  • genomic DNA quantification

    My lab is a biobank that sends samples out to external labs for next gen sequencing. However, we find it nearly impossible to get matching results from our internal quantification methods and the lab that we are sending to.

    My lab uses picogreen quantification to quantify the input genomic DNA, we also have a Nanodrop and an Agilent Bioanalyzer. We have found that our 3 pieces of equipment often disagree with each other and of course our measurements disagree with external lab's measurements. Most labs that perform sequencing double-check the DNA concentration prior to running and require a certain range in order to "pass" a sample in their QC. And of course everyone thinks that their measurements are right and everyone else's are wrong. Very frustrating.

    Does anyone know of a better, more accurate measurement for DNA than the 3 listed above? I feel like since NextGen is so picky about this, there needs to be a better method?

    Thanks,
    Roxanne.

  • #2
    Picogreen or Qubit should be sufficiently accurate. The Nanodrop and Bioanalyzer will often give more varying results. The Nanodrop tends to overestimate the DNA concentration due to contamination with ssDNA/nucleotides/RNA/etc. The Bioanalyzer depends on accurately pipetting very small volumes, so it's always a bit tricky.

    In my experience, those three methods frequently disagree with each other, but measurements with Qubit on the same sample at different times have given fairly good reproduction.

    Comment


    • #3
      You should be fine using all these three methods. Did you ever try to compare the references your service provider is using for fluroesecent measurement against your PicoGreen measurement?

      Comment


      • #4
        qPCR

        We use Solid Taqman assays for our Solid machines. They have such kits for Illumina also. If I remember correctly, you can buy them from Kappa and Agilent. They are expensive, but then again so isn't Bioanalyzer chips. The advantage is that you are actually measuring amplifiable template, not just every DNA fragement in your sample. That is what really counts with Next Gen seq, are how many of those will become template for cluster generation or ePCR. BioA should be good to bring your multiplexed samples to equal conc. or to ballpark your library, but we always do a final qPCR before preparing our ePCR for sequencing.

        Comment


        • #5
          Originally posted by Roxanne Kelly View Post
          My lab is a biobank that sends samples out to external labs for next gen sequencing. However, we find it nearly impossible to get matching results from our internal quantification methods and the lab that we are sending to.

          My lab uses picogreen quantification to quantify the input genomic DNA, we also have a Nanodrop and an Agilent Bioanalyzer. We have found that our 3 pieces of equipment often disagree with each other and of course our measurements disagree with external lab's measurements.
          Hi Roxanne,
          In your case I would recommend that after your genomic DNA prep, you treat the DNA with a potent RNAse, like "RiboCruncher" and then deplete low molecular weight solutes somehow. We use a few cycles of dilution/concentration on Millipore Microcons, but they will not remove any contaminating high molecular weight proteins. So it might be worth looking into low X Ampure. I have not tried it with genomic DNA, so it might be problematic.

          As touched on above, what is happening is that the nanodrop is confounded by all manner of substances commonly found in genomic DNA. pico green is probably giving you the right answer. But if your core is using another method they may flag the discrepancy. Agilent chips are useless for assaying full length genomic DNA -- genomic DNA is too long for the molecules to migrate past the detector before the run is finished.

          Typical genomic DNA preps are >90% RNA and while RNAse will degrade the RNA into short oligo and mononucleotides, that will not decrease the amount of UV absorbed by those short oligos and mononucleotides. Further, even very low concentrations of phenol absorbs very strongly at 260 nm. That is, 0.1% phenol will absorb at 260 nm as strongly as 100's of ng/ul of DNA. And there are other substances that absorb at 260 nm.

          So you need a method to degrade all the RNA and then remove it all, along with any other UV absorbing substance that is not DNA.

          --
          Phillip

          Comment


          • #6
            Originally posted by JeremyDay View Post
            We use Solid Taqman assays for our Solid machines. They have such kits for Illumina also. If I remember correctly, you can buy them from Kappa and Agilent. They are expensive, but then again so isn't Bioanalyzer chips. The advantage is that you are actually measuring amplifiable template, not just every DNA fragement in your sample. That is what really counts with Next Gen seq, are how many of those will become template for cluster generation or ePCR. BioA should be good to bring your multiplexed samples to equal conc. or to ballpark your library, but we always do a final qPCR before preparing our ePCR for sequencing.
            I may be mistaken, but I took original post to mean that the samples being sent out were genomic DNA. (Eg, library construction being done by the core not by Roxanne's lab.)

            If Roxanne's lab is constructing the libraries themselves, then I agree with you. Except I don't know of commercial qPCR kits for Illumina library Taqman. We use the SYBR green one from Kapa.

            --
            Phillip

            Comment


            • #7
              @Rocketknight, which assay of Qubit seems to give more accurate results, is BR assay consistent for your samples?

              Comment


              • #8
                Hi Roxanne,

                We (Kapa Biosystems) have a qPCR-based method for quantifying hgDNA that also provides a quantitative measure of DNA quality across a broad range of samples. We presented this system in a poster at AGBT earlier this year. Unfortunately, we have not yet launched a commercially listed version of the kit, but we are supplying some early adopters with custom versions, and we have an on-going test program for the kit.

                If you would like more information, or if you are interested in testing the system, please let me know.

                Best regards,

                Eric

                Comment


                • #9
                  We also had this problem. We quantify using Picogreen (Qubit) and check integrity on a 2% agarose gel. Nanodrop usually over estimates (sometimes by as much as 10X) but we use it to check on 260/280 and 260/230 ratios.
                  We still had lots of discrepancy between what we quantified the DNA at and what our service provider estimated (even using the same method of QC). I think it's just one of those things.
                  In the end, we usually send 3-5X the recommended quantity and hope for the best. We still get samples dropping out but it's not as bad as it was.

                  We intended to look into DNA storage options (such as PureLink) to improve shipping conditions but never got round to it. Has anyone had any experience with such kits? Do they affect QC or library prep in any way?

                  Comment


                  • #10
                    Originally posted by Smriti View Post
                    @Rocketknight, which assay of Qubit seems to give more accurate results, is BR assay consistent for your samples?
                    BR has been consistent for us. I only use HS rarely.

                    Comment


                    • #11
                      Genomic DNA quantification

                      I think I know what is going on here and I am surprised nobody has ever mentioned it in this forum. I just discovered it after working with DNA for many, many years, probably because the Next Gen protocols require such small amounts of DNA.

                      Picogreen or Qubit methods are definitely best for getting accurate quantitation, because they only measure the dsDNA. However, people still get variable results. So where is this variability coming from? I believe it is from trying to pipette small volumes of HMW DNA in order to make the required dilutions. When microliter volumes of HMW DNA are pipetted, some of the DNA (or even all of it, believe it or not!) can be dragged back out of the pipette tip as the tip is removed from the source tube. You think you just transferred 100 ng, but it might actually be much less. Why does this happen? Simple. Good HMW DNA preps are a massive tangle of strands. The higher the concentration, the worse the tangle. To ensure accurate quantitation, you must work with DNA solutions of lower concentration (no more than 0.1 ng/ul) and avoid transferring volumes less than about 5 ul when making dilutions.

                      By the way protocols usually warn of shearing DNA if you pipette too vigorously, but I don't think this is much of a problem. The above issue is much more serious.

                      Comment


                      • #12
                        Hi creeves,
                        Thanks for bringing this up. I actually had some guys in our lab complaining about the viscosity of a genomic DNA prep they were working with -- thinking something was terribly wrong with it. Turns out they had concentrated it to > 1ug/ul, so yes, that is just the way genomic DNA is. And you are completely correct, you are never going to get an accurate concentration of the whole prep by taking an aliquot of that with a pipette.

                        I spent some significant fraction of my scientific youth doing genomic DNA preps in a lab where everyone did genomic DNA preps. So, to me it was completely obvious. Whole volumes of genomic DNA lore pass by my eyes every time I think about it. I literally just spent the last five minutes day dreaming about various genomic DNA prep methods of various resultant purities. And the attendant issues and even (tangentially) the old assay for determining the length of DNA molecules by the viscosity of the solutions it forms.

                        So many possible misunderstandings. Surprising protocols work at all, really.
                        --
                        Phillip

                        Comment


                        • #13
                          Hi,
                          Sometimes I like to run an 0.8% agarose gel with controls of known concentration with my samples and estimate the concentration based on staining of the sample. It may not be as accurate as picogreen, but it does allow you to see what is going on in your sample such as quality, possible contaminants (protein/RNA) and degredation, because picogreen will bind degrading dsDNA. But I find that most of the time the gel does correspond to other methods of quantification.

                          Comment


                          • #14
                            Originally posted by Rocketknight View Post
                            BR has been consistent for us. I only use HS rarely.
                            What is the range of DNA concentrations that you quantify? I am using Qubit BR assay kit to quantify genomic DNA of 5 ng/ul and I was wondering if I need to switch to the HS kit.

                            Comment


                            • #15
                              Below 100ng then HS, 100ng-1ug then BR.
                              It's also good practice to re-QC normalised samples too.

                              Comment

                              Latest Articles

                              Collapse

                              • seqadmin
                                Strategies for Sequencing Challenging Samples
                                by seqadmin


                                Despite advancements in sequencing platforms and related sample preparation technologies, certain sample types continue to present significant challenges that can compromise sequencing results. Pedro Echave, Senior Manager of the Global Business Segment at Revvity, explained that the success of a sequencing experiment ultimately depends on the amount and integrity of the nucleic acid template (RNA or DNA) obtained from a sample. “The better the quality of the nucleic acid isolated...
                                03-22-2024, 06:39 AM
                              • seqadmin
                                Techniques and Challenges in Conservation Genomics
                                by seqadmin



                                The field of conservation genomics centers on applying genomics technologies in support of conservation efforts and the preservation of biodiversity. This article features interviews with two researchers who showcase their innovative work and highlight the current state and future of conservation genomics.

                                Avian Conservation
                                Matthew DeSaix, a recent doctoral graduate from Kristen Ruegg’s lab at The University of Colorado, shared that most of his research...
                                03-08-2024, 10:41 AM

                              ad_right_rmr

                              Collapse

                              News

                              Collapse

                              Topics Statistics Last Post
                              Started by seqadmin, Yesterday, 06:37 PM
                              0 responses
                              8 views
                              0 likes
                              Last Post seqadmin  
                              Started by seqadmin, Yesterday, 06:07 PM
                              0 responses
                              8 views
                              0 likes
                              Last Post seqadmin  
                              Started by seqadmin, 03-22-2024, 10:03 AM
                              0 responses
                              49 views
                              0 likes
                              Last Post seqadmin  
                              Started by seqadmin, 03-21-2024, 07:32 AM
                              0 responses
                              67 views
                              0 likes
                              Last Post seqadmin  
                              Working...
                              X