  • Bottom Middle Swath and other focus issues

    Is anyone else seeing a pretty frequent loss of lanes or substantially reduced lane output due to focus issues related to bubbles in lanes or other focus issues when using v3 chemistry on the HiSeq?

    In our case, there is an instrument dependency where some HiSeqs are better than others. I know Illumina is working towards a fix for this problem but I was hoping to get an idea if the frequency we are seeing this issue is unusual.

    On the worst instrument, we are seeing it in the majority of lanes, losing anything from 30-60% of the lane output. The data can be cleaned up, but cleaning things up on a tile-by-tile basis becomes more time consuming than it is worth. Best machines we see it for a cycle or two in a single lane or so with minimal, but measurable effect in the final data.

    Hoping to hear what others are seeing. Thanks.
    HudsonAlpha Institute for Biotechnology
    http://www.hudsonalpha.org/gsl

  • #2
    Yep, it's affecting a large swath (lol) of us:
    http://seqanswers.com/forums/showthread.php?t=13830

    Comment


    • #3
      We got a message from illumina regarding this quite recently (see attachment)
      Attached Files

      Comment


      • #4
        Loss of focus and data

        Originally posted by RPiddock View Post
        We got a message from illumina regarding this quite recently (see attachment)
        As far as we know, there has been no solution offered for the problem of bubbles/focus/BMS dropout yet. Anyone heard otherwise?

        Comment


        • #5
          No solution has been given to us I'm afraid.

          Comment


          • #6
            The latest HCS (1.5.1) should fix it:

            Originally posted by Illumina
            Bottom Middle Swath Update
            Jan 18 2012
            Occasionally, a HiSeq will be unable to correctly focus on a swath (1/6th of a lane on the flowcell) for a single cycle. This usually happens on the middle swath of the bottom surface of the flow cell, hence the name BMS (for "bottom middle swath"). Illumina has determined that when a bubble forms at the inlet of a lane, it can prevent accurate determination of the focus position needed to scan a swath from that lane. As a result, the data from that swath is sometimes out of focus and could result in no-call bases ("N" basecalls) for just that cycle. The wider lanes on the v3 flowcells cause this loss of focus most often on the bottom middle swath.

            Illumina R&D has developed a new version of the HiSeq firmware that reduces occurrences of BMS. This works by identifying bubbles in the region of the lane used for focusing and preventing focusing on bubbles. During testing of this firmware, they identified some changes that would lead to better identification and avoidance of bubbles. These changes were made in a new firmware version, which is currently undergoing final testing. We expect to release this firmware in February 2012.

            If you are experiencing occasional loss of focus resulting in a drop in intensity and base calls of "N" for the affected cycle, be sure you have upgraded to HCS 1.4.8 as a mitigation until the new firmware is released . If you have additional questions, please contact Illumina Technical Support at [email protected].

            Comment


            • #7
              Originally posted by GW_OK View Post
              The latest HCS (1.5.1) should fix it:
              If it's ever available.

              "Thank you for contacting Illumina Technical Support.

              We are still completing testing for this "fix". We don’t have an exact timeline at the moment, but we will let you know when the information is available. Thank you for your patience.

              If you have any additional questions, please feel free to email or call us.

              Kind regards,
              Zheng"

              Zheng Xu, Ph.D.
              Technical Applications Scientist
              Illumina Technical Support
              Illumina, Inc.
              North America: 800.809.4566
              Global contacts: www.illumina.com/support.ilmn
              Website: www.illumina.com

              Comment


              • #8
                But it is available? We've got it. It's on icom:

                Originally posted by icom
                HCS 1.5.15 is released for HiSeq and HiScan SQ

                Feb 21 2012

                HCS 1.5.15 is now available for download to your HiSeq and HiScan SQ systems. New features include support for dual indexing and warnings when there is not enough disk space available to start the sequencing run. HCS 1.5.15 also fixes a memory leak that could lead to the RTA software freezing up, and has a new tile layout that reduces the RTA memory footprint. Release notes and the installer can be found at https://my.illumina.com/download/sum...9UKStokGqhI96Q.

                For further assistance please contact [email protected].

                Comment


                • #9
                  not the 'fix'

                  The release notes say nothing about "BMS" or bubble problem.

                  Comment


                  • #10
                    Nope, 1.15.1 doesn't fix the BMS, supposedly the next release - but have heard that before.

                    Comment


                    • #11
                      Huh. Well.

                      Comment


                      • #12
                        Originally posted by GW_OK View Post
                        Huh. Well.
                        So does this mean you haven't been adversely affected by loss of focus?

                        Comment


                        • #13
                          I have had BMS dropouts, yes, but there seems to be less of it now, if any. Whether it's the new HCS, or the fiddling (something about changing the focus window) my FES did when he was here on an unrelated matter I can't say for sure.

                          Comment


                          • #14
                            We are still seeing LOTS of it, and engineers have been out here twice, disassembled the instrument and put it back together. No use. See the pretty picture. I can find lots more. Also note that this isn't bottom middle, it's bottom left and bigger than a single swath anyway. I checked our software and we're at the latest version (engineers did the installation). Guess your engineers are better fiddlers or we've got a lemon.
                            Attached Files

                            Comment


                            • #15
                              You picture is a bubble, it looks like, not exactly killing your entire swath. I've got bubbles like that for sure, but it's not the dreaded BM(LR)S.

                              Comment

                              Latest Articles

                              Collapse

                              • seqadmin
                                Advancing Precision Medicine for Rare Diseases in Children
                                by seqadmin




                                Many organizations study rare diseases, but few have a mission as impactful as Rady Children’s Institute for Genomic Medicine (RCIGM). “We are all about changing outcomes for children,” explained Dr. Stephen Kingsmore, President and CEO of the group. The institute’s initial goal was to provide rapid diagnoses for critically ill children and shorten their diagnostic odyssey, a term used to describe the long and arduous process it takes patients to obtain an accurate...
                                12-16-2024, 07:57 AM
                              • seqadmin
                                Recent Advances in Sequencing Technologies
                                by seqadmin



                                Innovations in next-generation sequencing technologies and techniques are driving more precise and comprehensive exploration of complex biological systems. Current advancements include improved accessibility for long-read sequencing and significant progress in single-cell and 3D genomics. This article explores some of the most impactful developments in the field over the past year.

                                Long-Read Sequencing
                                Long-read sequencing has seen remarkable advancements,...
                                12-02-2024, 01:49 PM

                              ad_right_rmr

                              Collapse

                              News

                              Collapse

                              Topics Statistics Last Post
                              Started by seqadmin, 12-17-2024, 10:28 AM
                              0 responses
                              32 views
                              0 likes
                              Last Post seqadmin  
                              Started by seqadmin, 12-13-2024, 08:24 AM
                              0 responses
                              48 views
                              0 likes
                              Last Post seqadmin  
                              Started by seqadmin, 12-12-2024, 07:41 AM
                              0 responses
                              34 views
                              0 likes
                              Last Post seqadmin  
                              Started by seqadmin, 12-11-2024, 07:45 AM
                              0 responses
                              46 views
                              0 likes
                              Last Post seqadmin  
                              Working...
                              X