    Single or Dual indices?

    Hi all,
    I am new to Illumina and am preparing a library of 24 samples with the Nextera XT. I plan on runing these samples on the MiSeq. I've read a couple threads where people mention single or dual indices such as http://seqanswers.com/forums/showthread.php?t=14676.
    Why you would do one over the other? Does it make a difference?
    Thanks!
  • kcchan
    Senior Member
    • Jul 2012
    • 186

    #2
    Nextera XT uses a dual index strategy which allows 24 unique barcodes using only 10 oligos (4 of Index 1 and 6 of Index 2) instead of 24 unique oligos for each barcode. You can mix and match dual and single indexed libraries in a sequencing run, but not for the sample prep. For Nextera you must use both index oligos for your sample preparation.

    Comment

    • pmiguel
      Senior Member
      • Aug 2008
      • 2328

      #3
      Just to be clear, if you use a Nextera kit, you get dual-indexed amplicons. You don't have to sequence both indexes (or any of them), but they are there, none the less.

      Since you are doing a MiSeq run, I would suggest running dual indexing since that is built into the MiSeq reagents, anyway. But for 24 indexes, you would have to deploy duals anyway, because the standard Nextera kit has a maximum of 12 single indexes. Of course you could "go rogue" and synthesize your own step-out primers to get you up to 24 single indexes. But the cheap Nextera adapter kit offers 24 index pairs, so why look further?

      --
      Phillip

      Comment

      • farrel75
        Member
        • May 2012
        • 11

        #4
        Thanks everyone! It's good to know that its recommended to use dual indexes for the MiSeq and Nextera XT. I guess I didn't make it clear that I was wondering why in the general sense would you do single vs. dual indexing. What are the benefits/problems of one index vs. two?

        Comment

        • pmiguel
          Senior Member
          • Aug 2008
          • 2328

          #5
          Dual costs you an extra 8 cycles + some dark cycles. But MiSeq reagent kits are rated for how many sequence (insert) sequence bases you read, with enough extra so that you can do all the extra cycles required for dual indexing.

          So, the only reasons I can think of to choose single index, would be if you really wanted your reads to be a few bases longer or you could not afford the extra hour or two required to run the second index.

          --
          Phillip

          Comment

          • austinso
            Member
            • Jun 2012
            • 77

            #6
            Off-hand does anyone know if you can specify reading only the second index "i5" and not the first "i7"?

            Comment

            • kcchan
              Senior Member
              • Jul 2012
              • 186

              #7
              Originally posted by austinso View Post
              Off-hand does anyone know if you can specify reading only the second index "i5" and not the first "i7"?
              It might take some hacking of the recipes to get that to work. A simpler workaround would be to use the P5 primer as the index 1 primer and then skip the "index2" read. You will have to burn an extra 7 cycles to get through the padding sequences before you start reading the barcodes, however.

              Comment

              • austinso
                Member
                • Jun 2012
                • 77

                #8
                Thanks for confirming that, kcchan.

                Comment

                • HeinKey
                  Member
                  • May 2009
                  • 21

                  #9
                  Hi Austinso,
                  If you use low complexity pooling strategies be aware to have sufficient balance in each index cycle.
                  Choose the indices properly, making sure the combination has 50% A/C and 50% G/T per cycle. Strong bias in variation is easy with only a few different indices. Basecalling can be very much off with biased indices.
                  The latest MCS version has a new algorithm and works much better with biased libraries though.
                  It's a miracle to me they don't implement this algorithm into the next HCS update also. "No that's not planned" was Illumina's reaction when I informt for this.

                  Comment

                  Latest Articles

                  Collapse

                  • SEQadmin2
                    Advanced Sequencing Platforms Tackle Neuroscience’s Toughest Genomics Problems
                    by SEQadmin2



                    Genomics studies in neuroscience face a special challenge due to the brain’s complexity and scarcity of samples. Mapping changes in cell type and state using conventional next-generation sequencing methods remains challenging. Advances in technologies like single-cell sequencing, spatial transcriptomics, and long-read sequencing have opened the door to deeper studies of the brain and diseases like Alzheimer’s, amyotrophic lateral sclerosis (ALS), and schizophrenia.
                    ...
                    07-09-2026, 11:10 AM
                  • SEQadmin2
                    Cancer Drug Resistance: The Lingering Barrier to Rising Survival
                    by SEQadmin2



                    Cancer survival rates have significantly increased in the last few decades in the United States, reaching a combined 70% 5-year survival rate by 2021. Behind this number, there are years of research to find new therapies, drug targets, and early detection methods. But there is one core challenge that keeps slowing down these advances, and it’s about drug resistance.

                    There is no single reason why many patients don’t respond to treatment as expected. Cancer is...
                    07-08-2026, 05:17 AM
                  • GATTACAT
                    Reply to Nine Things a Sample Prep Scientist Thinks About Before Sequencing
                    by GATTACAT
                    Love this - good data definitely starts from good input, and poor input can only give relatively poor data. I particularly like the mention of Nanodrop/absorbance based methods for quantification. It's such a toss up if you'll get an accurate reading or what amounts to a randomly generated number, and a lot of library/sequencing related issues can be traced back to poor quant.
                    07-01-2026, 11:43 AM

                  ad_right_rmr

                  Collapse

                  News

                  Collapse

                  Topics Statistics Last Post
                  Started by SEQadmin2, Yesterday, 10:26 AM
                  0 responses
                  13 views
                  0 reactions
                  Last Post SEQadmin2  
                  Started by SEQadmin2, 07-09-2026, 10:04 AM
                  0 responses
                  26 views
                  0 reactions
                  Last Post SEQadmin2  
                  Started by SEQadmin2, 07-08-2026, 10:08 AM
                  0 responses
                  16 views
                  0 reactions
                  Last Post SEQadmin2  
                  Started by SEQadmin2, 07-07-2026, 11:05 AM
                  0 responses
                  33 views
                  0 reactions
                  Last Post SEQadmin2  
                  Working...