Old 12-21-2012, 11:53 AM   #21
Papaveraceae
Junior Member
 
Location: Kansas

Join Date: Dec 2012
Posts: 5
Default

...........................

Last edited by Papaveraceae; 12-21-2012 at 12:02 PM.
Papaveraceae is offline   Reply With Quote
Old 12-21-2012, 11:58 AM   #22
Papaveraceae
Junior Member
 
Location: Kansas

Join Date: Dec 2012
Posts: 5
Default

Quote:
Originally Posted by pmiguel View Post
I think if your RNA is degrading during the DNAse digestion, then it was not purified sufficiently in the first place. You probably got off easy. If it degraded during DNAse treatment at least you know you need to modify your prep. Otherwise you submit it for sequencing and it degrades during library creation. Plus you end up sequencing DNA as well as RNA.

--
Phillip
Phillip,

I made no mention of my RNA degrading after DNAse treatment, in fact I havent treated them with DNAse at all yet, on the contrary the RIN of 9.3-9.2 from my non-DNAsed RNA (extracted using TRI reagent) listed on the electropherogram that I posted is indicative of high quality RNA. Nor am I a proponent of not using DNAse on RNA samples. My message was in regards to the Zymo colums, I was wondering if they had any ill effects on the RNA profile.

Thanks,

Dave

Last edited by Papaveraceae; 12-21-2012 at 12:35 PM.
Papaveraceae is offline   Reply With Quote
Old 01-03-2013, 05:00 AM   #23
pmiguel
Senior Member
 
Location: Purdue University, West Lafayette, Indiana

Join Date: Aug 2008
Posts: 2,315
Default

Quote:
Originally Posted by Papaveraceae View Post
Phillip,

I made no mention of my RNA degrading after DNAse treatment, in fact I havent treated them with DNAse at all yet, on the contrary the RIN of 9.3-9.2 from my non-DNAsed RNA (extracted using TRI reagent) listed on the electropherogram that I posted is indicative of high quality RNA. Nor am I a proponent of not using DNAse on RNA samples. My message was in regards to the Zymo colums, I was wondering if they had any ill effects on the RNA profile.

Thanks,

Dave
Hi Dave,

What do you mean by "RNA profile"? If you mean what the RNA sample looks like on an Agilent nanochip, then yes, various people have complained of decreasing RIN scores (sometimes drastically) after DNAse treatment. If that is what you mean, then I refer you to the answer I gave previously.

If by "RNA profile" you mean the actual composition of your RNA sample before and after DNAseing as, for example, assayed by RNAseq -- well I have no reason to suspect that would change. But any purification has the potential to bias sample composition. It would be pretty hard to assay though and seems far-fetched, so the chances that someone actually did an experiment of this sort seem low.

If you mean something else, then maybe you would be will to invest 10% of the time it takes me to answer your post in composing it -- specifically by defining "RNA profile" as it exists in your idiolect!

No, you do not "have" to run an Agilent chip prior to DNAsing your sample. I mean, unless you have some sort of contractual obligation to Agilent of which I am unaware. You could use some other method of quantitation, if you like. But, I am just stating the obvious here.

--
Phillip
pmiguel is offline   Reply With Quote
Old 02-27-2013, 09:02 AM   #24
Al Merry
Junior Member
 
Location: Oxford

Join Date: Feb 2013
Posts: 3
Default

I have total RNA samples prepared using the Ambion Mirvana kit containing miRNA which I am interested in sequencing, as well as larger RNA species, so I need both small and large RNAs to survive downstream procedures. I now need to DNAse them, hence my reply to this thread. I have done a few reactions using a Qiagen DNAse in solution then purifying on an RNEasy Minelute column, which typically gave me about 65-70% recovery from 1-3ug (which is acceptable, although if anyone routinely gets greater recovery I'd love to hear how you do it). But some samples repeatedly give only about 20% yield, which is obviously unacceptable - perhaps those particular RNA samples contained RNAses, but the RIN and everything was great before the reaction.

I have been advised to use the Ambion Turbo DNAse, which I avoided before because I hate using the inactivation resin - in the past using its predecessor, I used to spin the supernatant twice and leave quite a lot behind to ensure no resin carry over. I've been told that column purification is the way to go. My question is, do you use the Turbo DNAse inactivation resin before column cleanup or do you just put the DNAse reaction straight onto the column?

And I've read about people worrying that DNAse treatment will just chop up the DNA into smaller bits that will add background to miRNA sequencing - is this a valid concern?

Thanks for any help,

Al

Last edited by Al Merry; 02-28-2013 at 01:29 AM.
Al Merry is offline   Reply With Quote
Old 05-03-2013, 01:00 PM   #25
maryluzyl
Junior Member
 
Location: Bogota, Colombia

Join Date: Apr 2013
Posts: 2
Default help how can i use zymo columns

Hi Phillip
I have a problem with DNA contamination in my RNA samples for RNA seq. So you said you use zymo columns for DNAse remotion, what is the protocol for DNAse remotion, how many steps? Do you elute RNA with DEPC water?
Thanks so much
Mary Luz

Quote:
Originally Posted by pmiguel View Post
Listen, you can attempt to convince yourselves that DNAse treatment is not necessary all you want, but that does change the fact that failing to do it can result in DNA sequence contamination of your RNAseq results.

Also, if your RNA is degrading during DNAse treatment, the simple explanation is that your preps still have RNases in them. That is, your RNA prep has been unsuccessful at completely removing them. So, even if you skip the DNAse step, to spare your RNA, all you accomplish is a circumvention of the QC step -- your RNA will likely degrade during a later step.

A "standard paradigm protocol" would have you isolate intact RNA free of contaminating RNases and depleted of most DNA. Then DNase treat to remove the remaining DNA, followed by some purification to remove the DNase and oligonucleotides. (I like Zymo columns for this purpose.)

No one likes to see their RNA degrade during DNase treatment, but better it happens there than at a later step where library QC may not catch it.

--
Phillip
maryluzyl is offline   Reply With Quote
Old 06-22-2013, 12:12 PM   #26
Apis37
Junior Member
 
Location: California

Join Date: Sep 2011
Posts: 3
Default

I don't see how contamination (a little or a lot) can cause differential expression between genes in a RNA-Seq analysis. The DNA is not present is different numbers between samples, so if there is contamination its just noise. Saying you are going to reject any paper that does not do a DNAase treatment is nonsense. I think the biggest problem right now with genomics is people inventing rules based on what they imagine should be the case, when really these techniques are new and we do know exactly what is and isn't necessary.
Apis37 is offline   Reply With Quote
Old 12-03-2013, 04:13 PM   #27
armins
Junior Member
 
Location: Champaign, IL

Join Date: Dec 2013
Posts: 3
Default

I recently extracted RNA from 20,000 sorted cells that I am going to perform RNA-seq on. Using RNeasy Micro/Mini kits, I would normally get ~ 50 ng total RNA in 20 µL. Given the minute amounts of RNA I was dealing with, I decided to skip the on-column DNase treatment. My Bioanalyzer results show that with the micro kit I have DNA contamination in my samples. The samples prepared by the mini kit are usually free of DNA contamination - only based on the Bioanalyzer results.

I decided to go ahead with what I have and do the RNA-seq, hoping that the poly(A) selection in the TruSeq kit will get rid of most of the DNA. Plus, I did not really have any other choice (because my total sample is so little). If I were to repeat my samples preps, though, I would stick to RNeasy Mini rather than Micro, as it seemed to yield better results, even though advertised otherwise.

I will evaluate my sequencing reads for the presence of intron/intergenic sequences and post the results on here. Best of luck all -
armins is offline   Reply With Quote
Old 12-04-2013, 08:02 AM   #28
pmiguel
Senior Member
 
Location: Purdue University, West Lafayette, Indiana

Join Date: Aug 2008
Posts: 2,315
Default

Quote:
Originally Posted by armins View Post
I recently extracted RNA from 20,000 sorted cells that I am going to perform RNA-seq on. Using RNeasy Micro/Mini kits, I would normally get ~ 50 ng total RNA in 20 µL. Given the minute amounts of RNA I was dealing with, I decided to skip the on-column DNase treatment. My Bioanalyzer results show that with the micro kit I have DNA contamination in my samples. The samples prepared by the mini kit are usually free of DNA contamination - only based on the Bioanalyzer results.

I decided to go ahead with what I have and do the RNA-seq, hoping that the poly(A) selection in the TruSeq kit will get rid of most of the DNA. Plus, I did not really have any other choice (because my total sample is so little). If I were to repeat my samples preps, though, I would stick to RNeasy Mini rather than Micro, as it seemed to yield better results, even though advertised otherwise.

I will evaluate my sequencing reads for the presence of intron/intergenic sequences and post the results on here. Best of luck all -
I don't get it. How can a bioanalyzer chip estimate the amount of genomic DNA in your sample? Max size detectable on a bioanalyzer (nano or pico RNA chip) is around 10,000 nt. Genomic DNA is 50,000+ bp.

--
Phillip
pmiguel is offline   Reply With Quote
Old 12-04-2013, 08:38 AM   #29
armins
Junior Member
 
Location: Champaign, IL

Join Date: Dec 2013
Posts: 3
Default

Phillip,

All I know is that gDNA fragments are present in some of my samples, based on the shape of the "inter-region" (I'm trying to attach one of the Bioanalyzer results). I failed to mention that due to the processing of my samples, gDNA is sheared into pieces, so it would show up on the Bioanalyzer graph.
Attached Files
File Type: pdf Untitled.pdf (150.1 KB, 53 views)
armins is offline   Reply With Quote
Old 12-04-2013, 09:58 AM   #30
pmiguel
Senior Member
 
Location: Purdue University, West Lafayette, Indiana

Join Date: Aug 2008
Posts: 2,315
Default

Quote:
Originally Posted by armins View Post
Phillip,

All I know is that gDNA fragments are present in some of my samples, based on the shape of the "inter-region" (I'm trying to attach one of the Bioanalyzer results). I failed to mention that due to the processing of my samples, gDNA is sheared into pieces, so it would show up on the Bioanalyzer graph.
How did the genomic DNA get sheared?
Is the hump between the 18S and 28S peaks supposed to be DNA?

--
Phillip
pmiguel is offline   Reply With Quote
Old 12-04-2013, 10:25 AM   #31
armins
Junior Member
 
Location: Champaign, IL

Join Date: Dec 2013
Posts: 3
Default

Quote:
Originally Posted by pmiguel View Post
How did the genomic DNA get sheared?
Is the hump between the 18S and 28S peaks supposed to be DNA?
The lysate was passed through QIAshredder columns a couple of times which I believe breaks up the DNA. What that hump is? I am not positive, but according to Agilent in their case it is DNase-sensitive. Page 82 here:
https://www.chem.agilent.com/Library...991-3322EN.pdf
armins is offline   Reply With Quote
Old 06-11-2016, 09:16 PM   #32
Janan
Junior Member
 
Location: Kota Kinabalu, Sabah, Malaysia

Join Date: May 2016
Posts: 1
Default DNase treatment

Hello,

I am using mirVana™ miRNA Isolation Kit, to isolate small RNA from animals blood. Do I have to do DNase treatment for it?

Thanks,

Janan
Janan is offline   Reply With Quote
Reply

Tags
library prep, rna fragmentation, sample prep

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 07:47 PM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2019, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO