Old 01-10-2017, 06:45 AM   #1
thermophile
Senior Member
 
Location: CT

Join Date: Apr 2015
Posts: 167
Default MiSeq v3 low diversity

I haven't tried a v3 2x300 kit in over a year. Back then it had worse quality than v2 2x250 at similar cycle (v3 cycle 200 was much worse than v2 cycle 200) and the extra 50 bases were very poor. Now I'm having issues with a particular set of 16s samples (other 16s libraries work fine, this one set of samples is failing in R2) so talking to tech support who is telling me that v3 is now optimized for low diversity.

Has anyone see an improvement in low diversity libraries on v3 kits over the past year?
__________________
Microbial ecologist, running a sequencing core. I have lots of strong opinions on how to survey communities, pretty sure some are even correct.
thermophile is offline   Reply With Quote
Old 01-11-2017, 12:36 AM   #2
Tim NGS
Junior Member
 
Location: Antwerp, Belgium

Join Date: Mar 2015
Posts: 4
Default

We noticed an increase of quality since march/april last year. On our MiSeq's we mainly run targeted resequencing (amplicon), so also (quite) low diversity. Even when using v3 2x300, almost all runs were above Illumina specs (mostly Q30 above 80%). Error rates also seem to have improved.

What I can suggest is to keep cluster density quite low, we target below 900-850 K/mm˛ with a PhiX spike in ranging from 3-5%.

In general quality of read 2, based on SAV, can still drop quite low a the end of the reads.
Tim NGS is offline   Reply With Quote
Old 01-11-2017, 05:48 AM   #3
thermophile
Senior Member
 
Location: CT

Join Date: Apr 2015
Posts: 167
Default

v2 I generally aim to cluster 650-700 (though tech support is now telling me to go higher) and phiX and/or wgs 20% of the run
__________________
Microbial ecologist, running a sequencing core. I have lots of strong opinions on how to survey communities, pretty sure some are even correct.
thermophile is offline   Reply With Quote
Old 01-12-2017, 03:42 AM   #4
Tim NGS
Junior Member
 
Location: Antwerp, Belgium

Join Date: Mar 2015
Posts: 4
Default

I think you can aim for slightly higher cluster densities, but raising too much will definitely compromise on average quality.

Issues Illumina had with chemistry in 2015, are to our opinion, mostly solved.
Tim NGS is offline   Reply With Quote
Reply

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 05:49 AM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2017, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO