  • PacBio RS II SMRTcell performance variability

    We are observing increased variability in the performance of SMRTcells by lot, strip and within a strip. Using the same SMRTbell template on different sequencing runs the variability can be as much as a two fold difference between strips of SMRTcells. We would like advice from other users who may experiencing the same issue.

  • #2
    Are you referring to just successful loading? Or some other metrics (RL, accuracy)?

    Comment


    • #3
      @ribodoc: Was there a time when you were regularly getting consistent results and now you are not?

      Comment


      • #4
        Originally posted by ECO View Post
        Are you referring to just successful loading? Or some other metrics (RL, accuracy)?
        Within a week, with the same SMRTbell template and loading concentration went from 47.44% P1 to 26.64% P1. Accuracy dipped from 0.841 to 0.831. Polymerase read length N50 had a smaller drop from 16285 to 15849.

        Comment


        • #5
          Originally posted by GenoMax View Post
          @ribodoc: Was there a time when you were regularly getting consistent results and now you are not?
          Occasionally we see this occur and it was time to put it out to the community. We typically do a titration of SMRTbell concentration for loading and then do a production run on a strip of SMRTcells. There will be periods of time when the tritration and production runs are very consistent and then there are other times when they are not consistent and even not consistent within a strip of SMRTcells.

          Comment


          • #6
            Originally posted by ribodoc View Post
            Occasionally we see this occur and it was time to put it out to the community. We typically do a titration of SMRTbell concentration for loading and then do a production run on a strip of SMRTcells. There will be periods of time when the tritration and production runs are very consistent and then there are other times when they are not consistent and even not consistent within a strip of SMRTcells.
            That has been more or less the story. While I don't work at the bench I hear similar observations most of the time. There is no logical explanation that can be put forth either. Hopefully Sequel will be more reliable.

            Comment


            • #7
              Originally posted by GenoMax View Post
              That has been more or less the story. While I don't work at the bench I hear similar observations most of the time. There is no logical explanation that can be put forth either. Hopefully Sequel will be more reliable.
              Thank you for the response.

              Comment


              • #8
                @Genomax...the logical explanation is that the SMRTcell production process is dramatically more complicated than all other sequencing consumables, ZMW size and uniformity, the micromirror shape, aluminum coating, then on top of that the blocking of the surface and then the biotinylation.

                I know it's no solace to scientists trying to do reproducible production (we run two RSII's all the time), but it's kind of amazing just how well it all works, and is a testament to all the people that have contributed at PACB. (disclaimer I worked there for a long time and know a lot of people there still)

                Comment


                • #9
                  I have nothing but admiration for the amount of technology that went into making the original PacBio machine, consumables.

                  Based on anecdotal references keeping an RSII running ALL the time seems to be essential for its well being. Sadly we could not do that.

                  Comment


                  • #10
                    The SMRT-cell quality (and correlated to it the yields) can indeed be quite variable in general .
                    The very last batches of SMRT-cells seem to be more variable than usual (for us and also some other labs). These higher frequency problems seem to occur at least once a year.

                    Comment

                    Latest Articles

                    Collapse

                    • seqadmin
                      Current Approaches to Protein Sequencing
                      by seqadmin


                      Proteins are often described as the workhorses of the cell, and identifying their sequences is key to understanding their role in biological processes and disease. Currently, the most common technique used to determine protein sequences is mass spectrometry. While still a valuable tool, mass spectrometry faces several limitations and requires a highly experienced scientist familiar with the equipment to operate it. Additionally, other proteomic methods, like affinity assays, are constrained...
                      04-04-2024, 04:25 PM
                    • seqadmin
                      Strategies for Sequencing Challenging Samples
                      by seqadmin


                      Despite advancements in sequencing platforms and related sample preparation technologies, certain sample types continue to present significant challenges that can compromise sequencing results. Pedro Echave, Senior Manager of the Global Business Segment at Revvity, explained that the success of a sequencing experiment ultimately depends on the amount and integrity of the nucleic acid template (RNA or DNA) obtained from a sample. “The better the quality of the nucleic acid isolated...
                      03-22-2024, 06:39 AM

                    ad_right_rmr

                    Collapse

                    News

                    Collapse

                    Topics Statistics Last Post
                    Started by seqadmin, 04-11-2024, 12:08 PM
                    0 responses
                    18 views
                    0 likes
                    Last Post seqadmin  
                    Started by seqadmin, 04-10-2024, 10:19 PM
                    0 responses
                    22 views
                    0 likes
                    Last Post seqadmin  
                    Started by seqadmin, 04-10-2024, 09:21 AM
                    0 responses
                    16 views
                    0 likes
                    Last Post seqadmin  
                    Started by seqadmin, 04-04-2024, 09:00 AM
                    0 responses
                    47 views
                    0 likes
                    Last Post seqadmin  
                    Working...
                    X