Seqanswers Leaderboard Ad

Collapse

Announcement

Collapse
No announcement yet.
X
 
  • Filter
  • Time
  • Show
Clear All
new posts

  • Preparation of low concentration libraries for hybridization

    Hi,
    I know this has been covered by many threads (I've read them all) but am looking for a few clarifications.

    (1) I understand Illumina recommendations are such that you dont want to go past max 0.05 N NaOH during the DNA denaturization steps. But the threads I have seen here all had a 0.1N final NaOH concentraion during denaturing. Is this intentional/a mistake/a change in Illumina recommendations?

    (2) Do you prepare libraries with the hyb buffer right before hyb to cBot (we are preparing it in an institute separate from the genome facility running the samples) or can they be saved for a few days at -20C (similar to Illumina recommendations for phiX libraries?) Anyone have experience here?

    (3) Does the concentration at which the libraries are when they undergo denaturization matter? (One thread- http://seqanswers.com/forums/showthread.php?t=4435 - commented what the concentration was during denaturing was and I wasnt sure whether it was important)

    (4) would be super-happy if someone could give my protocol a quick run-through before I actually go and do it.... especially happy if someone can look at the different calculations in part 1 and see if they all work out right and if any is preferable over the other.

    1. Combine the following in a microcentrifuge tube:
    • template DNA (XXXXX μl)
    • 1 N NaOH (1 μl)
    • Tris-Cl 10 mM, pH 8.5 (up to 20 μl)
    XXXXX is such that the final concentration of these 20uL is 60pM. So for example, if you have a 1.5nM library, or 1500pM, you want to dilute it 25 fold (1500/60) so you would take 0.8uL library (0.8uL into 20uL is a 25 fold dilution).
    Just as another example, if you have 0.6nM library, to get to 60pM you want to dilute 10 fold, so you would take 2uL library into the NaOH denaturing reaction.
    (Alternatively- you can take more library and then further dilute with hyb buffer, but I am not sure why we would want to do that. For example, you could take from a 1.5nM library 13.3uL, which would give you a 1nM concentration in the 20uL (1.5*13.3/20). This you would then dilute in hyb buffer to a final concentration of 10pM in 120uL for center to load onto the cBot.)
    (Alternatively, you could take the 600pM library, take 4uL, add 1uL 0.25 N NaOH, add 1uL 0.25N HCl, which would give you a 400pM solution. Now you take 3 uL of this into 117 uL hyb buffer for a final concentration of 10pM.)
    2. Vortex briefly to mix the template solution.
    3. Centrifuge the template solution to 280 xg for one minute.
    4. Incubate for five minutes at room temperature to denature the template into single strands.
    5. Add 1uL 1N HCl to neutralize the NaOH. Vortex briefly and spin down.
    6. Transfer 20 μl of denatured template to a tube containing 100 μl of pre-chilled HT1 (Hybridization Buffer) to a final concentration of 10pM for loading onto chip (our center usually uses 8-12pM, recommended we use 10pM for a start).
    7. From this step, keep the denatured template DNA on ice until you are ready to proceed to hybridization by cBot.

  • #2
    Please see my responses in Red.

    Originally posted by Noa View Post
    Hi,
    I know this has been covered by many threads (I've read them all) but am looking for a few clarifications.

    (1) I understand Illumina recommendations are such that you dont want to go past max 0.05 N NaOH during the DNA denaturization steps. But the threads I have seen here all had a 0.1N final NaOH concentraion during denaturing. Is this intentional/a mistake/a change in Illumina recommendations?
    If you're following the MiSeq protocol, it uses 0.2N NaOH in equal volume with the library, so the denaturing conditions will be 0.1N. I don't believe that the cBot protocol has been changed and still uses 0.1N NaOH.

    (2) Do you prepare libraries with the hyb buffer right before hyb to cBot (we are preparing it in an institute separate from the genome facility running the samples) or can they be saved for a few days at -20C (similar to Illumina recommendations for phiX libraries?) Anyone have experience here?
    Keep your libraries at 2nM (or 10nM) in TE or Illumina RSB until ready for cBot. Diluted DNA does not store very well. If the concentrations are lower than 2nM don't mess with it at all.

    (3) Does the concentration at which the libraries are when they undergo denaturization matter? (One thread- http://seqanswers.com/forums/showthread.php?t=4435 - commented what the concentration was during denaturing was and I wasnt sure whether it was important)
    The library concentration does matter somewhat during denaturing, it's not as important as the final concentrations.

    (4) would be super-happy if someone could give my protocol a quick run-through before I actually go and do it.... especially happy if someone can look at the different calculations in part 1 and see if they all work out right and if any is preferable over the other.

