![]() |
|
![]() |
||||
Thread | Thread Starter | Forum | Replies | Last Post |
Two lane failures and low cluster density - expired reagents? | Turnerac0987 | Illumina/Solexa | 2 | 08-25-2011 11:46 AM |
library concentration and cluster density | cybog337 | Illumina/Solexa | 0 | 01-12-2011 02:44 AM |
Cluster density Illumina | CatVincent | Introductions | 2 | 11-23-2010 10:50 AM |
Factors affecting cluster density | SeqTruth | Illumina/Solexa | 3 | 11-09-2010 06:59 AM |
Troubles with Cluster Density | andibody | Illumina/Solexa | 9 | 11-03-2009 10:28 AM |
![]() |
|
Thread Tools |
![]() |
#1 |
Member
Location: Oregon Join Date: Jan 2011
Posts: 24
|
![]()
Greetings,
We have been getting low cluster densities, and as we increase the concentration, we are getting less clusters/tile/pM. It looks like there is no way for our samples to reach the optimal density of 400,000. It would seem that there is something wrong with our RNA-seq samples other than concentration that is causing this trend. Ex: 9.6pM – 175,000 cl/tile (18,000 cl/tl/pM) – one lane 10pM – 146,000 cl/tile (15,000 cl/tl/pM) – one lane 20pM – 279,000 cl/tile (14,000 cl/tl/pM) – one lane 22pM – 261,000 cl/tile (12,000 cl/tl/pM) – one lane 25pM – 120,000 to 240,000 cl/tl (5,000 to 10,000 cl/tl/pM) – seven lanes, seven diff samples We are multiplexing – 6 barcodes per lane. Samples were quantified with the Qubit flourometer. The bioanalyzer showed multiple peaks for our earlier samples that were daisy-chained. The extra peaks were eliminated when we changed the gel extraction protocol from 50C to RT, and the bioanalyzer showed very nice peaks at approx 350bp. Recently, one of our samples was checked with qPCR and found to be ¼ the concentration of the Qubit reading. It had previously been sequenced at 25pM and yielded 150,000 cl/tl. 1) Does anyone get such drastically different concentration readings? 2) Any ideas as to why so little of the sample is sequencing when the bioanalyzer shows our library is the target length in bp and PCR enrichment works well? 3) Any ideas as to why higher concentrations are not increasing (often decreasing) the cluster density? I greatly appreciate any ideas or help you may have on these problems. Thanks, thanks and thanks!! |
![]() |
![]() |
![]() |
#2 |
Member
Location: CA Join Date: Aug 2009
Posts: 22
|
![]()
In terms of increasing concentrations leading to lower densities: Are you correctly following Illumina's recommendations for high concentration library denaturation/dilution? Excess hydroxide carryover from the denaturation step certainly can lead to this phenomenon.
|
![]() |
![]() |
![]() |
#3 |
Member
Location: Oregon Join Date: Jan 2011
Posts: 24
|
![]()
We use the core facilities here for ngs. I have not looked at the high concentration directions from Illumina, but it is my understanding that the core facility dilutes samples so that the NaOH concentration is less than 0.8mN. Whether or not this is the "recommendation" I do not know.
|
![]() |
![]() |
![]() |
#4 |
--Site Admin--
Location: SF Bay Area, CA, USA Join Date: Oct 2007
Posts: 1,358
|
![]()
What technology and kit are you using for your final library purification?
|
![]() |
![]() |
![]() |
#5 |
Member
Location: Oregon Join Date: Jan 2011
Posts: 24
|
![]()
We are using the Qiagen gel extraction kit with minielute columns. We have changed our gel several times, but presently we are using a 2% NuSieve 3:1 agarose gel (with EtBr) at RT for 80min.
|
![]() |
![]() |
![]() |
#6 |
--Site Admin--
Location: SF Bay Area, CA, USA Join Date: Oct 2007
Posts: 1,358
|
![]()
Post final PCR as well? Or no final PCR?
|
![]() |
![]() |
![]() |
#7 |
Member
Location: Oregon Join Date: Jan 2011
Posts: 24
|
![]()
Yes. We PCR enrich the 350bp cDNA 12-18 cycles and then gel extract again. This final sample is checked on the bioanalyzer and the qubit before mixing with other samples for multiplexing and sequencing.
|
![]() |
![]() |
![]() |
#8 |
Junior Member
Location: CT Join Date: Dec 2010
Posts: 5
|
![]()
What technique are you using to add the barcodes?
We found that the 3 primer-based barcoding strategy from Illumina was very inefficient and many products lacked both flow cell attachment sites (P5/P7) and therefore could not undergo bridge amplification. If you are using this approach, if you run the PCR enriched product on the Bioanalyzer (before second gel extraction) you will likely see two peaks separated by 30 bp. |
![]() |
![]() |
![]() |
#9 |
Member
Location: Oregon Join Date: Jan 2011
Posts: 24
|
![]()
Thanks very much for your suggestion. Although we don't use the 3 primer-based barcoding method, it may be informative to run pre gel extracted product on the bioanalyzer.
We ligate barcoded Y-adapters to cDNA. We run the "smear" on electrophoresis gel and extract at 350bp. Then, we PCR enrich with PE 1.0 and PE 2.0. I have always assumed that getting lots of product after PCR enrichment meant the adapters had annealed. Is there something else that could be happening? |
![]() |
![]() |
![]() |
#10 |
Member
Location: Little Rock AR Join Date: Jul 2010
Posts: 12
|
![]()
I realize this thread is a little old, but we are having trouble with getting our cluster densities up. A little background. We have made non-multiplexed RNA-seq libraries using PE primer 1.0 and 2.0 (synthesized from IDT) and get very high clusters/mm2.
However, with indexed libraries (modified Illumina's design using two primers, where the PCR 2.0 primer and the indexing primer are synthesized as one single primer) the clusters/mm2 are around (250 -300 K). We use qPCR to quantitate the libraries (after gel purification) and use 9 pM library DNA for cluster on the c-Bot. Any hints on how to improve the cluster densities and explain the discrepancy between indexed and non-indexed libraries? |
![]() |
![]() |
![]() |
#11 |
Member
Location: Oregon Join Date: Jan 2011
Posts: 24
|
![]()
Hello kshankar, I too would like to know why indexed libraries may produce lower cluster densities. How many samples per lane are you indexing?
Also, I still believe there are some contaminants from my gel extraction that are inhibiting the sequencing reaction. Can you tell me what gel percentage, brand, buffer and which kit you are using for your final gel extraction? Thanks so much. |
![]() |
![]() |
![]() |
#12 | |
Member
Location: Houston, TX Join Date: Jul 2010
Posts: 14
|
![]() Quote:
I know that SCS1.8 identifies clusters based on the result of the first 5 or 6 cycles. Poorly designed indexes may lead to SCS to treat clusters close to each other as one cluster initially but eventually fail those clusters due to poor phasing/prephasing, hence leading to low cluster density. |
|
![]() |
![]() |
![]() |
Tags |
cluster density, rna seq |
Thread Tools | |
|
|