    Hello,
    I followed the protocol for the 550bp size TruSeq nano library preparation but my final libraries were much larger than the expected 670bp average size (see attached pic).

    Has anyone else come across larger than expected libraries?
    Do you have any advice as to what may have happened?

    Thanks in advance,
    Anna.
    Attached Files

  • #2
    I also saw a larger than expected product. I ran the Covaris to shear at 500 bp, and the final libraries had Bioanalyzer peaks at 780 bp. I don't have an explanation.

    Comment


    • #3
      Truseq pcr free libraries do run larger on a BA due to the tailed adapters but I have not seen this before with nano libraries. One explanation my be incomplete mixing of the size selection beads meaning that the incorrect amount of beads were added to the cleanup and a poor size selection. Might also be an idea to run the sheared samples on the BA before cleanup to check the size distribution from the shearing.

      Comment


      • #4
        Hello, I asked Illumina tech support and they said to try diluting the libraries as the chip looked to be overloaded. I made a 1 in 20 dilution and repeated the chip. An example of a trace is attached. The average size has gone down by around 100bp but it's still not the 670bp average stated in the manual. I have sent tech support these diluted library results and asked if they think they will sequence.

        Anna.
        Attached Files

        Comment


        • #5
          I have the same problem. I make a change on the fragmentation protocol whit covaris (from a 550 pb insert to a 400 pb insert) but the mean size was almost the same (900 bp aprox.). I think that could be a problem on the size selection. Any help?. Thanks.

          Comment


          • #6
            Hello,
            It's a size selection issue. The protocol doesn't need to be altered, just make sure your SP beads are vortexed a lot (I mean a lot!). I usually mix them in between each sample prep as a pre-caution. Also we've found that the libraries run larger on the tapestation in comparison to the bioanalyser. Libraries usually look around 50bp larger on the tapestation. When the insert sizes are checked in the analysis stage, they are usually around the right size.

            All the best,
            Anna.

            Comment

            Latest Articles

            Collapse

            • seqadmin
              Advancing Precision Medicine for Rare Diseases in Children
              by seqadmin




              Many organizations study rare diseases, but few have a mission as impactful as Rady Children’s Institute for Genomic Medicine (RCIGM). “We are all about changing outcomes for children,” explained Dr. Stephen Kingsmore, President and CEO of the group. The institute’s initial goal was to provide rapid diagnoses for critically ill children and shorten their diagnostic odyssey, a term used to describe the long and arduous process it takes patients to obtain an accurate...
              12-16-2024, 07:57 AM
            • seqadmin
              Recent Advances in Sequencing Technologies
              by seqadmin



              Innovations in next-generation sequencing technologies and techniques are driving more precise and comprehensive exploration of complex biological systems. Current advancements include improved accessibility for long-read sequencing and significant progress in single-cell and 3D genomics. This article explores some of the most impactful developments in the field over the past year.

              Long-Read Sequencing
              Long-read sequencing has seen remarkable advancements,...
              12-02-2024, 01:49 PM

            ad_right_rmr

            Collapse

            News

            Collapse

            Topics Statistics Last Post
            Started by seqadmin, 12-17-2024, 10:28 AM
            0 responses
            32 views
            0 likes
            Last Post seqadmin  
            Started by seqadmin, 12-13-2024, 08:24 AM
            0 responses
            48 views
            0 likes
            Last Post seqadmin  
            Started by seqadmin, 12-12-2024, 07:41 AM
            0 responses
            34 views
            0 likes
            Last Post seqadmin  
            Started by seqadmin, 12-11-2024, 07:45 AM
            0 responses
            46 views
            0 likes
            Last Post seqadmin  
            Working...
            X