  • Agilent Bioanalyzer Negative Peaks?!?

    Here is what I mean:

    Click image for larger version

Name:	NegativePeak.png
Views:	1
Size:	28.4 KB
ID:	309147

    This is some input DNA obtained by reverse-crosslinking sonicated chromatin submitted by someone interested in our doing library construction for them. It was run on an Agilent Bioanalyzer High Sensitivity Chip.

    Everything looks great except -- the peak produced has negative amplitude!
    The Agilent Bioanalyzer Trouble shooting manual suggests this could happen as a result of the sample containing "detergents" or "dyes".

    Anyone seen this and been able to figure out the cause?

    (Yes, I know I answered the post of someone asking nearly the same question 4 years ago. But having seen this a few time recently in our lab, I'm finding all the responses in that thread, including my own, unhelpful.)

    --
    Phillip
    Last edited by pmiguel; 10-19-2016, 10:39 AM.

  • #2
    We have seen this before (slightly less dramatic) also with ChIP DNA from customers. Perhaps there is a reagent in the Diagenode ChIP kits that can have this effect? The library prep and sequencing went well nevertheless.

    Comment


    • #3
      I have seen it when DNA is eluted with buffer containing 0.1% Tween 20 even though in the Chip it is further diluted. Running the same sample on Screentape HS5000 is OK.

      Comment


      • #4
        Did you check concentration on qubit? May be outside of upper range? Have seen similar, not so pronounced, but diluting helped in our case.

        Comment


        • #5
          Originally posted by RickC7 View Post
          Did you check concentration on qubit? May be outside of upper range? Have seen similar, not so pronounced, but diluting helped in our case.
          We re-ran the samples, loading 1/4th the amount. Same effect:
          Click image for larger version

Name:	NegativePeak2.png
Views:	1
Size:	27.2 KB
ID:	305173

          --
          Phillip

          Comment


          • #6
            Originally posted by nucacidhunter View Post
            I have seen it when DNA is eluted with buffer containing 0.1% Tween 20 even though in the Chip it is further diluted. Running the same sample on Screentape HS5000 is OK.
            Interesting!

            Thanks,
            --
            Phillip

            Comment


            • #7
              Originally posted by nucacidhunter View Post
              I have seen it when DNA is eluted with buffer containing 0.1% Tween 20 even though in the Chip it is further diluted. Running the same sample on Screentape HS5000 is OK.
              Alas, after checking with the people doing the sample prep, they don't use Tween. They use SDS. They also do a minelute clean-up afterwards.

              --
              Phillip

              Comment


              • #8
                Phillip,

                Have you checked this resource from Agilent - slide 12??



                Maybe the sample needs one more clean up??
                SeqDavis

                Comment


                • #9
                  This suggests multiple causes including detergents and possibly chaotropic salts remains from column purification. Probably columns need to be washed more times than manufacturer recommends.

                  Comment

                  Latest Articles

                  Collapse

                  • seqadmin
                    Advancing Precision Medicine for Rare Diseases in Children
                    by seqadmin




                    Many organizations study rare diseases, but few have a mission as impactful as Rady Children’s Institute for Genomic Medicine (RCIGM). “We are all about changing outcomes for children,” explained Dr. Stephen Kingsmore, President and CEO of the group. The institute’s initial goal was to provide rapid diagnoses for critically ill children and shorten their diagnostic odyssey, a term used to describe the long and arduous process it takes patients to obtain an accurate...
                    12-16-2024, 07:57 AM
                  • seqadmin
                    Recent Advances in Sequencing Technologies
                    by seqadmin



                    Innovations in next-generation sequencing technologies and techniques are driving more precise and comprehensive exploration of complex biological systems. Current advancements include improved accessibility for long-read sequencing and significant progress in single-cell and 3D genomics. This article explores some of the most impactful developments in the field over the past year.

                    Long-Read Sequencing
                    Long-read sequencing has seen remarkable advancements,...
                    12-02-2024, 01:49 PM

                  ad_right_rmr

                  Collapse

                  News

                  Collapse

                  Topics Statistics Last Post
                  Started by seqadmin, 12-17-2024, 10:28 AM
                  0 responses
                  33 views
                  0 likes
                  Last Post seqadmin  
                  Started by seqadmin, 12-13-2024, 08:24 AM
                  0 responses
                  49 views
                  0 likes
                  Last Post seqadmin  
                  Started by seqadmin, 12-12-2024, 07:41 AM
                  0 responses
                  34 views
                  0 likes
                  Last Post seqadmin  
                  Started by seqadmin, 12-11-2024, 07:45 AM
                  0 responses
                  46 views
                  0 likes
                  Last Post seqadmin  
                  Working...
                  X