    1. Combine the following in a microcentrifuge tube:
    • template DNA (XXXXX μl)
    • 1 N NaOH (1 μl)
    • Tris-Cl 10 mM, pH 8.5 (up to 20 μl)
    XXXXX is such that the final concentration of these 20uL is 60pM. So for example, if you have a 1.5nM library, or 1500pM, you want to dilute it 25 fold (1500/60) so you would take 0.8uL library (0.8uL into 20uL is a 25 fold dilution).
    Just as another example, if you have 0.6nM library, to get to 60pM you want to dilute 10 fold, so you would take 2uL library into the NaOH denaturing reaction.
    (Alternatively- you can take more library and then further dilute with hyb buffer, but I am not sure why we would want to do that. For example, you could take from a 1.5nM library 13.3uL, which would give you a 1nM concentration in the 20uL (1.5*13.3/20). This you would then dilute in hyb buffer to a final concentration of 10pM in 120uL for center to load onto the cBot.)
    (Alternatively, you could take the 600pM library, take 4uL, add 1uL 0.25 N NaOH, add 1uL 0.25N HCl, which would give you a 400pM solution. Now you take 3 uL of this into 117 uL hyb buffer for a final concentration of 10pM.)
    2. Vortex briefly to mix the template solution.
    3. Centrifuge the template solution to 280 xg for one minute.
    4. Incubate for five minutes at room temperature to denature the template into single strands.
    5. Add 1uL 1N HCl to neutralize the NaOH. Vortex briefly and spin down.
    6. Transfer 20 μl of denatured template to a tube containing 100 μl of pre-chilled HT1 (Hybridization Buffer) to a final concentration of 10pM for loading onto chip (our center usually uses 8-12pM, recommended we use 10pM for a start).
    7. From this step, keep the denatured template DNA on ice until you are ready to proceed to hybridization by cBot.

    I got lost trying to follow what you're saying here. But here's my procedure for preparing low concentration libraries for clustering. This method has worked for me with libraries as low as 200pM.
    1. Add 18uL of library
    2. Add 2uL of 1N NaOH
    3. Incubate 5 minutes at room temperature, then place on ice
    4. Dilute to desired clustering concentration in HT1
    5. Add 2uL of 1N HCl to normalize pH. You can also use lower concentrations of HCl or a buffer if you like. The goal is to lower the pH to more neutral levels.
    6. Add PhiX (optional)

    Comment


    • #3
      Thanks for taking the time to answer: Further clarifications/questions:

      (a) regarding the NaOH concentration- I am using the HiSeq 2500 - their protocol says to mix 10uL 0.1N NaOH with 10uL library, which would give a final concentration of 0.05N. Is this different in MiSeq?

      (b) my library concentrations are low (<1nM) so I won't dilute them at all before preparing for the cBot. What I meant here was, after denaturing and adding hyb buffer, do I nned to immediately load this on the cBot or will this keep for some time?

      (c) What range should I try to have my library concentration during denaturing?

      (d) Thanks for the protocol. So for example if you do this on a 200pM library, after the NaOH you get a concentration of 180pM. Then if I want to load 10pM onto the cBot I need to dilute this 18-fold, so I would take 6.7uL of my denatured library into 113.3uL hybridization buffer?

      Thanks!
      noa

      Comment


      • #4
        If you add HCl, you neutralise the pH, so concentration of NaOH doesn't become an issue.

        My protocol:

        XXuL of Library for a 200pM in 20uL final volume (eg if lib is 500pM, then add 8uL)
        XXuL water (for above example add 8uL water)
        2uL 2N NaOH
        --------------------
        Total volume = 18uL

        Incubate at RT for 5 mins

        Add 2uL 2N HCl

        Final volume is 200pM in 20uL. Dilute with HT1 to required concentration

        Comment


        • #5
          Are you also using MiSeq? Why is everyone using a final of 0.1N NaOH? The protocol I have from Illumina uses a final of 0.05N (see below):

          1 Combine the following volumes of template DNA and 0.1 N NaOH in a
          microcentrifuge tube:
          • 2 nM template DNA (10 μl)
          • 0.1 N NaOH (10 μl)
          2 Vortex briefly to mix the template solution.
          3 Centrifuge the template solution to 280 xg for one minute....

          Thanks for the info!

          Comment


          • #6
            Where are you getting your Illumina protocol from? All the versions I have seen call for you to dilute the NaOH to 0.2 N (200 ul 1 N + 800 ul H2O) then add an equal amount to your DNA - 0.1 N. This is for the MiSeq.

            Comment


            • #7
              the protocol is from our genome center- Illumina cBot user guide - (manual is Illumina part # 15006165 rev. K) from Oct 2012

              (attached)

              note it specifically says:

              NOTE
              If your application requires higher than a 20 pM final concentration of your library,
              make sure that your concentration of NaOH is not higher than 0.05 N in the
              denaturation solution and not more than 0.001 N (1 mM) in the final solution after
              diluting with HT1.

