Old 05-11-2016, 03:08 AM   #1
jpw39
Junior Member
 
Location: London

Join Date: Jan 2014
Posts: 2
Default Unexpected peaks on Bioanalyzer trace, KAPA hyper prep for ChIP-seq library

Hi all,

Just wondering if anyone has observed high molecular weight peaks on their bioanlyzer traces for library preps - KAPA or otherwise. These were made from ChIP material with KAPA hyper prep kit, TruSeq adapters and KAPA's primers for amplification from P5/P7 sequences.

Thanks in advance...
Attached Images
File Type: png KAPA_hyperprep_library.png (86.2 KB, 33 views)
jpw39 is offline   Reply With Quote
Old 05-11-2016, 05:00 AM   #2
nucacidhunter
Jafar Jabbari
 
Location: Melbourne

Join Date: Jan 2013
Posts: 1,182
Default

I wonder if you have any trace from input into library. Possible source of large fragments:
1- Suboptimal shearing of ChIP DNA
2- Anomaly in bioanalyser run due to library contaminants or run reagents

PCR would be less likely source of large fragments because polymerase will not be able to copy such large fragments
nucacidhunter is offline   Reply With Quote
Old 05-11-2016, 05:15 AM   #3
jpw39
Junior Member
 
Location: London

Join Date: Jan 2014
Posts: 2
Default

Thanks,

I suspect the input chromatin was not optimally sheared and I'm not surprised to see that the library material is quite a broad smear. However, the sharp peaks seen running larger than the 10kb marker are particularly strange to me, I can't think how the library prep would give rise to these.

A bioanalyzer anomaly seems more likely but I had 3 samples on the chip that had not been subjected to library prep which had no such peaks.
jpw39 is offline   Reply With Quote
Old 05-11-2016, 10:08 PM   #4
nucacidhunter
Jafar Jabbari
 
Location: Melbourne

Join Date: Jan 2013
Posts: 1,182
Default

If the library was run before clean up it will affect DNA migration speed on the Chip causing anomaly.

If you have left over from input material you can run them on the Chip and look for size distribution to see if larger fragments originate from input. The sharp peak could be unsheared chromatin mass.
nucacidhunter is offline   Reply With Quote
Reply

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 04:06 PM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2018, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO