    custom sequencing primers

    Hi,

    Does anyone use their own primers for sequencing on the GAII, i.e rather than the Illumina supplied primer? If so, what purification/modifications do you use? I'm wondering if, for example, any phosphorothioate bonds would be necessary... any thoughts appreciated!

    Thanks.
  • ScottC
    Senior Member
    • Jan 2008
    • 244

    #2
    I don't think the sequencing primers are modified. The sequences are freely available from Illumina if you ask (they may even be posted here somewhere).

    Cheers,

    Scott.

    Comment

    • james hadfield
      Moderator
      Cambridge, UK
      Community Forum
      • Feb 2008
      • 224

      #3
      The sequencing primers are indeed unmodified. We order PAGE purified although I am not sure there is a requirement for this. If you want to know the sequences I would suggest contacting Illumina techsupport directly and asking for the latest release of their "sequence information for customers" pdf.
      This way there should be no confusion.
      James.

      Comment

      • ovallon
        Junior Member
        • Feb 2011
        • 1

        #4
        a sequencing primer OTHER than the Illumina primer

        Hi !
        My first post here
        Originally posted by athos View Post
        Hi,

        Does anyone use their own primers for sequencing on the GAII, i.e rather than the Illumina supplied primer?
        Possibly the same question that you had in mind, but phrased differently.

        What I would like to know is whether it is possible, when you are doing amplicon sequencing, to use a primer WITH A SEQUENCE DIFFERENT from that used by Illumina.
        Let me explain. Imagine you have amplified a region, say the flanking sequences of a transposon, and made an Illumina library that has the Illumina primers at both ends. But you don't want to sequence the region immediately downstream of the primer, because you already know what it is. You want to start sequencing WITHIN the insert.
        Can you prime sequencing with an oligonucleotide that anneals to a common part of all your inserts, downstream of the adapter ?

        If yes, must you use the same sequencing primer for the whole chip, or can you use different sequencing primers in different lanes ?

        Final question: is it true that identification of the clusters is hampered when the first nucleotides of all (or most of) the sequences are identical ? This would be the case for example if you were sequencing through polyA-tails, or through a region that is common to all your inserts. I have heard that it is wise in this case to add 2 Ns right after the sequencing primer to make sure that sequences differ locally, allowing pixels to be grouped into clusters. True ?

        Comment

        • HESmith
          Senior Member
          • Oct 2009
          • 512

          #5
          1. You can use custom primers for sequencing. The Tm should be similar to the standard primer.

          2. You can use different primers in different lanes of the flow cell.

          3. Biased nucleotide composition in the first few cycles (I believe this number varies with different software versions) adversely affects cluster calling.

          -Harold

          Comment

          • fkrueger
            Senior Member
            • Sep 2009
            • 627

            #6
            It is indeed true that low-diversity at the start of sequences will cause problems with cluster identification. See this thread.

            Comment

            Latest Articles

            Collapse

            • GATTACAT
              Reply to Nine Things a Sample Prep Scientist Thinks About Before Sequencing
              by GATTACAT
              Love this - good data definitely starts from good input, and poor input can only give relatively poor data. I particularly like the mention of Nanodrop/absorbance based methods for quantification. It's such a toss up if you'll get an accurate reading or what amounts to a randomly generated number, and a lot of library/sequencing related issues can be traced back to poor quant.
              07-01-2026, 11:43 AM
            • SEQadmin2
              Nine Things a Sample Prep Scientist Thinks About Before Sequencing
              by SEQadmin2


              I’m not a sequencing expert. I’m a purification scientist who uses NGS to evaluate workflows my group develops. With this perspective, we think about the sample first and the NGS workflow second. The sequencer is an exceptionally honest reporter, but it can only report on what you give it, so whether you get clean, interpretable data from an NGS workflow is largely determined before you begin.

              Here are nine questions we think about, in roughly the order they matter, before...
              06-18-2026, 07:11 AM

            ad_right_rmr

            Collapse

            News

            Collapse

            Topics Statistics Last Post
            Started by SEQadmin2, 07-02-2026, 11:08 AM
            0 responses
            16 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 06-30-2026, 05:37 AM
            0 responses
            17 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 06-26-2026, 11:10 AM
            0 responses
            21 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 06-17-2026, 06:09 AM
            0 responses
            54 views
            0 reactions
            Last Post SEQadmin2  
            Working...