#1 |
Member
Location: Brasil Join Date: Jun 2011
Posts: 12
|
![]()
We started a MiSeq running and after the first cycle there was an error: "Best focus is too near edge of range: AutoFocus would have moved z to 0, outside soft limits os -0,295001144426642... -0,00499885557335775: ThroughFocusSacnCommand".
Is it a camera problem? May I start the running againg? Thanks in advance for your attention! |
![]() |
![]() |
![]() |
#2 |
Senior Member
Location: East Coast USA Join Date: Feb 2008
Posts: 7,087
|
![]()
Contact Illumina tech support and let them take a look at your sequencer.
|
![]() |
![]() |
![]() |
#3 |
Senior Member
Location: CT Join Date: Apr 2015
Posts: 243
|
![]()
There's no restarting a run with focus problems. It's unlikely to be a camera problem, more likely cluster failure, lens or z stage issue. Absolutely call tech support
__________________
Microbial ecologist, running a sequencing core. I have lots of strong opinions on how to survey communities, pretty sure some are even correct. |
![]() |
![]() |
![]() |
#4 |
Junior Member
Location: AFRICA Join Date: Mar 2018
Posts: 1
|
![]()
We have had this error several times as well. The run just stops without images to show whether there is under- or over-clustering. I am very doubtful if this is just a clustering issue. I think it is a hardware problem.
|
![]() |
![]() |
![]() |
#5 |
Junior Member
Location: USA Join Date: May 2018
Posts: 8
|
![]()
I had the same issue twice in a row at Miseq platform. How did you guys resolve the problem. I called Illumina and they just questioned a lot about my library.
Thanks |
![]() |
![]() |
![]() |
#6 |
Member
Location: QLD, Aus Join Date: Jan 2018
Posts: 19
|
![]()
I had the same issue - "Best focus too near edge of range" and after contacting tech support and being directed to run these tests:
Y Stage test M3 mirror test Z Stage test Compensator test Conduct volume test Thermal Ramp test Prime Reagent Lines Optics test I was told the run was significantly overclustered and to drop my loading concentration and reload. |
![]() |
![]() |
![]() |
#7 |
Junior Member
Location: USA Join Date: May 2018
Posts: 8
|
![]()
All the system checks passed.
The only issue for me why the quant of library measured by KAPA qPCR, Qubit, BioAnalyzer and LabChip GX are not consistent with each other. Which platform is reliable? Thanks |
![]() |
![]() |
![]() |
#8 |
Member
Location: Boston Join Date: Mar 2013
Posts: 24
|
![]()
KAPA is the most accurate, provided you calculate it correctly. We usually run librarys on BioA or AA to get the correct size distribution and use it to do the calculation with KAPA results.
|
![]() |
![]() |
![]() |
#9 |
Junior Member
Location: USA Join Date: May 2018
Posts: 8
|
![]()
ikripp, have you had a successful run recently? What is your loading concentration with how much % of PhiX? Thanks
|
![]() |
![]() |
![]() |
#10 |
Member
Location: QLD, Aus Join Date: Jan 2018
Posts: 19
|
![]()
We do approximately a run each week and vary rarely do we have what is considered an unsuccessful run. I base the loading concentration off of the pool concentrations as determined by a HSD1000 tape run on a tapestation 4200, again previous run pools. So its hard to say what loading concentration I have gone with but our standard PhiX is 10% which changes when we have a unique pool. I work with 16S amplicon pools for reference.
|
![]() |
![]() |
![]() |
#11 |
Junior Member
Location: USA Join Date: May 2018
Posts: 8
|
![]()
What is the loading concentration ( ng/ul) you got based on tape station?
|
![]() |
![]() |
![]() |
#12 |
Member
Location: QLD, Aus Join Date: Jan 2018
Posts: 19
|
![]() |
![]() |
![]() |
![]() |
#13 |
Junior Member
Location: Chile Join Date: Apr 2014
Posts: 8
|
![]()
Hi everyone!
Here I was searching about failed runs on MiSeq, and found this thread ![]() In my case, we do runs on MiSeq each month, and recently we've had problems with 2 runs (MiSeq v3, 600c). In the first failed run, it was a problem of focus ("Best focus is too near edge of range(...)"); on the 2nd failed run, it has no clustering ("No usable signal found in the images"). I'll post lot number, in case it helps someone else: MiSeq Reagent v3 600c box 1 of 2: LOT 20183241 and LOT 20270202 MiSeq Reagent v3 box 2 of 2: LOT 20178686 and LOT 20262956 We've also had the same answer from our local Illumina manager ... something like "it's your libraries fault" or "did you even load library on cartridge?". As we're no newbies, these answers doesn't help ![]() Has someone else had failed runs with MiSeq, or reagents v3?
__________________
Science is ok, but I'm hungry. |
![]() |
![]() |
![]() |
#14 | |
Member
Location: QLD, Aus Join Date: Jan 2018
Posts: 19
|
![]() Quote:
I've never seen an error of no usable signal. What do the thumbnail images look like? |
|
![]() |
![]() |
![]() |
#15 | |
Junior Member
Location: Chile Join Date: Apr 2014
Posts: 8
|
![]() Quote:
in the case of "no usable signal", we didn't even have thumbnail images, but we do have folder "Images", and there's a "Focus" -> "C1" folder, with .tif images file. Here an example of name: FocusImage_0_exp_200_Focus T_IM_zpos_-0.15_method_CoarseFocus.tif It looks like a huge black square. Literally no signal ![]()
__________________
Science is ok, but I'm hungry. |
|
![]() |
![]() |
![]() |
#16 |
Member
Location: Europe Join Date: Jul 2018
Posts: 12
|
![]()
Is a problem of focus alignment.
You need to realign the Focus position. |
![]() |
![]() |
![]() |