    Low input/adaptor dimers

    Hi all,

    I'm preparing libraries from very small amounts of RNA, around 100-1000x less than the 1ug recommended by illumina. Generating a "library" seems easy but on analysis it turns out to be mostly adaptor. Has anyone experimented with adding less adaptor? Or any other ways of optimizing library preparation for small amounts of nucleic acids?

    I get the impression this is a fairly common problem, I'll be on the phone to illumina in a bit and I'll pass along any gems they give me. In the meantime what do all you seqanswerers think?

    WW
  • arolfe
    Member
    • Jul 2011
    • 29

    #2
    I believe that different Illumina kits used different adapter concentrations. For example, I think the only difference between the ChIP-Seq and genomic (?) kits was a 100X difference in adapter concentration. So yes, it makes sense and you should be able to look around to get an idea of where to start.

    Comment

    • HESmith
      Senior Member
      • Oct 2009
      • 512

      #3
      The ratio of adaptor to sample is probably the most critical variable in library construction. You should be able to adapt the ChIP-Seq protocol, which works with 10ng dsDNA. You may need to perform an additional round of size selection to remove adaptor dimers. Alternatively, you can perform linear amplification of your starting material.

      Harold

      Comment

      • whw
        Member
        • Jun 2011
        • 19

        #4
        I have the Ultra-low input mRNA Seq Guide from Illumina which is optimized for 2ng of dsDNA. I'm thinking I'll try their adaptor dilutions and see what happens, hopefully the ligation chemistry won't be too different.

        Comment

        • griffes
          Junior Member
          • Oct 2011
          • 4

          #5
          what adapter dilutions are you using for RNA?

          Comment

          • ETHANol
            Senior Member
            • Feb 2010
            • 308

            #6
            Just in case you were wondering from information I've gathered from other threads the concentrations of the TruSeq adapters are:
            DNA Sample prep Kit adapters = 15 μM
            RNA Sample Prep Kit adapters = 0.25 μM

            And yes, adjusting the amount of adapter to the amount of DNA you want to ligate it to is good practice. The adapter should be in excess but I'd say not more then 10 fold and probably better to be less.
            --------------
            Ethan

            Comment

            • whw
              Member
              • Jun 2011
              • 19

              #7
              Originally posted by griffes View Post
              what adapter dilutions are you using for RNA
              The ultra-low input protocol suggests a dilution of 1:19 for 1-4ng of sample. I thought I'd start in that range.

              Originally posted by ETHANol View Post
              Just in case you were wondering from information I've gathered from other threads the concentrations of the TruSeq adapters are:
              DNA Sample prep Kit adapters = 15 μM
              RNA Sample Prep Kit adapters = 0.25 μM

              And yes, adjusting the amount of adapter to the amount of DNA you want to ligate it to is good practice. The adapter should be in excess but I'd say not more then 10 fold and probably better to be less.
              That's very helpful, thanks for the tip Ethan(ol)

              Comment

              Latest Articles

              Collapse

              • SEQadmin2
                Proteomic Platforms: How to Choose the Right Analytical Strategy to Improve Detection and Clinical Applications
                by SEQadmin2


                Proteomics platforms are evolving rapidly, with advances in mass spectrometry and affinity-based approaches expanding what researchers can detect and at what scale. As the field moves toward deeper proteome coverage and clinical applications, scientists face an increasingly complex landscape of tools. This article will explore how researchers are navigating these choices to find the right platform for their work.

                The systematic characterization of the human proteome has
                ...
                07-20-2026, 11:48 AM
              • SEQadmin2
                Advanced Sequencing Platforms Tackle Neuroscience’s Toughest Genomics Problems
                by SEQadmin2



                Genomics studies in neuroscience face a special challenge due to the brain’s complexity and scarcity of samples. Mapping changes in cell type and state using conventional next-generation sequencing methods remains challenging. Advances in technologies like single-cell sequencing, spatial transcriptomics, and long-read sequencing have opened the door to deeper studies of the brain and diseases like Alzheimer’s, amyotrophic lateral sclerosis (ALS), and schizophrenia.
                ...
                07-09-2026, 11:10 AM
              • SEQadmin2
                Cancer Drug Resistance: The Lingering Barrier to Rising Survival
                by SEQadmin2



                Cancer survival rates have significantly increased in the last few decades in the United States, reaching a combined 70% 5-year survival rate by 2021. Behind this number, there are years of research to find new therapies, drug targets, and early detection methods. But there is one core challenge that keeps slowing down these advances, and it’s about drug resistance.

                There is no single reason why many patients don’t respond to treatment as expected. Cancer is...
                07-08-2026, 05:17 AM

              ad_right_rmr

              Collapse

              News

              Collapse

              Topics Statistics Last Post
              Started by SEQadmin2, Today, 12:17 PM
              0 responses
              10 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, Yesterday, 11:41 AM
              0 responses
              11 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 07-20-2026, 11:10 AM
              0 responses
              23 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 07-13-2026, 10:26 AM
              0 responses
              37 views
              0 reactions
              Last Post SEQadmin2  
              Working...