#1 |
Senior Member
Location: Purdue University, West Lafayette, Indiana Join Date: Aug 2008
Posts: 2,317
|
![]()
Top panel, normal RNA nano ladder run on a single-strand assay Bioanalzyer chip (RNA nano). Sizes in nucleotides.
Bottom panel, same RNA nano molecular weight ladder run on a double-strand assay Bioanalyzer chip (High Sensitivity DNA). Sizes (incorrect) in base pairs. ![]() The first 3 fragments run at approximately double their true molecular weight. That is 200, 500 and 1000 nucleotides run as 236, 428 and 902 base pairs. Then 2000 and 4000 nucleotide fragments run as 10 and 27 thousand bp. Well, that is my interpretation. There could actually be some utility in this phenomenon for the lower MW molecules. If you have a mixture of ds and ss molecules and want to know their strand lengths, then you would get approximately the right answers. But if your are expecting them to run at their true MW then you might be confused by your results. For the converse phenomenon on (double stranded DNA on an RNA chip runs faster than expected) please see my earlier thread: http://seqanswers.com/forums/showthread.php?t=12501 -- Phillip Last edited by pmiguel; 07-20-2011 at 06:55 AM. |
![]() |
![]() |
![]() |
#2 |
Member
Location: ILLINOIS Join Date: Jan 2011
Posts: 30
|
![]()
How about running ssDNA on a ds DNA chip? Is it gonna work fine?
|
![]() |
![]() |
![]() |
#3 |
Senior Member
Location: Bethesda MD Join Date: Oct 2009
Posts: 509
|
![]()
I noticed that the fluorescence units for your DNA chip graph are pretty high. We observe the same phenomenon you observe when the chip is overloaded, even with dsDNA. You might want to dilute the sample to the appropriate concentration (<10 ng/ul) and retest.
-Harold |
![]() |
![]() |
![]() |
#4 | |
Senior Member
Location: Purdue University, West Lafayette, Indiana Join Date: Aug 2008
Posts: 2,317
|
![]() Quote:
http://seqanswers.com/forums/showthread.php?t=12501 -- Phillip |
|
![]() |
![]() |
![]() |
#5 | |
Senior Member
Location: Purdue University, West Lafayette, Indiana Join Date: Aug 2008
Posts: 2,317
|
![]() Quote:
Did you take a look at the converse result I saw when running dsDNA on an RNA chip? http://seqanswers.com/forums/showthread.php?t=12501 -- Phillip |
|
![]() |
![]() |
![]() |
#6 |
Member
Location: Seattle Join Date: Mar 2008
Posts: 29
|
![]()
may check with Agilent about effects of sample's salt/tween that change the migration
|
![]() |
![]() |
![]() |
#7 |
Senior Member
Location: Purdue University, West Lafayette, Indiana Join Date: Aug 2008
Posts: 2,317
|
![]()
Hi vasvale,
Have you ever tried communicating with Agilent? Personally, I have not found it to be productive. In any case, since we are loading the ladder from another chip type, it seems unlikely there is a difference in the salt/tween concentration between it and the double stranded ladder. -- Phillip |
![]() |
![]() |
![]() |
#8 |
Member
Location: USA Join Date: Jul 2010
Posts: 58
|
![]()
Hi pmiguel.
I love this work! you are so... I've never done this before but I used a dye of 1000DNA chip for HS chip and worked well. |
![]() |
![]() |
![]() |
#9 |
Member
Location: Madrid Join Date: Dec 2011
Posts: 13
|
![]() |
![]() |
![]() |
![]() |
#10 |
Senior Member
Location: Purdue University, West Lafayette, Indiana Join Date: Aug 2008
Posts: 2,317
|
![]() |
![]() |
![]() |
![]() |
#11 | |
Senior Member
Location: Purdue University, West Lafayette, Indiana Join Date: Aug 2008
Posts: 2,317
|
![]() Quote:
-- Phillip |
|
![]() |
![]() |
![]() |
#12 | |
Member
Location: Madrid Join Date: Dec 2011
Posts: 13
|
![]() Quote:
|
|
![]() |
![]() |
![]() |
#13 | |
Member
Location: Madrid Join Date: Dec 2011
Posts: 13
|
![]() Quote:
|
|
![]() |
![]() |
![]() |
#14 | |
Senior Member
Location: Purdue University, West Lafayette, Indiana Join Date: Aug 2008
Posts: 2,317
|
![]() Quote:
The lowest peak is not part of the ladder -- it is spiked in separately and is a different size for the nano (25 nt) and the DNA HS (35 nt). If you look at the relative areas of all the other major peaks, they are reasonable. Also looking at time, instead of MW would not necessarily help. The Agilent software actually alters the time to align the lanes. If you change the position of the upper or lower marker, you will find its reported "migration time" will change as well. Thanks for your comments about the mass of the MW weight standards for the pico and nano chip. If true, then the chip was overloaded. I should run another one, but I am not sure when or if I will get around to it. -- Phillip |
|
![]() |
![]() |
![]() |
#15 |
Junior Member
Location: SF bay area Join Date: Dec 2007
Posts: 6
|
![]()
Hi Phillip,
There are a few differences between the different bioanalyzer kits, esp. the RNA vs. DNA. First, as you'd expect they use different dyes, and some dyes bind dsDNA better, others prefer ssDNA or RNA. Second, the RNA chip formulations can be somewhat denaturing; they will remove some but not all secondary structure. Often RNA or ssDNA runs more slowly on a DNA chip due to residual secondary structure, and dsDNA runs faster on an RNA chip because it is rod-like. Finally, the different protocols (esp. the HS protocols) may actually use different voltages and buffers. I think the targets can be "preconcentrated" by varying voltage in a low-ionic-strength buffer. This is one reason the sample should not have much salt in it. Finally, two tips in case things are not running the way they used to: first, remove and scrub the pin electrode assembly under di-water, dry it and replace. The recommended rinse does not remove salt deposits as well as a good brushing under running water. Second, watch the expiry dates on the reagents, esp. the chips. You might expect that glass chips would last forever but there is plastic outgassing so it is not so simple. Full disclosure: I actually work in R&D at Agilent, though not in groups related to sequencing or the bioanalyzer. But I've used the Bioanalyzer a lot, often for off-label functions. I've had good luck running mixtures of dsDNA and oligos on the small RNA chip, though I had to run parallel standards of known size and structure. |
![]() |
![]() |
![]() |