Default UPS (Uninterruptible Power Supply) with Ion Torrent

Hi all!
Does anyone of you use an UPS for preventing power failures ? If yes...which one do you use ?
I'm preparing the lab for the installation of the Ion Torrent PGM and Torrent Server and I have a lot of doubts about the required equipment....thanks!!
Cristian
Liam_Gallagher is offline   Reply With Quote
Old 10-20-2011, 08:14 AM   #2
pmiguel
Senior Member
 
Location: Purdue University, West Lafayette, Indiana

Join Date: Aug 2008
Posts: 2,309
Default

If possible contact the sales rep and ask for a pre-installation guide. Applied Biosystems always has these available for their instruments. They will probably have a recommended UPS.

You might want to use a different one, but at least you have some specifications to start with.
--
Phillip
pmiguel is offline   Reply With Quote
Old 10-21-2011, 12:05 AM   #3
Liam_Gallagher
Member
 
Location: Europe

Join Date: Oct 2011
Posts: 18
Default

Quote:
Originally Posted by pmiguel View Post
If possible contact the sales rep and ask for a pre-installation guide. Applied Biosystems always has these available for their instruments. They will probably have a recommended UPS.

You might want to use a different one, but at least you have some specifications to start with.
--
Phillip
Thank you Phillip!
Yes, we alredy have the "Site Preparation Guide" form Life Technologies...but in this user manual it is only specified that an UPS (from Multiple Suppliers) is required and should be compatible with 1500W output or greater for 1 Torrent Server and 1 Torrent PGM Sequencer.
So...I am just wondering wich one (brand or model) you are using with Ion Torrent.
Thanks,

Cristian
Liam_Gallagher is offline   Reply With Quote
Old 10-21-2011, 07:13 AM   #4
Apollo704
Junior Member
 
Location: Boston

Join Date: Jul 2009
Posts: 7
Default

We are not using a UPS with our sequencers. The run time being as short as it is, we haven't been as concerned with losing power during a run as we would be with a Hi-seq.
Apollo704 is offline   Reply With Quote
Old 11-06-2011, 03:25 PM   #5
rbridenbecker
Junior Member
 
Location: Life Technologies: Carlsbad, CA

Join Date: Nov 2011
Posts: 5
Default

A UPS also provides protection from power surges and brownouts that can kill electronics so there is value beyond just battery backup.

The Torrent Server has an optional SW module to shutdown the server if the power goes out, so I'd make sure the Torrent Server UPS is compatible. There is some info in the User Manual. --rob
rbridenbecker is offline   Reply With Quote
Old 11-09-2011, 08:06 AM   #6
ptran
Junior Member
 
Location: Houston, TX

Join Date: Jun 2011
Posts: 3
Default

We are using an APC SMT2200

Torrent Server (Optiplex) 115V*10A =~1150VA
Ion Torrent PGM 110V *9A = ~990 VA

total ~ 2140 VA
ptran is offline   Reply With Quote
Old 11-10-2011, 12:47 AM   #7
Liam_Gallagher
Member
 
Location: Europe

Join Date: Oct 2011
Posts: 18
Default

Thank you all!
We just bought the same one suggested by ptran.
It also should be compatible with the SW module that shutdown the server if the power goes out.

Cristian
Liam_Gallagher is offline   Reply With Quote
Old 01-02-2012, 11:16 PM   #8
lek2k
Member
 
Location: Australia

Join Date: Aug 2011
Posts: 32
Default

We have ours connected to the UPS/Backup power point supplied by the Hospital we are based in. Unfortunately there is a split second delay before it kicks in which causes the Torrent Server to reboot or something. To compound matters they test this backup once a month! Anyways at least it's not a dirty shut down which takes fdsk (or whatever it's called) several hours to check the drives.
lek2k is offline   Reply With Quote
Old 01-03-2012, 09:50 AM   #9
HMorrison
Senior Member
 
Location: Massachusetts

Join Date: May 2009
Posts: 116
Default

Ours is from Franek Technologies--we have an emergency generator, so this is basically to protect it during the transition.

FT1-IONPGM-IT Certified Laboratory Power Protection System for an Ion Torrent PGM
Sequencer and WorkStation (computer & monitor)

Pluggable Unit - Input & Output: Rated and configured for the above listed
instrument.
Input, 120 VAC: Provided with a NEMA 5-15P plug on a 6' power cord.
Output, Correct voltage, power and receptacles to match the Ion Torrent
PGM Sequencer and WorkStation (computer & monitor)
Maximum Envelope dimensions in: 14.7"H x 7.6" W x 18.9" D
Maximum installed weight 69 lb.
Backup time has been optimized for laboratories that are equipped with an
emergency generator system with an Automatic Transfer Switch (ATS).
Typical Backup time: Approximately 6 minutes minimum at full
instrumentation load.
HMorrison is offline   Reply With Quote
Reply

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 03:51 PM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2019, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO