Seqanswers Leaderboard Ad

Collapse
X
 
  • Filter
  • Time
  • Show
Clear All
new posts
  • Newie
    Junior Member
    • Jul 2010
    • 4

    HiSeq2000/ cBot poll - input vs pf

    Every instrument/ library is same, same but different...

    Curious what concentration people are loading per lane vs pf cluster densities they are getting on their cBot/ HiSeq instruments (indicating library type and size distribution would be great)?

    We are planning on sending our libraries out for HiSequencing or generating fc's in house and sending them out - curious to know what to expect?

    Thanks for your help!
  • avadhanam
    Junior Member
    • Oct 2012
    • 1

    #2
    We use 8pM as the final concentration that goes into the flow cell. Although illumina recommends 10pM, we got good coverage for 8pM too. With this you get a cluster density of 400-600.
    RNA libraries of size 320bp or so is the final size verified using bioanalyzer. Hope this info helps.

    Comment

    • pmiguel
      Senior Member
      • Aug 2008
      • 2328

      #3
      For the current run we used 16 pM as measured by the KAPA qPCR kit, using the KAPA standards. We do correct the concentration for the average size of each library.
      This puts us at 800-850 K clusters/mm^2. This gave us 92-95% pass filter for clusters.

      We had problems with both cluster density and getting even numbers of clusters/library but I think we have worked through them. What we changed:

      (1) Tightened up the size distributions of our libraries. We now do "upper cut" ampures to remove higher molecular weight products.

      (2) Stopped trying to use phiX as the qPCR standard and switched entirely to the KAPA protocol. (Which also involved using much higher dilutions of the initial samples.)

      (3) Now we take "heroic measures" to mix samples during serial dilutions. 10 full volume pumps of the solution may not have been enough. Chase that with vortex mixing.

      If this stays stable we may move our cluster density up. The instrument appears to produce excellent results at 1000 K clusters/mm^2, so we could take advantage of that -- if we were confident we would not come in 50% high...

      --
      Phillip
      Last edited by pmiguel; 10-01-2012, 10:27 AM.

      Comment

      • mama
        Junior Member
        • Sep 2008
        • 9

        #4
        upper cut

        Phillip, can you explain me in which step during the library prep do you do the "upper cut" ampures to remove the higher molecular weight product?
        Did you try to do it in the chipSeq library prep?

        Comment

        • GW_OK
          Senior Member
          • Sep 2009
          • 411

          #5
          We load 18pM, shooting for 850-950K/mm^2 with ~90% PF. It does vary a bit, usually upwards, but as long as the cpmm2 is below ~1200K it usually works out with >80% PF. You have to be careful, though, as there's a steep PF drop lurking immediately above that.

          We also do the dual size selection with Ampure beads, post-PCR, and use the KAPA kit and included standards. And I've also noticed that mixing properly is key to a good qPCR.

          Comment

          • pmiguel
            Senior Member
            • Aug 2008
            • 2328

            #6
            Originally posted by mama View Post
            Phillip, can you explain me in which step during the library prep do you do the "upper cut" ampures to remove the higher molecular weight product?
            Did you try to do it in the chipSeq library prep?
            I have not tried upper molecular weight cuts for chipseq. The "upper cut" could be done before or after enrichment PCR.

            --
            Phillip

            Comment

            • huguesparri
              Member
              • May 2008
              • 97

              #7
              7 pM for DNA (gDNa and ChIpseq) application.
              8 pM for RNAseq application.
              We are aiming 750 to 850 thousand cluster/mm².
              Libraries are quantified using Kapa kit.

              Comment

              • pmiguel
                Senior Member
                • Aug 2008
                • 2328

                #8
                Originally posted by huguesparri View Post
                7 pM for DNA (gDNa and ChIpseq) application.
                8 pM for RNAseq application.
                We are aiming 750 to 850 thousand cluster/mm².
                Libraries are quantified using Kapa kit.
                What instrument is this?

                --
                Phillip

                Comment

                • huguesparri
                  Member
                  • May 2008
                  • 97

                  #9
                  Oups, sorry: we are using a HiSeq2000 and v3 chemistry.

                  Comment

                  • pmiguel
                    Senior Member
                    • Aug 2008
                    • 2328

                    #10
                    I was asking because your numbers seem so far away from what we see. Could that be 17/18 pM you use, instead of 7/8 pM?

                    Or are you libraries much shorter than the KAPA standard and thus, effectively, of higher concentration than what the KAPA kit tells you?

                    --
                    Phillip

                    Comment

                    • huguesparri
                      Member
                      • May 2008
                      • 97

                      #11
                      Nope, it's really 7 and 8 pM.
                      When we quantify a library using the Kapa kit, we take into account the length of our fragments.
                      For RNAseq, we define a "region" on the bioanalyzer and use the "average length" calculated by the bioanalyzer.
                      For gDNA, it's more simple as there is only one sharp peak.

                      Comment

                      • genomeseeker
                        Member
                        • May 2009
                        • 16

                        #12
                        We routinely use 11 and 12pM on HiSeq2000 and generally get densities in the 650-850K/mm2 range. We quantify with the KAPA kit, as well, using their standards and adjusting concentration for avg library size.

                        Comment

                        Latest Articles

                        Collapse

                        • seqadmin
                          Pathogen Surveillance with Advanced Genomic Tools
                          by seqadmin




                          The COVID-19 pandemic highlighted the need for proactive pathogen surveillance systems. As ongoing threats like avian influenza and newly emerging infections continue to pose risks, researchers are working to improve how quickly and accurately pathogens can be identified and tracked. In a recent SEQanswers webinar, two experts discussed how next-generation sequencing (NGS) and machine learning are shaping efforts to monitor viral variation and trace the origins of infectious...
                          03-24-2025, 11:48 AM
                        • seqadmin
                          New Genomics Tools and Methods Shared at AGBT 2025
                          by seqadmin


                          This year’s Advances in Genome Biology and Technology (AGBT) General Meeting commemorated the 25th anniversary of the event at its original venue on Marco Island, Florida. While this year’s event didn’t include high-profile musical performances, the industry announcements and cutting-edge research still drew the attention of leading scientists.

                          The Headliner
                          The biggest announcement was Roche stepping back into the sequencing platform market. In the years since...
                          03-03-2025, 01:39 PM

                        ad_right_rmr

                        Collapse

                        News

                        Collapse

                        Topics Statistics Last Post
                        Started by seqadmin, Today, 10:17 AM
                        0 responses
                        7 views
                        0 reactions
                        Last Post seqadmin  
                        Started by seqadmin, 03-20-2025, 05:03 AM
                        0 responses
                        49 views
                        0 reactions
                        Last Post seqadmin  
                        Started by seqadmin, 03-19-2025, 07:27 AM
                        0 responses
                        59 views
                        0 reactions
                        Last Post seqadmin  
                        Started by seqadmin, 03-18-2025, 12:50 PM
                        0 responses
                        50 views
                        0 reactions
                        Last Post seqadmin  
                        Working...