Default NexteraXT Bionalayzer - undertagmentation?

Hi,

I am using the NexteraXT kit to prep mtDNA. I just did a test prep of some samples and had them run on a HS DNA Chip on the Bioanalyzer (see attachment).

I would like a second opinion on these. Does the libraries look undertagmentated?

(see past sample B, I screwed that one up...)
A: Following the procedure recommendations (1ng)
C: Half of the recommendations (0.5ng)
D: Quarter of the recommendations (0.25ng)

Also, quantifying the libraries by Qubit revealed them to be on the low end, just under 2nM with standard 12cycles PCR. Should I be concerned that the output of the prep is so low?

Thanks!
Attached Files
File Type: pdf Bioanalyzer.pdf (1.15 MB, 100 views)

Last edited by Meyana; 11-01-2017 at 12:51 AM. Reason: Uploaded wrong file...
Meyana is offline   Reply With Quote
Old 11-07-2017, 04:10 PM   #2
Meyana
Member
 
Location: Japan

Join Date: Sep 2017
Posts: 20
Default

Would still appreciate any input
Meyana is offline   Reply With Quote
Old 11-07-2017, 05:19 PM   #3
nucacidhunter
Senior Member
 
Location: Melbourne

Join Date: Jan 2013
Posts: 1,166
Default

Would you elaborate on the experiment aim because interpretation would depend on the final aim.

Edit: Are these post PCR libraries cleaned with beads?

Last edited by nucacidhunter; 11-07-2017 at 05:23 PM.
nucacidhunter is offline   Reply With Quote
Old 11-07-2017, 06:00 PM   #4
Meyana
Member
 
Location: Japan

Join Date: Sep 2017
Posts: 20
Default

Quote:
Originally Posted by nucacidhunter View Post
Would you elaborate on the experiment aim because interpretation would depend on the final aim.

Edit: Are these post PCR libraries cleaned with beads?
The aim of the experiment is to identify mtDNA mutations (heteroplasmy). We willbe doing paired-end seq on the MiSeq. We will do around 5000X coverage to enable this.
Before prep I did an "enzymatic purification" to remove nDNA and only retain mtDNA from gDNA purified from mouse tissue.

These are post-PCR, AMPure purified libraries. Eluted in volume recommended by the protocol (and then 1/2 and 1/4 for samples C and D respectively, also using 1/2 and 1/4 beads for cleanup).
Meyana is offline   Reply With Quote
Old 11-07-2017, 11:21 PM   #5
nucacidhunter
Senior Member
 
Location: Melbourne

Join Date: Jan 2013
Posts: 1,166
Default

I think that A is better because the other two show signs of low diversity library. For optimum read number you would need to quantify the library with qPCR and take the average size of area under 100-950 bp when calculating molar concentration.

You can change bead ratio to modify the size cut if you want to merge R1 and R2. If you do 2x300 cycles on current library A you probably will get around 30% merge.
nucacidhunter is offline   Reply With Quote
Old 11-08-2017, 03:41 PM   #6
Meyana
Member
 
Location: Japan

Join Date: Sep 2017
Posts: 20
Default

Could you please elaborate on how you determine which of the libraries have lower diversity? Is it due to the "spiky" pattern?
(I am new to prep of samples, previously I just passed them on to another person in the lab, but I would like to know how to best assess the libraries).

Also, inherently my samples are very low complexity. Based on some quality checks of my protocol, I will estimate I have >95% mtDNA (compared to nDNA) in my sample before going into library prep. The mtDNA is only ~16kb. So no matter what, wouldn't this be considered a low complexity sample? And then it would make sense if the libraries show low complexity?

I am doing paired-end because I will be looking for mtDNA deletions which are very frequent in my model mouse system and may be important in disease symptom development. So as such, I will not merge R1 and R2.
I was thinking to use the v2 2x150bp kit.
Meyana is offline   Reply With Quote
Old 11-08-2017, 05:32 PM   #7
nucacidhunter
Senior Member
 
Location: Melbourne

Join Date: Jan 2013
Posts: 1,166
Default

Spiky pattern is good indicator of ow diversity.

Your samples if tagmented randomly would produce around 15,000x15,000 fragments so it is not low diversity. PhiX is only 5.4 kb.

This library would be fine for sequencing.
nucacidhunter is offline   Reply With Quote
Old 11-08-2017, 07:37 PM   #8
GA-J
Member
 
Location: USA

Join Date: Jul 2015
Posts: 28
Default

A and C looks good to me, D looks over-fragmented due to low starting mass.

B ? The size looks too large.

But I think you will have good data yield for A,C and D, may receive less data from B.
GA-J is offline   Reply With Quote
Old 11-08-2017, 09:19 PM   #9
Meyana
Member
 
Location: Japan

Join Date: Sep 2017
Posts: 20
Default

Thanks for the input from both of you!

nucacidhunter
For some of my samples I will not be able to input more than 500ng (at the max, working with very small brain tissue pieces) to the prep, which is why I was attempting to scale down the reaction. Do you think my output will suffer tremendously if I am to sequence something that looks like C or D?

GA-J
What makes you say D is over-fragmentated? When I compare to the NexteraXT manual and look at the examples provided by Illumima I would say all my libraries are a bit large...
Meyana is offline   Reply With Quote
Old 11-09-2017, 12:53 AM   #10
nucacidhunter
Senior Member
 
Location: Melbourne

Join Date: Jan 2013
Posts: 1,166
Default

Quote:
Originally Posted by Meyana View Post
Thanks for the input from both of you!

nucacidhunter
For some of my samples I will not be able to input more than 500ng (at the max, working with very small brain tissue pieces) to the prep, which is why I was attempting to scale down the reaction. Do you think my output will suffer tremendously if I am to sequence something that looks like C or D?.
I guess you mean 500 pg and that should be fine. With your 16kb target and 5000x coverage you require 80Mb sequencing per sample. The required read number (2x150) would be ~267k plus some reads to cover off target regions.

D might be pushing it and you might get more duplicates but should not affect final results.
nucacidhunter is offline   Reply With Quote
Old 12-07-2017, 05:46 PM   #11
Meyana
Member
 
Location: Japan

Join Date: Sep 2017
Posts: 20
Default

Thank you for all your kind input! The libraries are coming along very well. Will get BioAnalyzer results from the last ones today (had to redo a few)

Now comes the next part, pooling the libraries.

My concentrations varies from 1nM to 5.6nM.
According to the protocol, I should pool equimolar concentrations to achieve a 2nM final pooled library.

However, since some of my samples below this concentration, how should I deal with this?

The protocol recommends using 5uL 2nM library with 5uL 0.2N NaOH. Could I instead use 10uL 1nM library with 1uL 1N NaOH (and disregard the 1 extra uL, since everything will be diluted with the addition of 990uL HT1).

Thanks!
Meyana is offline   Reply With Quote
Old 12-07-2017, 06:14 PM   #12
GA-J
Member
 
Location: USA

Join Date: Jul 2015
Posts: 28
Default

You may pool equal mol for each sample, then find the final pooled lib con. Meanwhile, if you receive 0.5nM, look at the protocol, you could do as below: Starting Library Concentration Library 0.2 N NaOH
4 nM 5 l 5 l
2 nM 10 l 10 l
1 nM 20 l 20 l
0.5 nM 40 l 40

The protocol: NextSeq System
Denature and Dilute Libraries Guide Document # 15048776 v02
GA-J is offline   Reply With Quote
Old 12-07-2017, 06:56 PM   #13
Meyana
Member
 
Location: Japan

Join Date: Sep 2017
Posts: 20
Default

Thanks! - I have only looked at the MiSeq protocols (since I will be using the MiSeq), but I guess this is also applicable for the MiSeq?
Meyana is offline   Reply With Quote
Old 12-07-2017, 07:24 PM   #14
GA-J
Member
 
Location: USA

Join Date: Jul 2015
Posts: 28
Default

For Miseq, whatever you started, you may adjust to load 10pM. I have done this a few times, it worked well. You don't have to spike more than 5%, I used 1% spike.
GA-J is offline   Reply With Quote
Old 12-07-2017, 07:41 PM   #15
Meyana
Member
 
Location: Japan

Join Date: Sep 2017
Posts: 20
Default

Great to know. I was just about to ask how to choose between 6, 8, and 10pM, I haven't seen any info or guidelines on how you know what you should do. Could you add any further information for my future runs? (this is mostly a method test)

I was thinking to do a 5-10% PhiX spike. Some of my libraries show low diversity, as adviced above in this thread (the 'spiky' pattern on the BioAnalyzer, only for a few of the samples).
Meyana is offline   Reply With Quote
Old 12-14-2017, 07:23 PM   #16
Meyana
Member
 
Location: Japan

Join Date: Sep 2017
Posts: 20
Default

I just finished the run of my first 24 samples on the MiSeq. Thank you all for your great input!

Unfortunately, it was really under clustered, with only 305K/mm2. 88.2% of clusters passed filtering. Hopefully I will still be able to get some info from the samples.

I'm thinking I overestimated the concentration of my libraries (I use Qubit, I aimed to load 10pM) and this is causing the issue. Any other ideas or things I should consider are highly appreciated

Last edited by Meyana; 12-14-2017 at 07:26 PM.
Meyana is offline   Reply With Quote
Reply

Tags
bioanalyzer, library concentration, nextera xt, undertagmentation

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 08:14 PM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2018, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO