Old 11-18-2016, 12:56 PM   #1
ox_zoo
Junior Member
 
Location: Oxford

Join Date: Nov 2016
Posts: 5
Default Low QC scores?

I have just set up my very first sequencing experiment on the Next-Seq 500. It is now 45 cycles into PE sequencing, however it seems that my QC scores are a bit low. I only have a % >= Q30%: 25.2%

I have around:
280 million reads in the QC 10 bin.
25 million in the QC 20 bin
8 million in the QC 25 bin
65 million in QC 30 bin
44 million in QC 35 bin.

Is this a bad result? Apologies for my stupidness, just was wondering whether data could be salvaged from this?
ox_zoo is offline   Reply With Quote
Old 11-19-2016, 05:12 AM   #2
ox_zoo
Junior Member
 
Location: Oxford

Join Date: Nov 2016
Posts: 5
Default

Yes, now it has finished and the QC is only 15.2.

The machine also has a warning message on two of the lanes:

camera disabled:failed to detect clusters


Does this mean that the cameras are not working properly?

Thanks
ox_zoo is offline   Reply With Quote
Old 11-19-2016, 06:29 AM   #3
GenoMax
Senior Member
 
Location: East Coast USA

Join Date: Feb 2008
Posts: 6,190
Default

Probably. You should contact Illumina tech support. They can take a look at your sequencer remotely to diagnose the problem.

If your instrument is not directly accessible over the network then you will need to send them some diagnostic files.
GenoMax is offline   Reply With Quote
Old 11-19-2016, 06:37 AM   #4
ox_zoo
Junior Member
 
Location: Oxford

Join Date: Nov 2016
Posts: 5
Default

Thanks for your reply.

Yes I will do that - unfortunately since its a weekend they cannot reply.

I was thinking - if clusters are not formed in those two particular areas, would the cameras be disabled? My pooled library concentration was sufficient, so I did not anticipate any problems. Also, if the cluster density is low then I am not sure how QC would be affected..

Many thanks

Last edited by ox_zoo; 11-19-2016 at 06:39 AM.
ox_zoo is offline   Reply With Quote
Old 11-19-2016, 06:42 AM   #5
GenoMax
Senior Member
 
Location: East Coast USA

Join Date: Feb 2008
Posts: 6,190
Default

What was the cluster density for this run?

If the run was borderline overclustered to begin with, it is possible that sequencer lost the ability to distinguish the clusters over time (they become fatter). There is also the possibility of this being some sort of a hardware issue.
GenoMax is offline   Reply With Quote
Old 11-19-2016, 06:48 AM   #6
ox_zoo
Junior Member
 
Location: Oxford

Join Date: Nov 2016
Posts: 5
Default

The cluster density for the four lanes range from 0-250k/mm2.

The lanes:
Lane 1: 235 41
Lane 2: 237 38
Lane 3: 212 56
Lane 4: 230 35

I hope this helps.
ox_zoo is offline   Reply With Quote
Old 11-19-2016, 07:50 AM   #7
GenoMax
Senior Member
 
Location: East Coast USA

Join Date: Feb 2008
Posts: 6,190
Default

This run is borderline overclustered (as I recollect for NextSeq optimal cluster density is 170-220 K cluster/mm^2).

Since there is nothing to lose at this time contact tech support and see what they say. You may have to eat the cost on this one and re-run.
GenoMax is offline   Reply With Quote
Reply

Tags
illumina, next-seq, qc scores

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 01:14 AM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2017, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO