Old 05-06-2015, 12:20 PM   #1
Dr. Hybe
Member
 
Location: Cambridge

Join Date: Jun 2009
Posts: 14
Default Anyone had success with Illumina V2 Rapid 2x250?

Our institute is having trouble completing these long runs. They seem to die towards the end of both reads, camera goes out of focus, etc. They are being run on an upgraded HS2000. FYI, the samples all passed on MiSeq 2x300. We are looking for slightly more yield meta, amp.Has anyone had similar issues, any success, any advice?
Dr. Hybe is offline   Reply With Quote
Old 05-06-2015, 01:20 PM   #2
dpryan
Devon Ryan
 
Location: Freiburg, Germany

Join Date: Jul 2011
Posts: 3,480
Default

I guess it depends on how good of success you're after. We had a 2x250 v2 rapid run finish this morning and the results were relatively good. There was decreased quality toward the end of each read, but this was due primarily to two tiles (I think they were the exact same ones for each of the reads), so it turned out that the vast majority of reads were quite good. I should note that there was a bigger issue with the end of read 2, where a large block of tiles had problems. The latter issue combined with the same tiles showing some quality problems in both read 1 and 2 would be compatible with a camera problem, since there was an obvious per-tile difference...thereby suggesting that the problem isn't due to the chemistry itself.

I can't give any advice, as I don't personally deal with the sequencer, but that's at least the sort of thing that we're getting at the moment.
dpryan is offline   Reply With Quote
Old 05-06-2015, 01:22 PM   #3
GenoMax
Senior Member
 
Location: East Coast USA

Join Date: Feb 2008
Posts: 7,022
Default

With an n=1 we had no problems. Got about 250M reads per lane.
GenoMax is offline   Reply With Quote
Old 05-07-2015, 06:52 AM   #4
Dr. Hybe
Member
 
Location: Cambridge

Join Date: Jun 2009
Posts: 14
Default

Are you guys using a native 2500 V4 capable instrument, or an older upgraded 2000?
Dr. Hybe is offline   Reply With Quote
Old 05-07-2015, 07:19 AM   #5
GenoMax
Senior Member
 
Location: East Coast USA

Join Date: Feb 2008
Posts: 7,022
Default

Our run was on an upgraded older 2000.
GenoMax is offline   Reply With Quote
Old 05-08-2015, 07:54 AM   #6
pmiguel
Senior Member
 
Location: Purdue University, West Lafayette, Indiana

Join Date: Aug 2008
Posts: 2,317
Default

Quote:
Originally Posted by GenoMax View Post
With an n=1 we had no problems. Got about 250M reads per lane.
So, 125M pass filter clusters per lane, right?

--
Phillip
pmiguel is offline   Reply With Quote
Old 05-08-2015, 08:07 AM   #7
GenoMax
Senior Member
 
Location: East Coast USA

Join Date: Feb 2008
Posts: 7,022
Default

Quote:
Originally Posted by pmiguel View Post
So, 125M pass filter clusters per lane, right?

--
Phillip
Correct. ~140 M per lane. 79% > Q30.
GenoMax is offline   Reply With Quote
Old 05-28-2015, 07:14 AM   #8
Vinz
Member
 
Location: Germany

Join Date: Dec 2010
Posts: 80
Default

I hope the problem has been solved by now!
Anyway: With 2 native HS2500 we regularly get >300M cpf per flowcell and between 80 and 85% Q30 for amplicon data.
Vinz is offline   Reply With Quote
Old 05-29-2015, 06:15 AM   #9
Dr. Hybe
Member
 
Location: Cambridge

Join Date: Jun 2009
Posts: 14
Default

We've given up for the time being in favor of V1 rapids. Our instrument is an older one that we bought in 2011 and then upgraded to 2500. We're thinking that the camera/optics aren't capable of running 500 cycles?
Dr. Hybe is offline   Reply With Quote
Reply

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 11:38 PM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2020, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO