  • Second peak on Agilent HS DNA Chip Trace

    Hi All,

    We've been attempting to create an RNA-Seq Library using a Nugen NGS Target prep. The samples are PCR amplified for 15 cycles and cleaned using a Qiagen Gel Band Extraction kit. The samples were run on an Agilent HS DNA chip and we noticed an additional peak following the sample peak around 1000bp. A second gel band extraction was performed and a follow-up HS DNA chip was run...the peaks were still there! We then tried denaturing the samples at 96C for 2 minutes, but it only made the peaks worse.

    We've been told that the second peak can be caused by too many PCR cycles - is this what anyone else has found? Has anyone else experienced this/have some insight?

    Thank you!

    AGTC Staff
    Attached Files

  • #2
    It is too much PCR. Have a look at one of my previous posts:

    Techniques and protocol discussions on sample preparation, library generation, methods and ideas


    This thread in general should answer your questions.

    Comment


    • #3
      Very helpful, thanks!

      Comment


      • #4
        Originally posted by mpalermo View Post
        Hi All,
        [...] We then tried denaturing the samples at 96C for 2 minutes, but it only made the peaks worse.

        We've been told that the second peak can be caused by too many PCR cycles - is this what anyone else has found? Has anyone else experienced this/have some insight?

        Thank you!

        AGTC Staff
        Actually had not seen a double peak post for a while. Almost a feeling of nostalgia...

        Note that your denaturation/renaturation result strongly supports the "bubble product" hypothesis over my "daisy chain" hypothesis. That is, the inserts all have the same sequences flanking them. Thus, kinetically, the most likely outcome are two unrelated library molecules annealing together at their adapters creating a "bubble" in the middle of the amplicon where the unrelated inserts will not anneal.

        As to why that would run slower than a completely anneal amplicon of the same length, I have no idea. Or at least for agarose gel electrophoresis, I have no idea. I am pretty sure that single stranded molecules tend to run more slowly on BioAnalyzer chips than double stranded molecule of the same length. So the single stranded region in the middle of the amplicon would tend to cause it to run slower.

        But as you probably saw in the other thread, pbluescript shows the double peak phenomenon on agarose gels as well.

        --
        Phillip

        Comment


        • #5
          Hi!
          I came across the same problem, you were describing in your thread when preparing ChIP-Seq libraries.
          What I dont understand is why can't you get rid of this second peak when doing gel out? I cut a 200-300 bp fraction as well as a 300-400 fraction, still when analyzing my library on the Bioanalyzer I see a second peak around 1000 bp.
          Can you or someone explain that to me.
          Thanxx



          Originally posted by mpalermo View Post
          Hi All,

          We've been attempting to create an RNA-Seq Library using a Nugen NGS Target prep. The samples are PCR amplified for 15 cycles and cleaned using a Qiagen Gel Band Extraction kit. The samples were run on an Agilent HS DNA chip and we noticed an additional peak following the sample peak around 1000bp. A second gel band extraction was performed and a follow-up HS DNA chip was run...the peaks were still there! We then tried denaturing the samples at 96C for 2 minutes, but it only made the peaks worse.

          We've been told that the second peak can be caused by too many PCR cycles - is this what anyone else has found? Has anyone else experienced this/have some insight?

          Thank you!

          AGTC Staff

          Comment


          • #6
            Originally posted by riehle View Post
            Hi!
            I came across the same problem, you were describing in your thread when preparing ChIP-Seq libraries.
            What I dont understand is why can't you get rid of this second peak when doing gel out? I cut a 200-300 bp fraction as well as a 300-400 fraction, still when analyzing my library on the Bioanalyzer I see a second peak around 1000 bp.
            Can you or someone explain that to me.
            Thanxx
            If you unintentionally strand-denature your amplicons during isolation from agarose, when they re-anneal, the statistically favored produce will be a bubble product. Bubble products can run at nearly 2x the apparent size of completely double stranded products of the same length. At least on Agilent chips.

            During purification of your DNA from agarose, did you use a heating step?

            If so, most likely that is the cause.

            I should warn you that this double peaking phenomenon is more a curiosity or an annoyance than a real problem. So burning a lot of time on it is probably not warranted.

            --
            Phillip

            Comment

            Latest Articles

            Collapse

            • seqadmin
              Advancing Precision Medicine for Rare Diseases in Children
              by seqadmin




              Many organizations study rare diseases, but few have a mission as impactful as Rady Children’s Institute for Genomic Medicine (RCIGM). “We are all about changing outcomes for children,” explained Dr. Stephen Kingsmore, President and CEO of the group. The institute’s initial goal was to provide rapid diagnoses for critically ill children and shorten their diagnostic odyssey, a term used to describe the long and arduous process it takes patients to obtain an accurate...
              12-16-2024, 07:57 AM
            • seqadmin
              Recent Advances in Sequencing Technologies
              by seqadmin



              Innovations in next-generation sequencing technologies and techniques are driving more precise and comprehensive exploration of complex biological systems. Current advancements include improved accessibility for long-read sequencing and significant progress in single-cell and 3D genomics. This article explores some of the most impactful developments in the field over the past year.

              Long-Read Sequencing
              Long-read sequencing has seen remarkable advancements,...
              12-02-2024, 01:49 PM

            ad_right_rmr

            Collapse

            News

            Collapse

            Topics Statistics Last Post
            Started by seqadmin, 12-17-2024, 10:28 AM
            0 responses
            34 views
            0 likes
            Last Post seqadmin  
            Started by seqadmin, 12-13-2024, 08:24 AM
            0 responses
            50 views
            0 likes
            Last Post seqadmin  
            Started by seqadmin, 12-12-2024, 07:41 AM
            0 responses
            35 views
            0 likes
            Last Post seqadmin  
            Started by seqadmin, 12-11-2024, 07:45 AM
            0 responses
            46 views
            0 likes
            Last Post seqadmin  
            Working...
            X