              1 Combine the following volumes of template DNA and 0.1 N NaOH in a
              microcentrifuge tube:
              • 2 nM template DNA (10 μl)
              • 0.1 N NaOH (10 μl)
              Attached Files

              Comment


              • #8
                That's your problem! I think you're getting answers about the MiSeq and the HiSeq interchangeably. I think that the previous posts about using more concentrated NaOH, but neutralizing it after denaturation, should work. I would call or email Illumina tech support with your proposed method and see what they think.

                Comment


                • #9
                  Neutralizing the reaction in a small volume can lead to variability ( ostensibly by partial/reannealing of adapter regions), we've found that neutralizing with HCl after addition of HT1 dramatically reduces our variability.

                  ...and I've seen absolutely no difference betwen 50mM and 100mM final NaOH in the denaturation. Unclear why ILMN changed it.

                  Comment


                  • #10
                    So if you added 2 ul of 1N NaOH for denaturation, you would do all the dilutions up to 600 ul for the MiSeq (or whatever you wanted) with HT1, then add 2 ul of 1N HCl?

                    Comment


                    • #11
                      Has anyone ever used this protocol on a HiSeq (and not a MiSeq)?
                      Thanks!

                      Comment


                      • #12
                        You are correct, the cBOT protocol recommends 0.5N NaOH denaturation and the MiSeq protocol 0.1N NaOH.

                        Regardless of the concentration of NaOH used to denature libraries, 0.05N or 0.1N, the rule of thumb is that [NaOH] in your final hyb not exceed 1mM. If [NaOH] is >1mM, a buffer must be used to netrualize the NaOH, as others have suggested.

                        Comment


                        • #13
                          So does anyone have a clue why you would use different final NaOH concentrations to denature the same DNA? The libraries should be the same in the end regardless of whether they go on MiSeq or HiSeq, no? The same NaOH would denature the DNA I would assume? Just curious....

                          Comment


                          • #14
                            Illumina has been playing around with the denaturing protocol of the MiSeq quite a bit in the past year. The initial protocol was identical to the cBot. In revision B, they doubled the NaOH concentration for denaturing, but did nothing to reduce the final NaOH concentration. Now they're at revision G where they've changed the protocol again where you use 4nM of DNA at 5uL and 0.2N of NaOH at 5uL to dilute the libraries, thus bringing the final NaOH concentrations back down to where they started.

                            Comment


                            • #15
                              Originally posted by kcchan View Post
                              Illumina has been playing around with the denaturing protocol of the MiSeq quite a bit in the past year. The initial protocol was identical to the cBot. In revision B, they doubled the NaOH concentration for denaturing, but did nothing to reduce the final NaOH concentration. Now they're at revision G where they've changed the protocol again where you use 4nM of DNA at 5uL and 0.2N of NaOH at 5uL to dilute the libraries, thus bringing the final NaOH concentrations back down to where they started.
                              I fail to understand why people add back HCl at the end (I understand neutralising the NaOH, and bringing the pH back to "normal"), but shouldn't this be accomplished largely by the dilution of the denatured library in the HT1 buffer? In other words, when did the "add HCl" step start getting introduced and is there any evidence that it helps? Thanks.

                              Comment

                              Latest Articles

                              Collapse

                              • seqadmin
                                Recent Advances in Sequencing Analysis Tools
                                by seqadmin


                                The sequencing world is rapidly changing due to declining costs, enhanced accuracies, and the advent of newer, cutting-edge instruments. Equally important to these developments are improvements in sequencing analysis, a process that converts vast amounts of raw data into a comprehensible and meaningful form. This complex task requires expertise and the right analysis tools. In this article, we highlight the progress and innovation in sequencing analysis by reviewing several of the...
                                Yesterday, 07:48 AM
                              • seqadmin
                                Essential Discoveries and Tools in Epitranscriptomics
                                by seqadmin




                                The field of epigenetics has traditionally concentrated more on DNA and how changes like methylation and phosphorylation of histones impact gene expression and regulation. However, our increased understanding of RNA modifications and their importance in cellular processes has led to a rise in epitranscriptomics research. “Epitranscriptomics brings together the concepts of epigenetics and gene expression,” explained Adrien Leger, PhD, Principal Research Scientist...
                                04-22-2024, 07:01 AM

                              ad_right_rmr

                              Collapse

                              News

                              Collapse

                              Topics Statistics Last Post
                              Started by seqadmin, Today, 06:57 AM
                              0 responses
                              7 views
                              0 likes
                              Last Post seqadmin  
                              Started by seqadmin, Yesterday, 07:17 AM
                              0 responses
                              13 views
                              0 likes
                              Last Post seqadmin  
                              Started by seqadmin, 05-02-2024, 08:06 AM
                              0 responses
                              19 views
                              0 likes
                              Last Post seqadmin  
                              Started by seqadmin, 04-30-2024, 12:17 PM
                              0 responses
                              21 views
                              0 likes
                              Last Post seqadmin  
                              Working...
                              X