Seqanswers Leaderboard Ad

Collapse

Announcement

Collapse
No announcement yet.
X
 
  • Filter
  • Time
  • Show
Clear All
new posts

  • fastqc shows strange GC content

    Hi,

    we have a 36-samples ChIP-Seq experiment with a very strange GC-content behavior (s. images)
    The samples are from yeast. This strange behavior comes in the IP samples, some of them shows (the first image is from such a file) in the fastqc report an over-representative accumulation of certain sequences, such as
    Code:
    GGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGG
    TGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGG	
    TTGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGG
    AGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGG
    But some of them don't ( the second image is from a file with no(!) over-rep. sequences at all.
    Other than that the report looks quite good with no abnormalities (except the GC content).

    I was thinking that this curve might be due to the over-rep. polyG sequences, but what about the samples where there is no apparent accumulation of such sequences. Is there a possible biological reason for this thing?

    The problem occurs only in the IP samples. Is it possible that lack of biological material for the sequencer for this samples causes the machine to put a G (or identify a G ) intread of the real nt?
    we are using a nextseq500 with a high kit from Illumina

    thanks
    Assa
    Attached Files

  • #2
    I just wanted to add that this case is even stranger, as i have this sample, where the GC content is classified as good, even though it is completely shifted to the left.

    Q: How does fastqc calculate the theoretical GC content?
    Attached Files

    Comment


    • #3
      The plots that you are referring should show normal distribution around the GC content of genome in a random library. But IP is not expected to be random so there would be a bias. Extreme GC at the start of reads could also be due to library construction method where some non-template bases are added.

      Stretches of G in NextSeq data indicates that there was not any signal in those cycles which could be due to short inserts and adapter dimers. F

      From FastQC manual:

      “This module measures the GC content across the whole length of each sequence in a file and compares it to a modelled normal distribution of GC content.”

      “An unusually shaped distribution could indicate a contaminated library or some other kinds of biased subset. A normal distribution which is shifted indicates some systematic bias which is independent of base position. If there is a systematic bias which creates a shifted normal distribution then this won't be flagged as an error by the module since it doesn't know what your genome's GC content should be.”

      Comment


      • #4
        Originally posted by frymor View Post
        Hi,

        we have a 36-samples ChIP-Seq experiment with a very strange GC-content behavior (s. images)
        The samples are from yeast. This strange behavior comes in the IP samples, some of them shows (the first image is from such a file) in the fastqc report an over-representative accumulation of certain sequences, such as
        Code:
        GGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGG
        TGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGG	
        TTGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGG
        AGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGG
        But some of them don't ( the second image is from a file with no(!) over-rep. sequences at all.
        Other than that the report looks quite good with no abnormalities (except the GC content).

        I was thinking that this curve might be due to the over-rep. polyG sequences, but what about the samples where there is no apparent accumulation of such sequences. Is there a possible biological reason for this thing?

        The problem occurs only in the IP samples. Is it possible that lack of biological material for the sequencer for this samples causes the machine to put a G (or identify a G ) intread of the real nt?
        we are using a nextseq500 with a high kit from Illumina


        thanks
        Assa
        frymor,

        You correctly identified the cause in your last statement above. The Illumina NextSeq500 (and NovaSeq) use 2 Color Chemistry for tagging and identifying bases. Using this system G's are defined by the absence of a fluorescent signal. This means that anything which can cause a cluster to stop producing a signal with result in a polyG output. A much better and more thorough explanation can be found in this QC Fail post.

        Comment


        • #5
          Stretches of G in NextSeq indicates lack of signal in those cycles and it could be result of short inserts and adapter-dimer.

          In a shot gun library GC content peak will be representative of genome GC but IP pull down is not random and some bias is expected depending on what sequences are pulled down and enriched.

          From FastQC manual:

          "This module measures the GC content across the whole length of each sequence in a file and compares it to a modelled normal distribution of GC content."

          "An unusually shaped distribution could indicate a contaminated library or some other kinds of biased subset. A normal distribution which is shifted indicates some systematic bias which is independent of base position. If there is a systematic bias which creates a shifted normal distribution then this won't be flagged as an error by the module since it doesn't know what your genome's GC content should be."

          Comment

          Latest Articles

          Collapse

          • seqadmin
            Advancing Precision Medicine for Rare Diseases in Children
            by seqadmin




            Many organizations study rare diseases, but few have a mission as impactful as Rady Children’s Institute for Genomic Medicine (RCIGM). “We are all about changing outcomes for children,” explained Dr. Stephen Kingsmore, President and CEO of the group. The institute’s initial goal was to provide rapid diagnoses for critically ill children and shorten their diagnostic odyssey, a term used to describe the long and arduous process it takes patients to obtain an accurate...
            12-16-2024, 07:57 AM
          • seqadmin
            Recent Advances in Sequencing Technologies
            by seqadmin



            Innovations in next-generation sequencing technologies and techniques are driving more precise and comprehensive exploration of complex biological systems. Current advancements include improved accessibility for long-read sequencing and significant progress in single-cell and 3D genomics. This article explores some of the most impactful developments in the field over the past year.

            Long-Read Sequencing
            Long-read sequencing has seen remarkable advancements,...
            12-02-2024, 01:49 PM

          ad_right_rmr

          Collapse

          News

          Collapse

          Topics Statistics Last Post
          Started by seqadmin, 12-17-2024, 10:28 AM
          0 responses
          34 views
          0 likes
          Last Post seqadmin  
          Started by seqadmin, 12-13-2024, 08:24 AM
          0 responses
          49 views
          0 likes
          Last Post seqadmin  
          Started by seqadmin, 12-12-2024, 07:41 AM
          0 responses
          34 views
          0 likes
          Last Post seqadmin  
          Started by seqadmin, 12-11-2024, 07:45 AM
          0 responses
          46 views
          0 likes
          Last Post seqadmin  
          Working...
          X