Old 06-18-2015, 01:36 AM   #1
anna_m
Member
 
Location: Edinburgh

Join Date: Jun 2013
Posts: 17
Default qPCR issues

Hello, we use Kapa qPCR kits to quantify our libraries before sequencing. Recently, the NTC wells are giving low cq values which means there are contaminants and therefore the data is not reliable. Despite cleaning the wells, changing kits, and re-aliquoting reagents, the NTC wells are still giving the same result.

Has anyone else had this issue? Does anyone have any suggestions?

Thanks in advance.
anna_m is offline   Reply With Quote
Old 06-18-2015, 05:37 AM   #2
microgirl123
Senior Member
 
Location: New England

Join Date: Jun 2012
Posts: 200
Default

How low are your values? Occasionally my NTC wells come up with very, very low values (way below the final standard). I just ignore it - compared to the inaccuracies caused by dilution and pipetting, it's meaningless.
microgirl123 is offline   Reply With Quote
Old 06-18-2015, 06:17 AM   #3
anna_m
Member
 
Location: Edinburgh

Join Date: Jun 2013
Posts: 17
Default

We have a minimum cut off of 28 for the cq values and we have been getting as low as 18 recently.
anna_m is offline   Reply With Quote
Old 06-19-2015, 12:25 PM   #4
helskel
Junior Member
 
Location: Ann Arbor

Join Date: Jun 2015
Posts: 4
Default

What do you mean by "cleaned the wells"? Do you re-use plates? That could be a source. Also, how clean is the water that you use for dilutions? If you didn't get new water for your tests, that is also a possibility. What do your NTC wells contain? Do you just use prepared Master Mix, or do you also add water?
helskel is offline   Reply With Quote
Old 06-22-2015, 01:11 AM   #5
anna_m
Member
 
Location: Edinburgh

Join Date: Jun 2013
Posts: 17
Default

No, we don't re-use plates. We clean the plate-holder wells with bleach, water and ethanol on recommendation by Kapa once a month. We don't use water to dilute the libraries, we use EB-tween. We put master mix and EB-tween in the NTC wells.
anna_m is offline   Reply With Quote
Old 06-22-2015, 07:25 AM   #6
Jessica_L
Senior Member
 
Location: Washington, D.C. metro area

Join Date: Feb 2010
Posts: 118
Default

Other than a few incidents, my experience has been the same as microgirl123-- if my last standard has a Cq value of 35, I might see some signal in the NTC around cycle 38, which corresponds to such a low level of DNA contamination I basically just don't pay it any mind.

When I've had issues in the past with higher levels of contamination in NTCs I took a pretty extreme approach-- use a PCR enclosure, wipe everything down with bleach then ethanol, including the enclosure surfaces and pipettors. I'd also usually leave the UV light on in the enclosure for at least thirty minutes prior to setup and made a point to leave pipettors, tips, water, any other non-sensitive consumables (i.e. plates, plate seals) in the hood.

I've also noticed that not all plate seals form a good seal against the plate, which would occasionally cause me problems when the plate was full enough to have NTC wells near my unknowns.
Jessica_L is offline   Reply With Quote
Reply

Tags
kapa, library, ntc, qpcr

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 10:24 PM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2021, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO