    DNA 1000 Ladder Slow Migration - Agilent Bioanalyzer

    Hi all,

    I'm having a consistent problem with the bioanalyzer not reading the last peak or peaks for the DNA 1000 ladder. All of the other assays seem to read OK.

    I've tried:
    1) Replacing the primer parts
    2) increasing the drying time after cleaning the pins
    3) Replacing all of the reagents

    The problem seems to get worse throughout the day. Any suggestions are appreciated!
  • idedios
    Member
    • Mar 2014
    • 20

    #2
    Some Agilent reps told me to clean the pins with a brush after heavy use (like a toothbrush).

    Comment

    • ScottC
      Senior Member
      • Jan 2008
      • 244

      #3
      Some things that we've experienced with slow migration: Does the priming syringe return to the recommended level after you release the clamp? Are all your reagents within their expiry date? Is the gel-dye mix fresh? Have you centrifuged the gel-dye mix hard before use?

      Also, I'm not sure whether this would solve your problem, but on the topic of washing Bioanalyzer pins:

      On the odd occasion that we have some really bad pin contamination or if someone has forgotten about a chip and left it to dry in the machine, we clean our Bioanalyzer pins by soaking them in detergent and scrubbing very thoroughly with a toothbrush (holding the pin array at an angle and scrubbing across the side of each row of pins). We then rinse very thoroughly in distilled water and dry them for a couple of hours in our 50oC incubator. Maybe it's a bit harsh compared to the recommendations, but it works and it doesn't do any harm.

      Comment

      • jlei_face
        Member
        • Oct 2014
        • 10

        #4
        Sorry for the late post but I experienced this problem before also. For my lab, it turns out the ambient temperature was too cold. We ended up creating an enclosure for the BA which kept the air surrounding it much warmer and we encountered this issue much less.

        Comment

        • AStretton
          Member
          • Jul 2012
          • 25

          #5
          Hi jlei_face
          That's quite interesting what you were saying about the ambient temperature. We a have similar problem quite often with the High sensitivity ladder. What temperature did you find was optimum?
          Thanks

          Comment

          • jlei_face
            Member
            • Oct 2014
            • 10

            #6
            AStretton,
            I've been out of that lab for some time but I believe our optimum temperature was right at RT (25degrees C) or slightly above. The lab that contained the BA also contained other instruments that needed to be kept cool so the AC was usually on full blast.

            Comment

            • FiReaNG3L
              Member
              • Apr 2012
              • 17

              #7
              We had this problem a few years ago - make sure that your priming station is fine (the little white seal is unbroken, the syringe is new and set at the right position).

              For us the problem was that the little hole underneath the syringe was clogged - cleaning it solved everything. Look through in while pointing at a bright light to make sure the path is clear.

              Comment

              Latest Articles

              Collapse

              • SEQadmin2
                Beyond CRISPR/Cas9: Understand, Choose, and Use the Right Genome Editing Tool
                by SEQadmin2



                CRISPR/Cas9 sparked the gene editing revolution for both research and therapeutics.1 But this system still showed severe issues that limited its applications. The most prominent were the heavy reliance on PAM sequences, delivery limitations, double-stranded breaks that prompt unintended edits and cell death, and editing inefficiency (both in targeting and in knock-in reliability).

                Despite this, “CRISPR helped turn genome editing from a specialized technique into
                ...
                07-31-2026, 11:01 AM
              • SEQadmin2
                Proteomic Platforms: How to Choose the Right Analytical Strategy to Improve Detection and Clinical Applications
                by SEQadmin2


                Proteomics platforms are evolving rapidly, with advances in mass spectrometry and affinity-based approaches expanding what researchers can detect and at what scale. As the field moves toward deeper proteome coverage and clinical applications, scientists face an increasingly complex landscape of tools. This article will explore how researchers are navigating these choices to find the right platform for their work.

                The systematic characterization of the human proteome has
                ...
                07-20-2026, 11:48 AM

              ad_right_rmr

              Collapse

              News

              Collapse

              Topics Statistics Last Post
              Started by SEQadmin2, Yesterday, 12:22 PM
              0 responses
              13 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 08-11-2026, 10:35 AM
              0 responses
              14 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 08-06-2026, 07:41 AM
              0 responses
              31 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 08-03-2026, 10:13 AM
              0 responses
              50 views
              0 reactions
              Last Post SEQadmin2  
              Working...