recently, when we had a MiSeq installed at my current company, the FAS specifically mentioned cleaning any salt buildup there. I think she recommended using ~70% ethanol, but my memory is fuzzy and we haven't had to deal with it on this instrument that I'm aware of.
cmbetts is offline   Reply With Quote
Old 03-08-2016, 02:19 AM   #3
dannyhi321
Member
 
Location: uk

Join Date: Jul 2014
Posts: 51
Default

Quote:
Originally Posted by cmbetts View Post
The MiSeq at my former company had issues where there would be a poor seal with the flowcell gaskets, resulting in an accumulation of salts where the flowcell seats on the instrument. Occasionally wiping it down with some dI water when salts became visible seemed sufficient to prevent issues.

More recently, when we had a MiSeq installed at my current company, the FAS specifically mentioned cleaning any salt buildup there. I think she recommended using ~70% ethanol, but my memory is fuzzy and we haven't had to deal with it on this instrument that I'm aware of.
Thanks for that - why didn't they just fix the leak?
dannyhi321 is offline   Reply With Quote
Reply

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 10:01 PM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2020, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO