Old 04-04-2012, 03:44 AM   #1
sanju0891
Member
 
Location: germany

Join Date: Mar 2012
Posts: 15
Question qPCR Kapa Kit Experimental Design

Hello folks,

In our lab we are thinking to use the Kappa library quant kit for Roche GS FLX System on a Lightcycler 480 for the first time.

in the protocol of the manufacturer (Kappa) they suggest to do a triplicate for every sample and standard. so is it really necessary to do triplicates for all samples we have. we have like 40 samples to be sequenced, so if i do a triplicate for each sample one 96 well plate wont be sufficient.

did somebody also try the 10Ál reaction voulme instead of 20Ál? is there any difference between them.

and is it necessary to do 3-4 serial dilutions of the samples. which dilutions have worked better (1:500 or 1:1000)?

pls help me with these...........
sanju0891 is offline   Reply With Quote
Old 04-04-2012, 03:50 AM   #2
ETHANol
Senior Member
 
Location: Western Australia

Join Date: Feb 2010
Posts: 308
Default

No it is overkill, but it also depends on how well you can pipet and the quality of your pipets. I would choose one dilution and you could do duplicates to be safe. Use the least diluted sample that falls within the standard curve.
__________________
--------------
Ethan
ETHANol is offline   Reply With Quote
Old 04-04-2012, 03:56 AM   #3
sanju0891
Member
 
Location: germany

Join Date: Mar 2012
Posts: 15
Default

so i do not need to do triplicates for every sample?

sorry for asking this but im really confused about this. what protocol should i follow correctly
sanju0891 is offline   Reply With Quote
Old 04-04-2012, 04:04 AM   #4
sanju0891
Member
 
Location: germany

Join Date: Mar 2012
Posts: 15
Default

should i have to prepare the 6 DNA standards triplicates for each plate i run or only one of the six is enough for each plate????

so many questions regarding this? Please can anyone tel mein detail how to do this procedure for my 40 samples
sanju0891 is offline   Reply With Quote
Old 04-04-2012, 06:55 AM   #5
Tom Barker
Member
 
Location: Norwich

Join Date: Jun 2011
Posts: 10
Default

you will need to prepare all six standards as they are used for the standard curve which is used to calculate the concentration of your samples. You can set these up in triplicate but i only set duplicates. I've never used the light cycler before but on the Step One instrument i use you can choose to exclude wells from the standard curve if for example you have made a pipetting error when setting up the plate. If your careful with your pipetting you shouldn't need to set up triplicates.
Tom Barker is offline   Reply With Quote
Old 04-04-2012, 09:11 PM   #6
RCJK
Senior Member
 
Location: Australia

Join Date: May 2009
Posts: 155
Default

For rapid libraries I do 1:500 and 1:1000 in triplicate and leave out the first standard as it's too high. If your samples are amplicons you'll need to start at a higher dilution.
RCJK is offline   Reply With Quote
Old 05-26-2012, 10:20 PM   #7
mohdsos
Member
 
Location: Saudi Arabia

Join Date: Oct 2010
Posts: 15
Default

Guys would you dilute the 10^7 tube, or the original tube, because I am trying to pool 6 libraries.
mohdsos is offline   Reply With Quote
Old 05-30-2012, 03:40 AM   #8
sanju0891
Member
 
Location: germany

Join Date: Mar 2012
Posts: 15
Default

we dilute the original to further.
sanju0891 is offline   Reply With Quote
Old 05-31-2012, 11:33 AM   #9
genomeseeker
Member
 
Location: DC

Join Date: May 2009
Posts: 16
Default

We've found these dilutions work over 90% of the time and we do 10uL reactions in triplicate
rapid libs = 1:400
PE libs = 1:10,000
Amplicon libs = 1:50,000
genomeseeker is offline   Reply With Quote
Reply

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 11:18 AM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2020, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO