Old 11-16-2016, 09:27 AM   #1
EcoRInya
Member
 
Location: UK

Join Date: Apr 2016
Posts: 10
Default Weird overrepresented K-mers in FastQC

Hi! We have recently done the first run on a new NextSeq 500 machine which produced reads with weird FastQC results (attaching below).

This is a paired end sequencing of MNase-seq sample. Libraries were prepared using NEBNext Ultra kit using TruseqLT Illumina adapters. Sequencing was done using NextSeq 500 sequencing kit v2 high output.

I don't have a lot of experience with sequencing so far but FastQC results like these look worrying for me.

Could please anyone explain to me what are these numerous k-mers overrepresented along the reads, including the center of the read (which means that is it is unlikely to be untrimmed adapters)? Is this a problem with library prep or with the actual sequencing on the machine?
Attached Images
File Type: png Read2_KmerContent.png (105.5 KB, 13 views)
File Type: png Read2_Overrepresented_sequences_Adapter_content.png (25.5 KB, 8 views)
File Type: png Read1_Kmer_content.png (92.8 KB, 6 views)
File Type: png Read1_Overrepresentedsequences_adapterContent.png (22.2 KB, 4 views)
EcoRInya is offline   Reply With Quote
Old 11-16-2016, 10:48 AM   #2
GenoMax
Senior Member
 
Location: East Coast USA

Join Date: Feb 2008
Posts: 6,302
Default

I am not familiar with MNase-seq but perhaps that enriches some sequence that may be leading to that particular enrichment of k-mers.

Different criteria (pass/fail) used by FastQC are settings that can be changed and are configured (default) for normal genomic sequencing. Any experiment that deviates from the norm is likely to generate some fail flags. That does not automatically make your data bad. You should keep those observations in mind as you proceed with your analysis. If you are unable to get reasonable alignments then you can circle back and check if these k-mers are responsible for it.

The string of poly G's is a characteristic of NextSeq data. NextSeq uses a 2-color chemistry so absence of any signal is considered a G. See this link. You can trim those reads out using GGGGG as seed for the trimming program.

Last edited by GenoMax; 11-16-2016 at 10:50 AM.
GenoMax is offline   Reply With Quote
Old 11-17-2016, 12:53 AM   #3
EcoRInya
Member
 
Location: UK

Join Date: Apr 2016
Posts: 10
Default

Thanks, GenoMax! I guess we can test it the next time when we sequence a different type of libraries (like from cross-linked ChIP) and look if we still see these weird k-mers. The prediction is that we shouldn't. I didn't know about GGGGG in NextSeq data. Thank you for pointing this out!
EcoRInya is offline   Reply With Quote
Reply

Tags
fastqc, kmer, nextseq 500

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 11:14 AM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2017, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO