Old 02-18-2015, 05:54 PM   #1
dlawrence
Junior Member
 
Location: South Australia

Join Date: Mar 2012
Posts: 8
Default HiSeq2500 - how to sequence pair after single end run

Our machine was prepped for a Paired End run, but we forgot to enter the read length for the paired read, so the machine ran it through as single end.

The job has completed, and the flow cell is sitting there.... it seems that all we would need to do is convince the Illumina software to turn the molecules over and start the second run.

Illumina tech support say this is impossible.

Does anyone know how to save this run?

I renamed/move all of the files produced during the run (netcopy files, Processing & Data directories), modified the runParameters.xml and restarted the program, hoping to resume a sequencing run, but it didn't work.

Edit: Idea

We have thought of an idea, which is to do a second run, with 0 bases (or 1) for the first read, then continue on with the pair. We can then match reads from the 1st run with the 2nd via tile ID.

Does anyone have any comments on whether this would work?

Last edited by dlawrence; 02-18-2015 at 06:06 PM.
dlawrence is offline   Reply With Quote
Old 02-19-2015, 02:26 PM   #2
ScottC
Senior Member
 
Location: Monash University, Melbourne, Australia.

Join Date: Jan 2008
Posts: 246
Default

Hi,

I don't think you'll be able to do another run with 0 bases, because the instrument will try to carry out clustering again, which would probably be bad for your already clustered flowcell.

I suspect that you might have to start again.

Cheers,

Scott.
ScottC is offline   Reply With Quote
Old 02-19-2015, 04:29 PM   #3
ECO
--Site Admin--
 
Location: SF Bay Area, CA, USA

Join Date: Oct 2007
Posts: 1,355
Default

I doubt this is possible, IME Illumina's tech support will tell you off-label ways to do things (with disclaimers) if they know how it's possible.

You will not be able to match up the new read2's because the XY positions of the clusters will be different.

Also, moving this to Illumina forum.
ECO is offline   Reply With Quote
Old 02-19-2015, 09:41 PM   #4
HESmith
Senior Member
 
Location: Bethesda MD

Join Date: Oct 2009
Posts: 499
Default

The most straightforward option is to rehybridize the flow cell and start a new paired-end run. You lose the existing read one data, but salvage the PE flow cell/clustering. It might be technically feasible to modify the existing run data to convert it to a paired end run and restart it but, without guidance from Illumina, it's virtually impossible to determine which of the many files require editing to make this work.
HESmith is offline   Reply With Quote
Reply

Tags
hiseq 2500, hiseq paired end

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 04:22 AM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2018, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO