Default Poly-G Failed Runs on NextSeq

Approximately a month ago, our lab performed an RNASeq using a TruSeq Stranded mRNA library prep and a v2 75cyc High Output flowcell. The data that came back was almost entirely G's and was severely underclustered. When we went to re-run the same library, the machine died and would soon thereafter require all of the boards to be replaced.

We assumed that the poly-G run was an early symptom of the board failure, until we tried to run a genomic library prepared with Nextera DNA reagents on v2 300cyc High Output flowcell and again got nearly all G's (also very underclustered). There were no shared reagents between the two library preps and our lab has performed many of each type of prep without failure.

Thinking that it was a hardware issues still, we did a phiX run on a 300cyc Mid Output flowcell... which of course came out perfect. Illumina is of the opinion that it is a transient issue and that we should just continue trying, but there's got to be more to it than that. Have any of you run into this problem before or have thoughts on what may be going on?
djs150 is offline   Reply With Quote
Old 11-27-2016, 04:46 AM   #2
GenoMax
Senior Member
 
Location: East Coast USA

Join Date: Feb 2008
Posts: 6,497
Default

Have you tried to sequence the libraries on a different sequencer?
GenoMax is offline   Reply With Quote
Old 11-27-2016, 06:07 AM   #3
djs150
Junior Member
 
Location: Pittsburgh

Join Date: Oct 2015
Posts: 6
Default

We have not tried them on a different machine yet. Illumina was first insistent that it was a library prep issue, so at the time it seemed wasteful to run on a separate machine.
djs150 is offline   Reply With Quote
Old 11-27-2016, 11:12 AM   #4
GenoMax
Senior Member
 
Location: East Coast USA

Join Date: Feb 2008
Posts: 6,497
Default

You may want to spike one of the libraries with someone else (as long as the tags allow) who may be running on a MiSeq/HiSeq. That would confirm that quality of your libraries before you try a re-run on NextSeq.
GenoMax is offline   Reply With Quote
Old 11-27-2016, 08:45 PM   #5
wdecoster
Member
 
Location: Antwerp, Belgium

Join Date: Oct 2015
Posts: 95
Default

Given that NextSeq uses two color chemistry, poly G essentially means an absence of signal...
wdecoster is offline   Reply With Quote
Old 11-28-2016, 12:04 AM   #6
Tim NGS
Junior Member
 
Location: Antwerp, Belgium

Join Date: Mar 2015
Posts: 4
Default

Make sure no custom primers were selected during the run setup. Information can be found at bottom of the "runparameters.xml"-file. Just look for "<UsesCustomReadOnePrimer>true"

Sounds stupid, but we also had it once when the custom primer button was pressed accidentally when initiating an experiment. The system will use the wrong primer for first read sequencing, so no signal is generated and thus no clusters will be detected. Due to the chemistry, there is no signal from G-bases. Run will look severely underclustered and possibly only G's will be called.
Tim NGS is offline   Reply With Quote
Old 11-28-2016, 06:04 AM   #7
djs150
Junior Member
 
Location: Pittsburgh

Join Date: Oct 2015
Posts: 6
Default

At this point, I welcome things that sound stupid if it means we can get this wrapped up. Unfortunately the run parameters show false for the custom primers and everything else in there looks as expected.
djs150 is offline   Reply With Quote
Old 11-28-2016, 06:29 AM   #8
GenoMax
Senior Member
 
Location: East Coast USA

Join Date: Feb 2008
Posts: 6,497
Default

Only options I see is to cross your fingers and try a re-run with NextSeq or check the libraries on a MiSeq/HiSeq to confirm that they are sequenceable before coming back to NextSeq.
GenoMax is offline   Reply With Quote
Old 11-29-2016, 04:19 PM   #9
djs150
Junior Member
 
Location: Pittsburgh

Join Date: Oct 2015
Posts: 6
Default

As an update, we've now run the same library on a different NextSeq and obtained the same results. The library was run with a 1% PhiX control, which sequenced perfectly.

We thought that, due to the high GC nature of the libraries, the strands weren't denaturing. The NaOH stock used was over pH 13 and a Qubit before and after denaturing showed that they are certainly single stranded when they are being loaded.

So the problem is in the library prep, somewhere. The only common reagent is the Omega Bio-tek beads from the PCR clean up. Is there any rationale that the beads could be causing this problem somehow?
djs150 is offline   Reply With Quote
Old 11-29-2016, 04:51 PM   #10
GenoMax
Senior Member
 
Location: East Coast USA

Join Date: Feb 2008
Posts: 6,497
Default

This is likely an indication of a bad library prep and you may be better off starting over.

As a last try do a MiSeq run, if you can. MiSeq is a resilient sequencer and may be able to sequence this sample. Having 4 color chemistry helps as well. If the library has low nucleotide diversity then increase the phiX concentration.

Last edited by GenoMax; 11-29-2016 at 04:54 PM.
GenoMax is offline   Reply With Quote
Old 11-29-2016, 05:16 PM   #11
Brian Bushnell
Super Moderator
 
Location: Walnut Creek, CA

Join Date: Jan 2014
Posts: 2,695
Default

I think if the same library failed with the same error mode on a different machine, it would be a waste of resources to run it again. Considering the PhiX reads sequenced perfectly, I cannot imagine that a different platform would give different results. It sounds like you just don't have any sequence (since as wdecoster noted, 'G' means no signal). But to clarify, are the sequences 100% G, or are they ~60bp of adapter and then poly-G?
Brian Bushnell is offline   Reply With Quote
Old 11-29-2016, 05:28 PM   #12
GenoMax
Senior Member
 
Location: East Coast USA

Join Date: Feb 2008
Posts: 6,497
Default

I think two separate libraries were originally tried but at least one of the libraries was run again and failed.

If this is an irreplaceable sample then try the last ditch effort I mentioned above otherwise I agree that it would be best to start over and make new libs.
GenoMax is offline   Reply With Quote
Old 11-29-2016, 05:47 PM   #13
Brian Bushnell
Super Moderator
 
Location: Walnut Creek, CA

Join Date: Jan 2014
Posts: 2,695
Default

Quote:
Originally Posted by GenoMax View Post
If this is an irreplaceable sample then try the last ditch effort I mentioned above
Good point, I did not consider that aspect. If that's the situation I concur with Genomax.
Brian Bushnell is offline   Reply With Quote
Old 11-29-2016, 05:48 PM   #14
djs150
Junior Member
 
Location: Pittsburgh

Join Date: Oct 2015
Posts: 6
Default

They are 100% G, adapter and all. And yes, there were originally two unrelated libraries that failed and the second was re-run on a different machine.

The samples are by no means irreplaceable as we have the original DNA to remake the libraries. However, if I'm missing something obvious that is going to lead to another failed run (i.e. bad reagent), then remaking the libraries isn't going to help.

I'm simply at a loss to explain why Qubit and Tapestation indicate there is tons of DNA there and yet I can't seem to get a viable read on in it.
djs150 is offline   Reply With Quote
Old 11-29-2016, 06:26 PM   #15
Brian Bushnell
Super Moderator
 
Location: Walnut Creek, CA

Join Date: Jan 2014
Posts: 2,695
Default

I wonder if...

A) The adapters never attached to the reads
or
B) The reads were to long to bridge-amplify (in which case a different platform might make a difference)

I don't know how to diagnose those issues, but they seem like the most likely culprits to me.
Brian Bushnell is offline   Reply With Quote
Old 11-29-2016, 11:02 PM   #16
nucacidhunter
Senior Member
 
Location: Iran

Join Date: Jan 2013
Posts: 1,060
Default

Quote:
Originally Posted by djs150 View Post

So the problem is in the library prep, somewhere. The only common reagent is the Omega Bio-tek beads from the PCR clean up. Is there any rationale that the beads could be causing this problem somehow?
I wonder if this is the first time that you have used this brand or batch of bead.
nucacidhunter is offline   Reply With Quote
Old 11-30-2016, 08:08 AM   #17
djs150
Junior Member
 
Location: Pittsburgh

Join Date: Oct 2015
Posts: 6
Default

All of the reagents and procedures have been used in identical fashions for many months without issue. Regarding the beads specifically, we moved to this brand after having issues with Agencourt. It also didn't hurt that they were cheaper.

I agree with the idea that the adapters are not attaching to the reads, I'm going to run with that hypothesis and see where I end up.
djs150 is offline   Reply With Quote
Old 11-30-2016, 05:33 PM   #18
nucacidhunter
Senior Member
 
Location: Iran

Join Date: Jan 2013
Posts: 1,060
Default

To identify adapter ligation issues for RNA-Seq library you can do qPCR quantification and compare results to Qubit. If adapters have not been ligated then qPCR quantification should be substantially less than Qubit.

Although you have not run Nextera library for the second time but I doubt that they will have adapter issue because without addition of adapter sequences it will be impossible to PCR amplify those libraries.
nucacidhunter is offline   Reply With Quote
Old 12-01-2016, 12:20 PM   #19
RickC7
Member
 
Location: Baton Rouge, Louisiana

Join Date: Feb 2010
Posts: 27
Default

Similar but different issue on Next-Seq.

I am of course screwed b/c I'm using a custom Tag-sequencing protocol, no custom primers as we have Gex adapters. Lanes 1 and 3 look great, but lanes 2 and 4 contain a high number of poly-G reads. Do a phix run and everything looks good. Have not repeated the custom libraries a 2nd time, but a smaller sample subset did sequence fine on our Mi-Seq. I don't understand how the cluster is ID if there is no signal? The run is a short 51bp read and I have reads that are all G's, like 100 million of these reads.

Illumina actually mentioned they are aware of lane-lane variation when sequencing some custom libraries and that since the library is not validated and PhiX works, basically too bad. That the libraries are sequencing fine in lane 1 and 3 should be enough validation. I just feel something else is at work and not library issue in our case.

Anyone see similar when sequencing custom libraries on the Next-Seq?

EDIT: Quantified using Kapa qPCR for Illumina, got great cluster density using the concentration obtained.

Last edited by RickC7; 12-01-2016 at 12:23 PM.
RickC7 is offline   Reply With Quote
Old 01-04-2017, 08:20 AM   #20
Yepler
Member
 
Location: Tucson, AZ, USA

Join Date: Oct 2010
Posts: 22
Default

Quote:
Originally Posted by RickC7 View Post
Similar but different issue on Next-Seq.

Lanes 1 and 3 look great, but lanes 2 and 4 contain a high number of poly-G reads. Do a phix run and everything looks good. Have not repeated the custom libraries a 2nd time, but a smaller sample subset did sequence fine on our Mi-Seq. I don't understand how the cluster is ID if there is no signal? The run is a short 51bp read and I have reads that are all G's, like 100 million of these reads.

Illumina actually mentioned they are aware of lane-lane variation when sequencing some custom libraries and that since the library is not validated and PhiX works, basically too bad. That the libraries are sequencing fine in lane 1 and 3 should be enough validation. I just feel something else is at work and not library issue in our case.

Anyone see similar when sequencing custom libraries on the Next-Seq?
.
Yep. I had the exact same issue (two lanes okay, two lanes pretty much all poly-G) when using a custom sequencing primer on the NextSeq. The NS is really sensitive to primer Tm. Are you using a custom sequencing primer? Do you know what the Tm of your primer is, and are you able to raise it? I ended up adding some LNA bases to mine (couldn't increase the length) and that fixed the problem. I'd be happy to talk in more detail if you like, just send a message.
Yepler is offline   Reply With Quote
Reply

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 07:09 AM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2017, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO