Default Whole-genome amplification prior to Illumina sequencing?

Hello,

Does anyone have any experience sequencing samples on the Illumina platform that have been whole-genome amplified? Has anyone explicitly tested whether it introduces allelic/amplification bias that could be a problem for SNP calling?

We've heard rumors and impressions that say it's a problem and should be avoided. But we have valuable samples that we would like to sequence and WGA seems to be the only way.

Thanks in advance.
Crawford505 is offline   Reply With Quote
Old 11-11-2010, 07:35 AM   #2
henry.wood
Member
 
Location: Leeds, UK

Join Date: Apr 2010
Posts: 63
Default

How tiny are your samples. If you are gentle you can make a perfectly good library with a lot less than 1ug.
henry.wood is offline   Reply With Quote
Old 11-11-2010, 07:40 AM   #3
Crawford505
Junior Member
 
Location: Ithaca, NY

Join Date: Jun 2009
Posts: 4
Default

Many of them are in the 400-500ng region. Have you had success with samples in this range. Do you modify the standard protocol to accommodate the lower grammage?
Thanks.
Crawford505 is offline   Reply With Quote
Old 11-11-2010, 07:46 AM   #4
henry.wood
Member
 
Location: Leeds, UK

Join Date: Apr 2010
Posts: 63
Default

We don't bother tweaking anything until we get to less than 100ng, so hopefully 400-500ng should be fine using normal protocols. If you are feeling twitchy you could practice using 400ng of a non-precious sample. You can tell if a library has worked by looking at the agilent/picogreen results without having to sequence it.
Once we get to low concentration we start to fiddle around with adaptor concentrations and PCR cycles.
henry.wood is offline   Reply With Quote
Old 11-11-2010, 07:57 AM   #5
Crawford505
Junior Member
 
Location: Ithaca, NY

Join Date: Jun 2009
Posts: 4
Default

Thanks very much that is helpful.

However, we are considering outsourcing our sequencing (the price is right) and the facility that would do it for us is requiring what i'm assuming they are considering 'safe' amounts of starting material.

So whole-genome amplification is still something we may want to look into. Anyone have any experience with this?
Crawford505 is offline   Reply With Quote
Old 11-11-2010, 08:24 AM   #6
jasonwood
Member
 
Location: RI

Join Date: May 2010
Posts: 10
Default

I used to use Sigma WGA to amplify ChIP samples for microarrays. Haven't needed to do this since I routinely make illumina libraries from 20ng or so starting material, and have been getting good sequencing results. One thing to keep in mind is that many WGA systems are PCR based, so if you then make libraries from them, you'll just end up sequencing the WGA adapter oligo instead of the sample. Maybe another WGA technique like MDA might work, but I don't have any experience with that.
But since the illumina library prep already has an amplification built in, i don't see any need to pre-amplify. you can use very small starting amounts. Find a place that will sequence pre-made libraries if necessary, and make them yourself.
jasonwood is offline   Reply With Quote
Old 11-13-2010, 01:39 PM   #7
csquared
Member
 
Location: Huntsville, AL

Join Date: May 2008
Posts: 67
Default

We haven't had any trouble making libraries down to 10ng input in our service lab and lower in the R&D lab. You need to appropriately alter the ratios in the library prep but it's nothing difficult.
__________________
HudsonAlpha Institute for Biotechnology
http://www.hudsonalpha.org/gsl
csquared is offline   Reply With Quote
Old 11-14-2010, 06:22 PM   #8
krobison
Senior Member
 
Location: Boston area

Join Date: Nov 2007
Posts: 747
Default

A number of vendors are introducing kits for smaller amounts of DNA; my notes say that NuGEN's requests 200ng and I know somewhere another company claims to have kits with a 100ng requirement (but those notes aren't making themselves found).

I also have data from WGA & at a crude level it looks fine. One thing to be careful of is any procedure using PCR -- you do risk having the PCR primer(s) occupying the ends of many of your reads.
krobison is offline   Reply With Quote
Old 11-19-2010, 03:47 AM   #9
Crawford505
Junior Member
 
Location: Ithaca, NY

Join Date: Jun 2009
Posts: 4
Default

Just so we're clear, we are interested in library preps of genomic DNA for DNA-seq runs. Are the numbers referred to here applicable to gDNA preps?
Crawford505 is offline   Reply With Quote
Old 11-19-2010, 09:20 AM   #10
jasonwood
Member
 
Location: RI

Join Date: May 2010
Posts: 10
Default

Yes, you should be fine with just a few ng of fragmented gDNA. Most of my samples are chip-seq (which is also fragmented genomic), and i start with 20ng if available, but routinely go down below 10ng without many problems if the IP is low efficiency. Certainly anything in the 20-100 range is quite robust (with diluted adapter primers).
jasonwood is offline   Reply With Quote
Old 11-20-2010, 08:56 AM   #11
kainsteven
Member
 
Location: SF Bay Area

Join Date: Jul 2009
Posts: 16
Default

The Ovation WGA FFPE System from NuGEN can be used for DNA from FFPE or Fresh Frozen (FF) material with input amounts as low as 100 ng.

Steve
kainsteven is offline   Reply With Quote
Old 01-20-2011, 01:19 PM   #12
rthornton4
Junior Member
 
Location: Houston, Texas

Join Date: Sep 2010
Posts: 6
Default

Hi,
We just attempted to sequence a WGA ChIP-seq library on our GA-II and it failed miserably. The adapter that Sigma uses in their WGA kit confused the RTA software and what the sequencer did call was junk. From talking with our FAS, the problem is that the new software needs to identify all four bases to call any single cluster and this adapter does not accommodate that. Currently we are being told that the only work-around is to design a sequencing primer that will hybridize to the adapter sequence or run these samples on the older software.
rthornton4 is offline   Reply With Quote
Old 01-20-2011, 05:33 PM   #13
frozenlyse
Senior Member
 
Location: Australia

Join Date: Sep 2008
Posts: 136
Default

Alternatively you could concatamerise your WGA'd material and refragment, that way the WGA primer sequences will not be on the end of your fragments - of course your downstream analysis will be made that much harder... You are probably better off optimising Illumina library prep for small amounts of DNA (which many labs have done)
frozenlyse is offline   Reply With Quote
Old 01-21-2011, 04:54 AM   #14
rthornton4
Junior Member
 
Location: Houston, Texas

Join Date: Sep 2010
Posts: 6
Default

Thanks frozenlyse for the advice! Unfortunately I didn't prepare the WGA sample. It was prepped by an outside user and I made the library. I do agree with optimizing library preps with smaller amounts of input but I don't have all of the information on why the user WGA'd their material. Another user had done this to compare WGA material vs. non-WGA material to see what, if any, the differences were in the sequencing. They had found the WGA material was highly comparable to the non-WGA material. I don't know if that was this users intent. All I do know is that they are not happy with Illumina currently.
rthornton4 is offline   Reply With Quote
Old 01-26-2011, 02:28 AM   #15
elaney_k
Member
 
Location: Dublin, Ireland

Join Date: Mar 2008
Posts: 55
Default

Quote:
Originally Posted by rthornton4 View Post
Thanks frozenlyse for the advice! Unfortunately I didn't prepare the WGA sample. It was prepped by an outside user and I made the library. I do agree with optimizing library preps with smaller amounts of input but I don't have all of the information on why the user WGA'd their material. Another user had done this to compare WGA material vs. non-WGA material to see what, if any, the differences were in the sequencing. They had found the WGA material was highly comparable to the non-WGA material. I don't know if that was this users intent. All I do know is that they are not happy with Illumina currently.
Hi,

I've managed to sequence samples for a collaborator on the Illumina GA that were target-enriched using the nimblegen/454 prep. The input DNA had the 454 sequencing adapters on either end of the DNA (22 bases long on the 5' end). I thought about concatamerising the DNA before doing an Illumina prep but figured the downstream bioinformatics might be a bit of a nightmare so prepped them without it. I made sure not to overload the flowcell and recalled all the data from the saved images. I got great usable data from it. I have to say in your case Illumina can't really be blamed, if you understand how the cluster calling algorithm/chemistry works then it's much easier to predict these kinds of problems.

Elaine
elaney_k is offline   Reply With Quote
Old 04-13-2011, 07:59 AM   #16
sshell
Junior Member
 
Location: Boston

Join Date: Jan 2009
Posts: 5
Default Adjustments for low input

Quote:
Originally Posted by jasonwood View Post
Haven't needed to do this since I routinely make illumina libraries from 20ng or so starting material, and have been getting good sequencing results.
Jasonwood, can you tell me what modifications (if any) you make to your library construction protocol when starting with low inputs such as 20 ng? Do you use less adapter, for example?

Thanks so much!
sshell is offline   Reply With Quote
Old 04-13-2011, 09:19 AM   #17
jasonwood
Member
 
Location: RI

Join Date: May 2010
Posts: 10
Default

I dilute the adapters down 1:20 (from 50uM stock concentration). I will usually increase the PCR cycles from 15 to 17-18, especially for ChIP samples.
I use Ampure XP beads for all purification steps, as I've found them less lossy than QIagen columns for low amounts of DNA. Also I size select a pretty wide band (200-500bp).
jasonwood is offline   Reply With Quote
Old 02-29-2012, 08:52 AM   #18
lterhune
Member
 
Location: Boulder, CO

Join Date: Sep 2011
Posts: 19
Default

Jasonwood, are you using the TruSeq kit to prepare the ChIP libraries or the Paired-End kit? We had lousy results with our first round of TruSeq ChIP libraries, but the protocol is not yet optimized.

We are also trying to create a library with about 200ng of gDNA- we have both the TruSeq and Paired-End kits, and our own reagents for our "homemade" Paired-End protocol.

Thanks!
lterhune is offline   Reply With Quote
Old 02-29-2012, 09:01 AM   #19
jasonwood
Member
 
Location: RI

Join Date: May 2010
Posts: 10
Default

I don't use Illumina kits, everything is homebrew. For Truseq adapters, the conditions are a little different than what I outlined above. They self-ligate much more readily than the older style short adapters, so I dilute them down more. For 20ng ChIP input, I do 1:50 at least. And change the Ampure bead ratio from 1.8:1 to 1:1 for the cleanup steps after the ligation and PCR enrichment (this effectively depletes fragments shorter than 200bp). Finally, I started measuring input to the PCR enrichment, and I put 2ng DNA in for 10 cycles. This gives you plenty of product and ensures you don't run out of PCR primers. If the starting amount is especially low (e.g. 5ng) or the post-ligation yield is very low, I will sometimes boost the cycle number, but generally now I use fewer cycles than I used to. I've had great results with this method, and minimal adapter contamination in my reads.
jasonwood is offline   Reply With Quote
Old 02-29-2012, 09:10 AM   #20
lterhune
Member
 
Location: Boulder, CO

Join Date: Sep 2011
Posts: 19
Default

Awesome, thank you .

We are also trying to use as few PCR cycles as possible because we having issues with high duplication rates. I haven't tried limiting the input ligation for the PCR enrichment before...
lterhune is offline   Reply With Quote
Reply

Tags
illumina, snp calling, solexa, whole genome amplifcation

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 04:12 PM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2020, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO