Default low fluorescence intensities read 2, HiSeq 2000

hi everyone
I have a problem with PE module Hiseq 2000
Basically the instrument read well the read 1 and the read index, but when arrive to the read 2, the fluorescence intensities are very low and sometime with Q30 around 0 (RNAseq libraries read at 50 PE).
I have opened a case with Illumina 2 month ago but I haven't the solution yet.
Someone have had the same problem?

Thanks a lot
LUR is offline   Reply With Quote
Old 01-03-2017, 11:37 AM   #2
pmiguel
Senior Member
 
Location: Purdue University, West Lafayette, Indiana

Join Date: Aug 2008
Posts: 2,236
Default

Well, unless you are using crazy old single read Illumina adapters, it sounds like something went awry during turn-around.
If you are under a service contract you should ask Illumina to replace the reagents so you can give it another shot.

--
Phillip
pmiguel is offline   Reply With Quote
Old 01-04-2017, 05:54 AM   #3
helskel
Junior Member
 
Location: Ann Arbor

Join Date: Jun 2015
Posts: 4
Default Have you checked your NaOH?

It is also possible that you have an issue with the NaOH. If the concentration of your NaOH in the PE module is off, you will not get through turnaround efficiently, and that can lead to all of the issues that you described.
helskel is offline   Reply With Quote
Old 05-05-2017, 01:15 AM   #4
LUR
Junior Member
 
Location: Milan

Join Date: Sep 2011
Posts: 2
Default

The problem was with the wash and mantainence wash on HiSeq. Probably a contamination with something but we don't know the origin of this contamination. We have solved the problem restored the mantainance wash with NAOH 0.1M. Now the instrument work well.
LUR is offline   Reply With Quote
Old 05-05-2017, 01:29 PM   #5
misterc
Member
 
Location: Livermore, CA

Join Date: Jan 2016
Posts: 20
Default

If you switch back to the 0.5% Tween + 0.03% Proclin maintenance wash solution, be sure and rinse the NaOH from your lines. For good measure, I'd do two complete "maintenance wash" runs with just de-ionized water loaded in all reagent positions. This makes sure you don't get some nasty interaction between NaOH and the Tween/Proclin mix. The Tween/Proclin maintenance wash solution I've had much better luck with preventing clogs and such than with 0.1M NaOH, but it has the drawback of creating possible biofilms in the lines since Tween supports microbial growth if the Proclin doesn't sufficiently prevent it. Switching from Tween/Proclin to the 0.1M NaOH wash would also be best advised with a couple long water washes to rinse out the former.
misterc is offline   Reply With Quote
Reply

Tags
fluorescence, hiseq 2000, low intensity, paired end reads

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 09:36 AM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2017, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO